Help adjusting time in my Philips Respironics System 560p?

Post by jagzoo » Wed Nov 05, 2014 11:21 am

I am trying to time correlate my ZEO data that I have imported into Sleepyhead from ZeoViewer (with the kind and generous help of wilsonintexas).
My CPAP is 6.5 minutes fast! Is there anyway to change the time? I have the provider manual and have looked through the settings many times - can't find any date/time.

I know the ZEO is correct as I have correlated it's time with my iPhone (accurate) and written down the time that I have started both ZEO (waiting for the symbol of connection) and APAP at the same time. Times that Zeo is reporting match my recorded time. My Philips Respironics System One (60 Series) is fast approximately 6.5 min. I know I can manually change the ZEO times after exporting the .csv file from ZeoViewer and before importing into Sleepyhead, but would rather the CPAP was accurate.

I am trying to ascertain any correlation between my Events and Sleep Stages.

Any ideas - Pugsy?

Thanks in advance

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Heated Tubing; mask sometimes is TAP PAP mask; now P10 nasal pillows; use Sleepyhead software
APAP 9-12

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Help adjusting time in my Philips Respironics System 560p?

Post by Wulfman... » Wed Nov 05, 2014 12:36 pm

jagzoo wrote::?: I am trying to time correlate my ZEO data that I have imported into Sleepyhead from ZeoViewer (with the kind and generous help of wilsonintexas).
My CPAP is 6.5 minutes fast! Is there anyway to change the time? I have the provider manual and have looked through the settings many times - can't find any date/time.

I know the ZEO is correct as I have correlated it's time with my iPhone (accurate) and written down the time that I have started both ZEO (waiting for the symbol of connection) and APAP at the same time. Times that Zeo is reporting match my recorded time. My Philips Respironics System One (60 Series) is fast approximately 6.5 min. I know I can manually change the ZEO times after exporting the .csv file from ZeoViewer and before importing into Sleepyhead, but would rather the CPAP was accurate.

I am trying to ascertain any correlation between my Events and Sleep Stages.

Any ideas - Pugsy?

Thanks in advance
Nope. You'd need special equipment and software to do it. The older generations of Respironics machines (Legacy and M Series) had bad habits of losing about a minute or two a month and when the first generation of the System One came out, supposedly DID have a way of setting time.......but, they "fixed" that in subsequent firmware releases. Basically, the machines' internal clocks were set to GMT and the software compensated for the time zone differences......but apparently the internal clock/time loss is still an "issue".


Sorry.

Edit: About four years ago I acquired the cables and software to do it with the older Legacy and M Series machines, but they apparently don't work with the System One machines. And, they weren't cheap.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Help adjusting time in my Philips Respironics System 560p?

Post by archangle » Wed Nov 05, 2014 12:52 pm

The lowlifes at Respironics haven't seen fit to allow the end user to properly set the clock. Not only can you not fix the clock yourself, most DME's can't set it either and have to send it back to the factory.

Shame on you, Respironics.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
CPAPER Paul
Posts: 58
Joined: Sun Aug 31, 2014 3:09 am
Location: Atlanta, GA

Re: Help adjusting time in my Philips Respironics System 560p?

Post by CPAPER Paul » Wed Nov 05, 2014 12:57 pm

Sleepyhead has a "clock Drift Adjustment" in "Preferences" under "CPAP". I have not tried to use it, but it might be what you are looking for....

Paul

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepMapper, SleepyHead Software

User avatar
jagzoo
Posts: 340
Joined: Tue Jan 15, 2013 7:14 pm
Location: Nh

Re: Help adjusting time in my Philips Respironics System 560p?

Post by jagzoo » Wed Nov 05, 2014 1:00 pm

Oh well, Thanks anyway. Mine isn't loosing, it is actually fast 6 minutes.

Is there anyway to compensate in Sleepyhead? Or maybe Sleepyhead is trying to already and just isn't quite accurate?

Will do my best to adjust in the conversion/import of ZEO data but it doesn't seem straight forward. Maybe, it is just good enough!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Heated Tubing; mask sometimes is TAP PAP mask; now P10 nasal pillows; use Sleepyhead software
APAP 9-12

User avatar
CPAPER Paul
Posts: 58
Joined: Sun Aug 31, 2014 3:09 am
Location: Atlanta, GA

Re: Help adjusting time in my Philips Respironics System 560p?

Post by CPAPER Paul » Wed Nov 05, 2014 1:06 pm

The Sleepyhead "CPAP Clock Drift" adjustment can be either plus OR minus (hours, minutes and/or seconds). Look in Sleepyhead Preferences under the heading CPAP and in the upper left you will see the adjustment.

Good luck,

Paul

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepMapper, SleepyHead Software

User avatar
jagzoo
Posts: 340
Joined: Tue Jan 15, 2013 7:14 pm
Location: Nh

Re: Help adjusting time in my Philips Respironics System 560p?

Post by jagzoo » Wed Nov 05, 2014 2:40 pm



Sweet!!! Works!!

Thank you, thank you, thank you!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Heated Tubing; mask sometimes is TAP PAP mask; now P10 nasal pillows; use Sleepyhead software
APAP 9-12