Your pressure is not quite sufficient to prevent the snores which are the pre cursor warning signs of an impending airway trying to collapse....small reductions in air flow that don't meet the criteria for hyponea or obstrutive apnea.
Since your AHI is low for the most part those minor flow reductions don't seem to progress in reduction or duration to meet the criteria for hyponea or OA flag. Why sometimes everything okay and other time the clusters and the higher AHI?
Usually the 2 prime suspects are either supine sleeping or REM stage sleep or maybe a little of both. It's common for OSA to be worse when supine sleeping and/or REM stage sleep. During those times you see the clusters and higher AHI the pressure is more than just a little insufficient.
There RERAs are flagged due to the machine sensing a specific breathing pattern that might point to some sort of arousal..usually associate with less than optimal sleep quality and frequent awakenings.
How come so much fragmented sleep? Some nights show a lot of breaks in therapy...any idea why?
The Monday Oct 6 report below is probably the worst I could see. One wouldn't load for me for some reason or other.
Looks like you need a little more pressure at times and sometimes you need it a lot more than other times.
If you will copy the IMG address and just paste that address into the body of your text here the image will show up straight away and we don't have to click on links.