EPR versus BiPap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ATXFerret
Posts: 31
Joined: Fri Apr 18, 2014 5:56 pm

EPR versus BiPap

Post by ATXFerret » Thu Oct 02, 2014 8:15 am

My sleep doctor, after the most recent sleep study, want's to put me on a bipap. Thing is, she wants to put it on a setting of 14/18. I'm on a ResMed CPAP right now that is set at 18 with EPR at max. Is there really a point in going to a bipap from that?

User avatar
Pugsy
Posts: 65375
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR versus BiPap

Post by Pugsy » Thu Oct 02, 2014 8:26 am

Why the need? Aerophagia maybe?

But there is a little bit of difference between the 3 EPR drop and the 4 cm drop that is proposed with those settings.
To me 4 is more comfortable than 3 but isn't like 3 is horrible but then I don't have any aerophagia issues or centrals to worry about. I use 4 or 5 PS (the difference between inhale and exhale) just for the comfort factor.

Also I did try EPR at 3 (CPAP mode on my S9 VPAP) and used my Bilevel machine with PS of 3...while there wasn't a huge difference in how it "felt" I thought I felt a little difference in the timing of the reduction and the response time back up to inhale. Not a deal breaker difference though and it may have been mostly in my head.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Thu Oct 02, 2014 8:33 am, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65375
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR versus BiPap

Post by Pugsy » Thu Oct 02, 2014 8:27 am

Oh...one other thing...maybe the doc is thinking might need more than 20 cm at some time in the future...the bilevel machine will go to 25.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

ATXFerret
Posts: 31
Joined: Fri Apr 18, 2014 5:56 pm

Re: EPR versus BiPap

Post by ATXFerret » Thu Oct 02, 2014 8:30 am

Maybe. I just hate to spend that kind of money on it right now on a maybe. But I guess that's how medicine works in America, isn't it?

User avatar
VikingGnome
Posts: 591
Joined: Fri Feb 03, 2012 6:46 pm
Location: Mississippi

Re: EPR versus BiPap

Post by VikingGnome » Thu Oct 02, 2014 9:09 pm

You get EPR or "Easy Breathe" with ResMed VPAP anyway. I started out with CPAP at 15. Went up to 17. At that point my doctor prescribed BiLevel for me. My current pressure is 22/18 with EPR at 2. So that essentially means 22/16 with EPR setting.

The IPAP (higher number) controls hypopneas. The EPAP (lower number) controls apneas. If your current pressure on CPAP is controlling apneas but not hypopneas, you need a higher IPAP. If you are getting close to 20, you may need BiLevel to control both hyponeas AND apneas.

_________________
Mask: AirFit™ F10 For Her Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure Setting 22/18; Pregnancy pillow to keep me sleeping on side

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: EPR versus BiPap

Post by palerider » Thu Oct 02, 2014 9:58 pm

VikingGnome wrote:You get EPR or "Easy Breathe" with ResMed VPAP anyway. I started out with CPAP at 15. Went up to 17. At that point my doctor prescribed BiLevel for me. My current pressure is 22/18 with EPR at 2. So that essentially means 22/16 with EPR setting.
you keep saying that, ignoring the fact that you're *wrong*.

EPR is not available on resmed vpaps in vpap mode, only in cpap mode.

easy breathe is NOT the same as EPR.

EPR is a limited bi-level function, it would be redudant, and *stupid* to have it on vpap mode.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
VikingGnome
Posts: 591
Joined: Fri Feb 03, 2012 6:46 pm
Location: Mississippi

Re: EPR versus BiPap

Post by VikingGnome » Thu Oct 02, 2014 10:29 pm

palerider wrote:
VikingGnome wrote:You get EPR or "Easy Breathe" with ResMed VPAP anyway. I started out with CPAP at 15. Went up to 17. At that point my doctor prescribed BiLevel for me. My current pressure is 22/18 with EPR at 2. So that essentially means 22/16 with EPR setting.
you keep saying that, ignoring the fact that you're *wrong*.

EPR is not available on resmed vpaps in vpap mode, only in cpap mode.

easy breathe is NOT the same as EPR.

EPR is a limited bi-level function, it would be redudant, and *stupid* to have it on vpap mode.
Then what do you thiink "Easy Breathe" is? By manual it functions the same a EPR. Here is the Table of Contents of the Clinician Manual for all ResMed VPAP models. Note Page 5 discusses EPR.

