Minimum pressure -- trying to understand the algorithm

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Minimum pressure -- trying to understand the algorithm

Post by Pugsy » Tue Oct 21, 2014 3:04 pm

Ooops...I get the locked out symbol for dropbox.

Try imgur like Palerider uses. I just now tried it. Pretty simple.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum pressure -- trying to understand the algorithm

Post by palerider » Tue Oct 21, 2014 3:13 pm

cathyf wrote:
Pugsy wrote:Those are pressure graphs and not the flow limitation graph...can you do just a couple of flow limitation graphs with pressure graph right next to it?
Ah... dropbox is much less trouble... Same order as before:

Image

Image

Image

Image

Image

Image

Image

Image

Image

Image

Image
only, dropbox doesn't work. all the rest of see is "image' where there should be images.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum pressure -- trying to understand the algorithm

Post by palerider » Tue Oct 21, 2014 3:14 pm

Pugsy wrote:Ooops...I get the locked out symbol for dropbox.

Try imgur like Palerider uses. I just now tried it. Pretty simple.
you can even do bulk uploads

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Minimum pressure -- trying to understand the algorithm

Post by Wulfman... » Tue Oct 21, 2014 3:27 pm

Pugsy wrote:I forgot that you didn't get the for Her model.

So I dunno what is going on.
Maybe it's "Frank", again. (lurking in the bushes)


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Minimum pressure -- trying to understand the algorithm

Post by Wulfman... » Tue Oct 21, 2014 3:38 pm

Those are some wild pressure reports. I think I'd recommend narrowing the pressure range a bit.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
cathyf
Posts: 515
Joined: Mon Sep 01, 2014 11:56 am

Re: Minimum pressure -- trying to understand the algorithm

Post by cathyf » Tue Oct 21, 2014 3:39 pm

palerider wrote:
Pugsy wrote:Ooops...I get the locked out symbol for dropbox.

Try imgur like Palerider uses. I just now tried it. Pretty simple.
you can even do bulk uploads
Ok, take 3!

Image

Image

Image

Image

Image

Image

Image

Image

Image

Image

Image

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum pressure -- trying to understand the algorithm

Post by palerider » Tue Oct 21, 2014 3:44 pm

Wulfman... wrote:
Pugsy wrote:I forgot that you didn't get the for Her model.
So I dunno what is going on.
Maybe it's "Frank", again. (lurking in the bushes)
FrankenDen!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Minimum pressure -- trying to understand the algorithm

Post by Pugsy » Tue Oct 21, 2014 3:49 pm

The flow limitations are keeping the pressure up...that's why it doesn't go down to that minimum of 7.

See how active your flow limitation graph is compared to mine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum pressure -- trying to understand the algorithm

Post by palerider » Tue Oct 21, 2014 3:51 pm

Pugsy wrote:The flow limitations are keeping the pressure up...that's why it doesn't go down to that minimum of 7.

See how active your flow limitation graph is compared to mine.
yup., pressure is too low to prevent FL's.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Minimum pressure -- trying to understand the algorithm

Post by Wulfman... » Tue Oct 21, 2014 3:55 pm

And, obviously you're not sleeping very long with all that pressure going bonkers. Having to hit the "reset" periodically.

Not good.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum pressure -- trying to understand the algorithm

Post by palerider » Tue Oct 21, 2014 4:00 pm

Wulfman... wrote:And, obviously you're not sleeping very long with all that pressure going bonkers. Having to hit the "reset" periodically.

Not good.
on brief exam, it looks like most of the wakeups were after pressure dropped, which probably caused more arousals from either FL or other problems, not just higher pressure causing issues.

higher base pressure would probably calm things down a good bit.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Minimum pressure -- trying to understand the algorithm

Post by Pugsy » Tue Oct 21, 2014 4:02 pm

How come the wake ups? Any idea?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Minimum pressure -- trying to understand the algorithm

Post by Wulfman... » Tue Oct 21, 2014 4:03 pm

Or, maybe a straight pressure........


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum pressure -- trying to understand the algorithm

Post by palerider » Tue Oct 21, 2014 4:09 pm

Wulfman... wrote:Or, maybe a straight pressure........
ones one figures out what it should be.

you must *love* the cpap-check feature of the 460/560.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Minimum pressure -- trying to understand the algorithm

Post by Wulfman... » Tue Oct 21, 2014 4:11 pm

palerider wrote:
Wulfman... wrote:Or, maybe a straight pressure........
ones one figures out what it should be.

you must *love* the cpap-check feature of the 460/560.
I have no idea what you're referring to, but I'd suggest a straight pressure of about 9 or 10 cm.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow