Mouth breathing as clear events

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 5:25 am

Last two nights I've had AHIs of 2.45 of which 1.5 was clear airway apneas. With a nasal pillow and some allergy congestion could I be mouth breathing at times and triggering false clear airway events? Obstructive and hypopneas make up the other 1, no unclassifieds.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Mouth breathing as clear events

Post by JDS74 » Fri Aug 22, 2014 5:36 am

What does you leak rate look like around the time of a CA event?

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 7:24 am

Sub 20 but flow rate is peaking. What does that imply?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

User avatar
LSAT
Posts: 13328
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Mouth breathing as clear events

Post by LSAT » Fri Aug 22, 2014 8:01 am

Why do you assume that the Clear Airway events are false? IMO..you are nit picking. If your AHI is low and you are sleeping and feeling well, don't worry about the details.

User avatar
Pugsy
Posts: 65064
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Mouth breathing as clear events

Post by Pugsy » Fri Aug 22, 2014 8:17 am

Remember that a few centrals are considered normal to have show up anyway. Like sleep onset centrals or sleep stage transition centrals.
Could mouth breathing alone get flagged by mistake? Not unless you pause your breathing itself. Hold your breath type of thing.
You can look at the flow rate and see if the flow is reduced.
Could you be mouth breathing and be holding your breath inadvertently while awake or semi awake ? Sure and then the machine would likely give that pause a central flag.
Mouth breathing doesn't interrupt the flow of air and the machine looks for interruptions/reduction of air flow. If the mouth breathing isn't huge then it may not even show up on the leak graphs. I have tested this myself while awake and have proof.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Mouth breathing as clear events

Post by BlackSpinner » Fri Aug 22, 2014 8:41 am

Are you taking allergy meds? If you are that could be causing the problem. I sleep lighter and have more disruptions when I use them.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 9:27 am

No allergy meds but not sleeping as well this week. More congestion with seasonal allergies I think.

So, Clear Airway and centrals are the same thing?

If I get a congested nose and then stop breathing until my mouth opens reflexively will that look like this?

Not hugely worried LSAT but annoying that I've gone from 1 to 2+ on AHI...

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

User avatar
Pugsy
Posts: 65064
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Mouth breathing as clear events

Post by Pugsy » Fri Aug 22, 2014 9:32 am

englandsf wrote:So, Clear Airway and centrals are the same thing?
Yes.

When SleepyHead was first written it was written for a Respironics machine which uses the term "Clear Airway" and the ResMed machines were added later and ResMed uses the term Central. They are both open airway cessation of breathing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 10:32 am

Thanks

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mouth breathing as clear events

Post by palerider » Fri Aug 22, 2014 12:17 pm

englandsf wrote: With a nasal pillow and some allergy congestion could I be mouth breathing at times and triggering false clear airway events?
just a point of order, if your nose was congested to the point where you'd have to breath through your mouth, and the machine scored an apnea, it would score an *obstructive* apnea, because the airway to your lungs would be blocked.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 12:27 pm

Good point pr, but I do get some weird air flow when my mouth is open and my nose is partially blocked.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 12:33 pm

Maybe I'm adjusting the pillows and creating another confusing or odd flow

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mouth breathing as clear events

Post by palerider » Fri Aug 22, 2014 1:03 pm

englandsf wrote:Maybe I'm adjusting the pillows and creating another confusing or odd flow
you'd have high leaks if that were it... did you?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Mouth breathing as clear events

Post by englandsf » Fri Aug 22, 2014 1:06 pm

Nope, just high flow.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mouth breathing as clear events

Post by palerider » Fri Aug 22, 2014 2:06 pm

englandsf wrote:Nope, just high flow.
and do you know what that means?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.