Image

_________________
Mask: AirFit™ F10 For Her Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure Setting 22/18; Pregnancy pillow to keep me sleeping on side

Guest

Re: EPR versus BiPap

Post by Guest » Thu Oct 02, 2014 10:58 pm

If you are paying out of pocket consider buying used here on the forum or from Craigslist.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: EPR versus BiPap

Post by palerider » Fri Oct 03, 2014 12:00 am

VikingGnome wrote:Then what do you thiink "Easy Breathe" is? By manual it functions the same a EPR. Here is the Table of Contents of the Clinician Manual for all ResMed VPAP models.
1) that's not nearly all vpap models, some that aren't listed, VPAP ST-A, VPAP Adapt, VPAP COPD, VPAP Tx.

2) was was very recently pointed out, here: viewtopic.php?f=1&t=101012&p=942808&hil ... he#p942706 easy breathe is a pressure SHAPING technique that changes the abruptness of the transition between epap and ipap, pictures of an example are in a post by robysue that are referenced in that post.

easy breathe is something that is *ADDED TO* EPR, but it is NOT EPR, and it is not a pressure REDUCTION, like EPR is.

and, it's not available on the ST, or Vauto modes of the VPAP ST or the VPAP Auto.

and none of them offer EPR in any VPAP mode, because it's redundant.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Oct 03, 2014 2:56 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

ATXFerret
Posts: 31
Joined: Fri Apr 18, 2014 5:56 pm

Re: EPR versus BiPap

Post by ATXFerret » Fri Oct 03, 2014 1:45 pm

Guest wrote:If you are paying out of pocket consider buying used here on the forum or from Craigslist.
I'm not paying out of pocket, but I'm sure there will still be a deductible. At least a few hundred.

ATXFerret
Posts: 31
Joined: Fri Apr 18, 2014 5:56 pm

Re: EPR versus BiPap

Post by ATXFerret » Fri Oct 03, 2014 1:47 pm

VikingGnome wrote:You get EPR or "Easy Breathe" with ResMed VPAP anyway. I started out with CPAP at 15. Went up to 17. At that point my doctor prescribed BiLevel for me. My current pressure is 22/18 with EPR at 2. So that essentially means 22/16 with EPR setting.

The IPAP (higher number) controls hypopneas. The EPAP (lower number) controls apneas. If your current pressure on CPAP is controlling apneas but not hypopneas, you need a higher IPAP. If you are getting close to 20, you may need BiLevel to control both hyponeas AND apneas.
My whole point was that, with the EPR on my current CPAP, I'm already around 18/16, so what's the point in going to a bipap that is also 18/16?

Guest

Re: EPR versus BiPap

Post by Guest » Fri Oct 03, 2014 2:49 pm

ATXFerret wrote:My whole point was that, with the EPR on my current CPAP, I'm already around 18/16, so what's the point in going to a bipap that is also 18/16?
Have you asked your doc that question?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: EPR versus BiPap

Post by palerider » Fri Oct 03, 2014 2:58 pm

ATXFerret wrote:
VikingGnome wrote:You get EPR or "Easy Breathe" with ResMed VPAP anyway. I started out with CPAP at 15. Went up to 17. At that point my doctor prescribed BiLevel for me. My current pressure is 22/18 with EPR at 2. So that essentially means 22/16 with EPR setting.

The IPAP (higher number) controls hypopneas. The EPAP (lower number) controls apneas. If your current pressure on CPAP is controlling apneas but not hypopneas, you need a higher IPAP. If you are getting close to 20, you may need BiLevel to control both hyponeas AND apneas.
My whole point was that, with the EPR on my current CPAP, I'm already around 18/16, so what's the point in going to a bipap that is also 18/16?
the only real advantage of going to a bipap at 18/16 is that you can twiddle with the behavior of the epap >ipap>epap transition, making it more or less sensitive, changing the timing, etc. not a big deal for the average patient, but it can be a big deal for someone with obstrucive/restrictive lung conditions.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: EPR versus BiPap

Post by Jay Aitchsee » Fri Oct 03, 2014 5:17 pm

VikingGnome wrote:The IPAP (higher number) controls hypopneas. The EPAP (lower number) controls apneas.
Could you provide a reference for that statement, please? I'm interested in learning more.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video