Tubing Size for Nuance Pro gel pillows mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MalabarFlash
Posts: 26
Joined: Wed Nov 13, 2013 12:22 pm

Tubing Size for Nuance Pro gel pillows mask

Post by MalabarFlash » Sat Aug 23, 2014 8:27 am

Philips Respironics RemStar Pro with C-Flex+ (460P)

I am using a 12 ft. 22mm hose. However, the pillows mask I'm now trying has an attached short 15mm hose. What is the correct tubing setting for my CPAP? My respiratory therapist said 22? Also, I would like the source of the information. I could not find anything in my manuals.

User avatar
Pugsy
Posts: 65167
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Tubing Size for Nuance Pro gel pillows mask

Post by Pugsy » Sat Aug 23, 2014 9:09 am

I can't give you an official document source as I have never seen one but machine's hose selection size is always based on the diameter of the longest hose and not any short hose that is part of the mask. I have never looked for that source because I don't need official documentation. It's common sense. If you just have to have it.....start pouring through all the Respironics literature available online. Call up Respironics and ask them...if they will talk to you.

You have a 60 series machine. If you had the heated hose with this machine and attached it the machine would automatically sense that hose and would default to the 15 mm size hose which is the size of the heated hose.

Your RT is correct.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.