Intermittent popping sound
Intermittent popping sound
I've been using a REMstar Auto A-flex with a full face mask for 6 years. All of a sudden there's an intermittent popping sound when I inhale that keeps me awake. I can't figure out where it's coming from. Tried mask adjustments, can't find leaks anywhere. Ideas?
Re: Intermittent popping sound
Hi rick:rickpdx wrote:I've been using a REMstar Auto A-flex with a full face mask for 6 years. All of a sudden there's an intermittent popping sound when I inhale that keeps me awake. I can't figure out where it's coming from. Tried mask adjustments, can't find leaks anywhere. Ideas?
A couple of comments.
1) please register and list your equipment. It is likely that your problem is mask specific and will be best answered by a knowledgeable user of your same mask.
2) That having being said, it may be that the anti-asphyxiation valve on your mask is sticking and either should be replaced or carefully cleaned. The exact mask will help in making that determination.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Intermittent popping sound
If there is a small amount of water in your hose...it will cause a popping noise. Make sure that your CPAP is lower than your mattress so any water runs down hill back to the humidifier. (Rainout)
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
-
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: Intermittent popping sound
I've heard a popping noise using a nasal pillow system, and usually take it as a sign that the pillow is getting old. It is a fluttering pop, and comes and goes. I have a heated hose and hose cover, so rain-out is pretty rare for me. I usually just change out the pillow (mask) and the noise goes away.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Intermittent popping sound
Some F/P Flexifit FF masks do that when the AA valve gets hung up due to a cough or sneeze.
Look in the elbow if it is one of these masks.
Look in the elbow if it is one of these masks.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Intermittent popping sound
This is my thought, too. Too high of a humidity setting or not keeping the hose warm enough will cause the water vapor to condense (back into liquid) in a low spot in the hose.LSAT wrote:If there is a small amount of water in your hose...it will cause a popping noise. Make sure that your CPAP is lower than your mattress so any water runs down hill back to the humidifier. (Rainout)
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05