Question re: F&P Auto - Water in Hose

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ldobler74
Posts: 8
Joined: Mon May 13, 2013 10:16 pm

Question re: F&P Auto - Water in Hose

Post by ldobler74 » Fri Oct 10, 2014 10:54 am

Hello!

Using the machine I find that sometimes in the morning I'll be hearing water gurgle and sure enough there is some water in the "L Connector" (my own term but the part that connects the machine to the hose) and I have to tilt the machine and let some of the water out. Is this normal or should I think about getting the machine looked at?

Thanks in advance!

P.S. I wanted to ping here first because it is a PITA to get to my DME provider hehe

_________________
Mask

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Question re: F&P Auto - Water in Hose

Post by Julie » Fri Oct 10, 2014 10:58 am

Have you got the machine positioned lower than your head/mask? The water can't run into the mask from below.

User avatar
ldobler74
Posts: 8
Joined: Mon May 13, 2013 10:16 pm

Re: Question re: F&P Auto - Water in Hose

Post by ldobler74 » Fri Oct 10, 2014 11:07 am

The machine is a little lower than the mask but relatively even. Should that change? Thanks!

_________________
Mask

User avatar
fat rat
Posts: 23
Joined: Mon Sep 01, 2014 5:39 am
Location: Germany, Rhein-Main-Area

Re: Question re: F&P Auto - Water in Hose

Post by fat rat » Fri Oct 10, 2014 11:28 am

Have you got a heated hose as displayed if clicking your profil? Perhaps it does not work or the setting should be changed?

Greetings
fat rat

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Weinmann JOYCEeasy FF Mask; Software: SleepyHead, Encore Basic

User avatar
ldobler74
Posts: 8
Joined: Mon May 13, 2013 10:16 pm

Re: Question re: F&P Auto - Water in Hose

Post by ldobler74 » Fri Oct 10, 2014 11:30 am

I do. It is marked as being "on" from the settings menu.

_________________
Mask

User avatar
fat rat
Posts: 23
Joined: Mon Sep 01, 2014 5:39 am
Location: Germany, Rhein-Main-Area

Re: Question re: F&P Auto - Water in Hose

Post by fat rat » Sat Oct 11, 2014 4:56 pm

It may be difficult to determine the function of the heated tube, since the temperature should has just so few above room temperature. It is best to compare the hose temperature by touch another object.
If the heated hose works, can you increase the temperature?

Greetings
fat rat

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Weinmann JOYCEeasy FF Mask; Software: SleepyHead, Encore Basic

User avatar
Pugsy
Posts: 65381
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question re: F&P Auto - Water in Hose

Post by Pugsy » Sat Oct 11, 2014 5:03 pm

I am not sure how the F & P humidifier is constructed but my S9 will sometimes do that gurgling thing if I have over filled the water chamber and the air flowing through the water chamber blows a little water into the area where the hose connects to the humidifier. If the opening in the chamber is anywhere near the water level line...this might be happening in your situation. It would be odd that this moisture at this area right next to the humidifier would be condensation so close to the humidifier chamber...not impossible but odd.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Sat Oct 11, 2014 8:01 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

mgaggie
Posts: 870
Joined: Sun Apr 21, 2013 4:30 am
Location: Victoria, Australia

Re: Question re: F&P Auto - Water in Hose

Post by mgaggie » Sat Oct 11, 2014 7:05 pm

I concur with Pugsy.

Sounds like you have filled the tank a bit too much (easy to do, I've done it a couple of times). I would also move the machine so its lower than your head.

User avatar
ldobler74
Posts: 8
Joined: Mon May 13, 2013 10:16 pm

Re: Question re: F&P Auto - Water in Hose

Post by ldobler74 » Sun Oct 12, 2014 12:59 pm

Thanks everyone!

_________________
Mask