ResMed Air Sense 10 Service Manual

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

ResMed Air Sense 10 Service Manual

Post by Reuven Gruber » Sun Apr 09, 2017 3:37 pm

Does anyone know where I can get the ResMed Air Sense 10 Service Manual (how to do maintenance and repairs)

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Air Sense 10 Service Manual

Post by palerider » Sun Apr 09, 2017 3:43 pm

Reuven Gruber wrote:Does anyone know where I can get the ResMed Air Sense 10 Service Manual (how to do maintenance and repairs)
no.

what do you want to know? I know someone who's taken a bunch of them apart.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
WRX03
Posts: 137
Joined: Fri Apr 29, 2016 9:37 am
Location: Roanoke, Virginia

Re: ResMed Air Sense 10 Service Manual

Post by WRX03 » Mon Apr 10, 2017 6:06 am

Other than cleaning why? What item is broken? You shouldn't be able to buy any internal parts. Plus replacing many parts you have to run a calibration on it afterwards, also data may need reloaded before it will work again. Once you pop it open ResMed wants certain seals replaced. There are two versions of the main circuit board; if you have the older you have to replace the Cellular Access Module too. The outlet silicone seal is user/DME replaceable. It's the seal between the water chamber and outlet port. Part# 37310

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Actually PCA is a "For Her". It's DIY piece together from spare parts CPAP. It's multi color! LOL

User avatar
LSAT
Posts: 13359
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ResMed Air Sense 10 Service Manual

Post by LSAT » Mon Apr 10, 2017 7:55 am

Reuven Gruber wrote:Does anyone know where I can get the ResMed Air Sense 10 Service Manual (how to do maintenance and repairs)
Here is the Clinical Manual, but I don't think you will find anything to show you how to do maintenance or repairs.
http://osahelp.com/clinician-manuals/S9 ... al-eng.pdf

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: ResMed Air Sense 10 Service Manual

Post by chunkyfrog » Mon Apr 10, 2017 9:33 am

What is broken?
Are you planning to cannibalize one to fix another?
Please share images. The ghoul in me drools.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: ResMed Air Sense 10 Service Manual

Post by prodigyplace » Mon Apr 10, 2017 10:03 am

LSAT wrote:
Reuven Gruber wrote:Does anyone know where I can get the ResMed Air Sense 10 Service Manual (how to do maintenance and repairs)
Here is the Clinical Manual, but I don't think you will find anything to show you how to do maintenance or repairs.
http://osahelp.com/clinician-manuals/S9 ... al-eng.pdf
I think that is for the S9 manual, not the S10.
I believe this is the S10 one. https://www.respshop.com/manuals/ResMed ... %20her.pdf

The link says "for her" but the manual is for both.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

Guest

Re: ResMed Air Sense 10 Service Manual

Post by Guest » Tue Apr 11, 2017 3:39 pm

I have an out of warranty machine where the autostart function doesn't work. I would like basic dissembly instruction instructions and a good block diagram

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed Air Sense 10 Service Manual

Post by Reuven Gruber » Tue Apr 11, 2017 3:40 pm

I have an out of warranty machine where the autostart function doesn't work. I would like basic dissembly instruction instructions and a good block diagram

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed Air Sense 10 Service Manual

Post by Pugsy » Tue Apr 11, 2017 3:44 pm

When did the problem start?
How long have you been using the P10 nasal pillow mask?
What is your starting pressure?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Air Sense 10 Service Manual

Post by palerider » Tue Apr 11, 2017 7:26 pm

Reuven Gruber wrote:I have an out of warranty machine where the autostart function doesn't work. I would like basic dissembly instruction instructions and a good block diagram
the smartstart function is twofold, both sensing when you put the mask on and start breathing, and sensing when you take it off.

this function involves calculations from the flow and pressure sensors, when you put the mask on, the machine senses the tiny amount of flow that is caused by inhaling or exhaling, and starts up, when you take the mask off, it measures the greatly reduced pressure and increased flow, and tries to guess whether it's a bad leak or mask off.

when I first switched to the P10, I was disappointed that my machine didn't turn off by itself, then I remembered I'd forgotten the mask type setting, I changed that from full face to pillow, and it's been working ever since.

the smaller openings of the pillow mask didn't allow enough pressure drop for the machine to think "ah, mask off" compared to the larger opening and flow of the full face mask.

in other words, there's nothing you can do about this, because if either of the two sensors involved were faulty, the machine wouldn't run in the first place, it would fault and pop a service message.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed Air Sense 10 Service Manual

Post by Reuven Gruber » Wed Apr 12, 2017 3:00 am

The function worked fine for a long time and then stopped functioning

Smart start is NOT working with a full face mask as well as with pillows

I have another Air Sense 10 and the smart start function works fine with both the pillows and the full face mask

My settings are 4 for the starting pressure and 8 for the maximum pressure

PALERIDER - please put me in touch with the person you mentioned who has taken a bunch of these machines apart

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
WRX03
Posts: 137
Joined: Fri Apr 29, 2016 9:37 am
Location: Roanoke, Virginia

Re: ResMed Air Sense 10 Service Manual

Post by WRX03 » Wed Apr 12, 2017 6:03 am

Reuven Gruber wrote:The function worked fine for a long time and then stopped functioning

Smart start is NOT working with a full face mask as well as with pillows

I have another Air Sense 10 and the smart start function works fine with both the pillows and the full face mask

My settings are 4 for the starting pressure and 8 for the maximum pressure

PALERIDER - please put me in touch with the person you mentioned who has taken a bunch of these machines apart
When a PAP comes in with this issue I:
1. Check settings
2. Check for water/dust/dropped damage
3. Run a calibration
4. Change main PCB

The Air 10 is pretty tight in design there is only about a 1/4" between the sensors on the PCB and ports. If they didn't have a leak before, the silicone seals weren't messed with, or the PAP wasn't dropped; I would give it a 99.99% change a leak wouldn't just start on it's own. Take several deep breathes and see if it triggers on. If it does a re-cal is all it may need.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Actually PCA is a "For Her". It's DIY piece together from spare parts CPAP. It's multi color! LOL

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: ResMed Air Sense 10 Service Manual

Post by prodigyplace » Wed Apr 12, 2017 6:40 am

Reuven Gruber wrote:
PALERIDER - please put me in touch with the person you mentioned who has taken a bunch of these machines apart
You probably shroud PM him using the button under his avatar. I doubt he would post that information publicly.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

Guest

Re: ResMed Air Sense 10 Service Manual

Post by Guest » Wed Apr 12, 2017 11:18 am

WRX03 wrote:
Reuven Gruber wrote:The function worked fine for a long time and then stopped functioning

Smart start is NOT working with a full face mask as well as with pillows

I have another Air Sense 10 and the smart start function works fine with both the pillows and the full face mask

My settings are 4 for the starting pressure and 8 for the maximum pressure

PALERIDER - please put me in touch with the person you mentioned who has taken a bunch of these machines apart
When a PAP comes in with this issue I:
1. Check settings
2. Check for water/dust/dropped damage
3. Run a calibration
4. Change main PCB

The Air 10 is pretty tight in design there is only about a 1/4" between the sensors on the PCB and ports. If they didn't have a leak before, the silicone seals weren't messed with, or the PAP wasn't dropped; I would give it a 99.99% change a leak wouldn't just start on it's own. Take several deep breathes and see if it triggers on. If it does a re-cal is all it may need.
Can you send me a procedure for calibration (I am a Medical Engineer and should be able to follow the procedure accurately)

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed Air Sense 10 Service Manual

Post by Wulfman... » Wed Apr 12, 2017 3:20 pm

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05