Having trouble w waking -- is APAP at 15 cm h20 normal?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Sibbs
Posts: 68
Joined: Fri Oct 10, 2014 7:17 am

Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Sibbs » Tue Nov 04, 2014 9:06 am

Hi folks,

Still having a fair amount of trouble adjusting to therapy. I fall asleep with my mask quite comfortably, but wake every night multiple times having a hard time breathing against the machine. SleepyHead shows the pressure pretty quickly ramps up to my max setting of 15 cm h20 (min is 5).

The pressure at that value is great enough that it's pretty hard for me to exhale against it-- my brain is setting pretty strong "you're suffocating" signals that it's hard to fall back to sleep. Do folks really adjust to that level of pressure, or is something amiss with my setup or my bod?

Thanks!
Machine: Dreamstation auto BiPap
Machine: Phillips RemStar Auto with A-Flex (560P), min=9, max = 13
Mask: AirFit P10 nasal pillows

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Pugsy » Tue Nov 04, 2014 9:11 am

Are you using the exhale relief that is available on your machine?
If so at which setting?
If you are using constant pressure of 15 that probably means CFlex .

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
cathyf
Posts: 515
Joined: Mon Sep 01, 2014 11:56 am

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by cathyf » Tue Nov 04, 2014 10:10 am

I also have a lot of problems with frequent waking... One thing that I have noticed is that my pressures rise pretty quickly when I fall asleep, then the pressures fall, then I wake up. Here is last night for me:
Image
The places where the pressure stays flat are when I am awake, because I don't have much in the way of flow limitations when I'm awake. There is 45 minutes that started right before midnight, and the hour-and-a-half between 4:00 and 5:30. Between 4:30 and 4:45 I actually got up and read for awhile which made me a little sleepier. Anyway, that the 7:45 "asleep" was really more like 5-1/2...

The high pressures don't seem to bother me -- I started with the machine at 7-20 with a ramp at 4. First I turned the ramp off, and then gradually raised the minimum pressure to 11. Last night -- during the hour and a half I was awake -- I had the idea to turn my EPR back down to zero. I started with 1 (which is the factory default) and raised it to 3, thinking that might help with the awakenings. But last night I realized that the EPR makes my mask pulse on my face with each breath, and maybe THAT is waking me up? I'm going to try turning the EPR off for a couple of nights.

I have developed the habit of reaching over and hitting the STOP button on the cpap each time I wake up, so I get separate sessions generated for each wake up. (I have AUTO-start turned on, so it comes right back on with a new session with the next breath.) Here's the picture that shows the gruesome sleep fragmentation story:
Image

Anyway... I'm in the same boat as you, and am open to any suggestions as to what to try to get better sleep. My AHI is fabulous, and as long as I'm asleep the machine chases snores and flow limitations and keeps the apneas and hypopneas away. It's just that I keep waking up and am frequently having lots of trouble getting back to sleep.

Sibbs
Posts: 68
Joined: Fri Oct 10, 2014 7:17 am

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Sibbs » Tue Nov 04, 2014 10:16 am

C-flex is set at 3.
Pressure is 5 min, 15 max.

Here's a screenshot of the first segment of my night last night if that's at all helpful. Happy to supply other data as would be helpful. Thanks!

Image
Machine: Dreamstation auto BiPap
Machine: Phillips RemStar Auto with A-Flex (560P), min=9, max = 13
Mask: AirFit P10 nasal pillows

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Pugsy » Tue Nov 04, 2014 10:22 am

Just do the overall image with no segments so we can get the general idea overall of how the night went.
Be sure to include the AHI bar graphs on the left side.

Are you using APAP mode or CPAP mode? If APAP mode...try AFlex..and try a setting of 2 or 3 and see if it is easier.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by palerider » Tue Nov 04, 2014 11:07 am

cathyf wrote: The high pressures don't seem to bother me --
am open to any suggestions as to what to try to get better sleep. .
higher minimum pressure, your machine has to raise pressure to get your FL under control, then when it backs off, you get them again and wake up.... try to break the cycle.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Starlette
Posts: 1728
Joined: Mon Feb 08, 2010 9:30 pm
Location: Denver, CO

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Starlette » Tue Nov 04, 2014 11:24 am

Welcome to the forum, I have prepared a welcome wagon just for you: Hot chocolate milk and chocolate chip cookies, and peppermint schnaaps for flavoring. NOTE: You will be carded since alcohol is provided. When you arrive, please sign in (your login name) so we'll know you stopped by. Then my post will return back to the top of posts for other newbies to stop by for a treat.
Newbie Welcome Wagon

Enjoy!
Starlette

_________________
MachineMask
Additional Comments: Silent Nite Dental Appliance

Sibbs
Posts: 68
Joined: Fri Oct 10, 2014 7:17 am

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Sibbs » Tue Nov 04, 2014 11:48 am

Pugsy wrote: If APAP mode...try AFlex..and try a setting of 2 or 3 and see if it is easier.
I'm APAP mode (and pressure do vary throughout the night)-- just changed flex from the original setting of CFlex to AFlex per your suggestion and will see what happens tonight.
Machine: Dreamstation auto BiPap
Machine: Phillips RemStar Auto with A-Flex (560P), min=9, max = 13
Mask: AirFit P10 nasal pillows

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Wulfman... » Tue Nov 04, 2014 11:54 am

Sibbs wrote:C-flex is set at 3.
Pressure is 5 min, 15 max.

Here's a screenshot of the first segment of my night last night if that's at all helpful. Happy to supply other data as would be helpful. Thanks!
That's your problem. You need a higher minimum and maybe a lower maximum. Looking at your report, you appear to have fewer events at around 10 cm. I'd suggest trying to get the machine to optimize your therapy on that pressure.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by Pugsy » Tue Nov 04, 2014 12:27 pm

That one report image is just from around 11 PM to shortly after 1 AM...so about a 2 hour segment.
Need to see the entire night to have a better feel for where the pressure wants to go and stay.
You probably need a higher minimum pressure but if you are having trouble exhaling it's going to mess with sleep quality...Also with the starting pressure of 5 cm the full benefit of the exhale feature isn't utilized because the most it can reduce with AFlex is 2 cm (at a setting of 3) so it can only go down to 4 cm.

I am thinking that if you could increase that minimum starting pressure from 5 cm to something a little higher it will help both with ease in exhaling because you can get the full benefit of the exhale relief feature and probably help with overall therapy effectiveness..

I have seen it often...bring up the minimum pressure a little and the maximum pressure will reduce and it's likely that the maximum is the disturbing factor...harder to exhale against that pressure in general.
Bring up the minimum so the machine does a better job preventing the events in the first place and the machine won't have to go so high to fix them.
Don't know just how much more minimum....need to see more than 2 hours of data and the statistics side of the detailed report showing the AHI breakdown and the statistics below the AHI....that's the stuff on the left side of the daily detailed report.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Having trouble w waking -- is APAP at 15 cm h20 normal?

Post by palerider » Tue Nov 04, 2014 1:46 pm

Sibbs wrote:Here's a screenshot of the first segment of my night last night if that's at all helpful.
post the whole screenshot, not a cut and paste, click view, take screenshot.

a bit of a couple charts is like hearing a couple notes out of the song and then being told to guess it

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.