Humidifier chamber keeps popping open

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Finnc
Posts: 70
Joined: Thu Sep 25, 2014 9:54 pm
Location: Charlotte

Humidifier chamber keeps popping open

Post by Finnc » Sat Oct 04, 2014 8:12 pm

For some reason last night I couldn't use my machine because every time I would turn it on the humidifier chamber would pop open. I checked the water and it was fine, I checked to make sure the water chamber was placed in correctly and it was. I looked over the manual and couldn't find anything about this, so I though I would ask here if anyone had ever had this happen.

TIA

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Software resmed S9 wireless

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Humidifier chamber keeps popping open

Post by chunkyfrog » Sat Oct 04, 2014 8:16 pm

When you close the lid, make sure the little slide button is all the way to the left.
It usually needs a little nudge to get there. The lid will pop open if the catch is anywhere but all the way to the left.
Been there--many times.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
LSAT
Posts: 13374
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Humidifier chamber keeps popping open

Post by LSAT » Sat Oct 04, 2014 8:49 pm

chunkyfrog wrote:When you close the lid, make sure the little slide button is all the way to the left.
It usually needs a little nudge to get there. The lid will pop open if the catch is anywhere but all the way to the left.
Been there--many times.
I agree with the frog...I have had the same problems. Just make sure that the catch clicks when you close it.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Humidifier chamber keeps popping open

Post by archangle » Sat Oct 04, 2014 11:46 pm

Mine's a little finicky as well. Sometimes it doesn't want to latch.

Take out the water tank and see if the humidifier latches properly without the water tank inside.

Check the clear silicone rubber gasket inside the humidifier lid for problems. Look for any irregularities inside the humidifier.

The lid does detach from the humidifier, so maybe it could be misaligned. Warning, it can be difficult to reattach if you detach it.

Be sure the water tank is all snapped together properly.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

TyroneShoes
Posts: 107
Joined: Sun Sep 07, 2014 2:45 pm

Re: Humidifier chamber keeps popping open

Post by TyroneShoes » Sun Oct 05, 2014 10:48 am

That latch does appear to be the weakest link in an otherwise well-designed system

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead on Mac OS, EPR at 3, no ramp