IS IT REALLY NEW

Post by Sly1234 » Fri Sep 26, 2014 3:40 pm

I suspect my new CPAP Resmed s9 is not really new, possibly preowned, but sold as new.
Does anyone know how I can confirm this. I have a dishonest supplier. Billed my insurance twice, once for mask kit and again on the same invoice for the individual components. Also kept two of the nose pillows, kit has s,m,l. Only gave me one. They did this twice. I ordered another mask this one containing xs,s,m. They kept xs and m. They cut open the plastic and resealed it to make it look like it was not tampered with.
They rent cpaps and now I wonder if my S9 is new or not.
How can I find "blower hours"? Is that the same as run hours. And how do I know if my software version is current.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: IS IT REALLY NEW

Post by palerider » Fri Sep 26, 2014 3:48 pm

hold down both silver buttons till the display changes, then use the knob to scroll down to the wrench, look at the hours shown.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: IS IT REALLY NEW

Post by archangle » Fri Sep 26, 2014 3:53 pm

Most of the time, insurance requires CPAP machines to be a monthly rental for a few months. You are not entitled to a new machine when you're renting it.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

Sly1234
Posts: 5
Joined: Sun Jul 27, 2014 8:45 am

Re: IS IT REALLY NEW

Post by Sly1234 » Fri Sep 26, 2014 4:21 pm

This was a purchase, not a rental. Even if it was a rental, its a rent to buy. And still should be new.

Sly1234
Posts: 5
Joined: Sun Jul 27, 2014 8:45 am

Re: IS IT REALLY NEW

Post by Sly1234 » Fri Sep 26, 2014 4:31 pm

palerider wrote:hold down both silver buttons till the display changes, then use the knob to scroll down to the wrench, look at the hours shown.
Thank you for your reply, I believe the wrench tells me my run hours, the Resmed rep, told me there is another measurement of hours, called service hours which represents the blower hours in total and the hours appear in red. Those can not be reset and would tell me if there is a discrepancy. I have gone into the clinician mode but can not find the area that tells me blower hours and current software being used.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: IS IT REALLY NEW

Post by palerider » Fri Sep 26, 2014 4:53 pm

Sly1234 wrote:
palerider wrote:hold down both silver buttons till the display changes, then use the knob to scroll down to the wrench, look at the hours shown.
Thank you for your reply, I believe the wrench tells me my run hours, the Resmed rep, told me there is another measurement of hours, called service hours which represents the blower hours in total and the hours appear in red. Those can not be reset and would tell me if there is a discrepancy. I have gone into the clinician mode but can not find the area that tells me blower hours and current software being used.
that's all there is, you can't reset run hours.... if you get something outta your resmed rep that's different, please share it with us.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Sep 26, 2014 5:25 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: IS IT REALLY NEW

Post by Jay Aitchsee » Fri Sep 26, 2014 5:09 pm

viewtopic.php?f=1&t=97732&p=908990#p908990
STL Mark wrote:I got in a new S9 Autoset this week fresh from ResMed. SX474-1201 is the current level. Prior to this what I would see is SX474-0907.
Image
"Used Hrs" in the Sleep Report section can be reset, not "Run Hrs" shown above.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: IS IT REALLY NEW

Post by TangledHose » Fri Sep 26, 2014 5:51 pm

Just a reminder, the run hours will be a little higher than the used hours as the blower continues to run at a low setting after you shut of the therapy to help dry out the machine and components, so the run hours will tend to be slightly higher.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: IS IT REALLY NEW

Post by archangle » Fri Sep 26, 2014 8:47 pm

The S9 "Used hours" is part of the daily, weekly, monthly, 3 month, or 1 year data. It can be reset.

Apparently, "Run hours" can't be reset. At least doing an "erase data" function didn't reset it.

I couldn't find a "service" hours. Maybe there's some ubersecret menu or software that will show this.

Maybe the person you were talking to was thinking about Respironics PRS1 machines. PRS1 machines have "Therapy, Blower, and Machine" hours. Therapy is average hours. Blower and Machine are lifetime totals. Blower can be reset, machine can't.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: IS IT REALLY NEW

Post by Jay Aitchsee » Sat Sep 27, 2014 7:36 am

Sly1234 wrote:And how do I know if my software version is current.
In case you didn't catch it, software (SW) version SX474-1201 was the most current known at the time of this posting, viewtopic.php?f=1&t=97732&p=908990#p908990 , of May 22, 2014.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: IS IT REALLY NEW

Post by Krelvin » Sat Sep 27, 2014 9:09 pm

Just added to that post tonight, my new machine came with SX474-1203 but it is a bilevel. Not sure if that makes a difference.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: IS IT REALLY NEW

Post by palerider » Sat Sep 27, 2014 9:45 pm

Krelvin wrote:Just added to that post tonight, my new machine came with SX474-1203 but it is a bilevel. Not sure if that makes a difference.
it would be interesting to compare sw revs of different models... hmm.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Sly1234
Posts: 5
Joined: Sun Jul 27, 2014 8:45 am

Re: IS IT REALLY NEW

Post by Sly1234 » Mon Sep 29, 2014 7:10 am

looks like under service, run hours are 96, software is SX474-0907
Used hours are 94.4
I am not certain if run hours=blower hours, but I will call again.

Resmed said from my serial number, they can only tell me when and to whom it was shipped

It's probably fine but with all the other improprieties, I started to doubt the S9 Status.
Does anyone have the S10?

Thanks for all your replies

User avatar
torontoCPAPguy
Posts: 1015
Joined: Mon Dec 28, 2009 11:27 am
Location: Toronto Ontario/Buffalo NY

Re: IS IT REALLY NEW

Post by torontoCPAPguy » Mon Sep 29, 2014 8:38 am

I have seen discussions regarding the resetting of both 'used hours' and 'motor/total hours' being reset and have never seen an answer to resetting total hours on the motor. The resetting of 'used hours' is fairly straightforward and generally part of the clinician menu. While one can view the other (total hours/motor hours/etc.) nobody has come up with the process to reset this and I suspect that it is for exactly the reason that you mention and that I, too, have experienced, but not quite to the same extent. I bought what was supposed to be a brand new S9 Auto from a large dealer online some years ago shortly after they came out. I actually bought TWO of them. One as a backup as they were pretty inexpensive at the time. The first one had a faint plastic odour while the second one, although it pretty much looked pristine, did NOT have that plastic "new equipment" smell but rather a very mild "perfume" smell. I investigated to see if the motor hours reading could be reset and have always kept my eye open for that.... but have never seen a means of doing a total reset INCLUDING reset of the motor hours. My curiousity will have the better of me until I see someone post that process and try it out myself. As a computer guy with a computer guy for a #1 son (taking Computer Software Engineering at University) we both believe that if it can be written to and read, there is likely a means of changing the data or resetting it... the question is the process.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Respironics Everflo Q infusing O2 into APAP line to maintain 95% SaO2; MaxTec Maxflo2 Oxygen Analyzer; Contec CMS50E Recording Pulse Oxymeter
Fall colours. One of God's gifts. Life is fragile and short, savour every moment no matter what your problems may be. These stunning fall colours from my first outing after surviving a month on life support due to H1N1.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: IS IT REALLY NEW

Post by palerider » Mon Sep 29, 2014 8:54 am

torontoCPAPguy wrote: The first one had a faint plastic odour while the second one, although it pretty much looked pristine, did NOT have that plastic "new equipment" smell but rather a very mild "perfume" smell.
perfumey smell comes from some silicone.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.