Saw a different DR today, his intrpretation of high CA Index

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 64988
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Saw a different DR today, his intrpretation of high CA Index

Post by Pugsy » Sun Feb 08, 2015 2:33 am

archangle wrote:I think sometimes, insurance requires you to try bilevel before getting an ASV.
Arch...he's already using bilevel...he's got the PR S1 BiPap Auto.
I wonder if the doc thinks the AirCurve 10 VAuto is the ASV model....it isn't.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
cwillis
Posts: 71
Joined: Sun Dec 14, 2014 10:47 pm
Location: SF Bay Area
Contact:

Re: Saw a different DR today, his intrpretation of high CA Index

Post by cwillis » Sun Feb 08, 2015 12:06 pm

Pugsy wrote:Scratching my head here...the AirCurve VAuto is like the S9 VPAP Auto and the PR S1 Bipap Auto.
Could have got the same settings on the Bipap Auto.
Hi Pugsy so I took your advise and changed my EPAP min last night to 9. Actually had a better night, did notice that the Centrals usually follow a period after I was awakened and used bathroom. So maybe they are a transitional sleep generated apnea? This morning as I was in bed half awake I had numerous. So without those my CAI would have been lower say 3.54. Should I try these new settings they are prescribing on my PRS1?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: SLEEPY Head/ PR System One BiPap Auto w/Bi-Flex

User avatar
Pugsy
Posts: 64988
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Saw a different DR today, his intrpretation of high CA Index

Post by Pugsy » Sun Feb 08, 2015 12:30 pm

cwillis wrote:Should I try these new settings they are prescribing on my PRS1?
Sure. I don't see any reason not to at least try them.

PS setting on the VPAP is fixed and won't vary so if you are using a PS minimum and maximum setting on the BiPap Auto you will need the PS minimum and maximum to equal each other to mimic how the VPAP works.
cwillis wrote:did notice that the Centrals usually follow a period after I was awakened and used bathroom. So maybe they are a transitional sleep generated apnea?
Very possible that they are sleep onset centrals and/or even simply awake.semi awake breathing irregularities getting flagged by mistake if you don't go back to sleep quickly.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Saw a different DR today, his intrpretation of high CA Index

Post by archangle » Sun Feb 08, 2015 12:43 pm

Pugsy wrote:
archangle wrote:I think sometimes, insurance requires you to try bilevel before getting an ASV.
Arch...he's already using bilevel...he's got the PR S1 BiPap Auto.
I wonder if the doc thinks the AirCurve 10 VAuto is the ASV model....it isn't.
Thanks, I noticed that. Maybe there's some sort of perceived insurance process to cause that. Maybe it's just doctor stupidity and doc is too lazy to learn to read PRS1 data instead of ResMed data.

Actually, if the doctor is using ResScan and Encore, that alone is enough to make the doctor insist on ResMed if you think about it.

I think it would be a good idea to ask the doctor if there's actually any reason to use a ResMed vs. the existing PRS1 machine.

I suspect it's just dummy doctor stuff, and they didn't notice OP already has an auto bilevel and just wrote a prescription for the ResMed machine without thinking.

cwillis, you should definitely ask the doc why he wants a ResMed auto bilevel instead of your existing PRS1 auto bilevel machine.

You definitely need to be sure the insurance will pay for the change between models of the same device. You could easily get screwed with the scam "full price" rental rates towards the end of your "rent to own" period if you aren't careful.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
cwillis
Posts: 71
Joined: Sun Dec 14, 2014 10:47 pm
Location: SF Bay Area
Contact:

Re: Saw a different DR today, his intrpretation of high CA Index

Post by cwillis » Sun Feb 08, 2015 3:22 pm

Pugsy wrote:
cwillis wrote:Should I try these new settings they are prescribing on my PRS1?
Sure. I don't see any reason not to at least try them.

PS setting on the VPAP is fixed and won't vary so if you are using a PS minimum and maximum setting on the BiPap Auto you will need the PS minimum and maximum to equal each other to mimic how the VPAP works.

Not sure what you mean pressure is fixed? That means my pressure will always be IPAP 17 and EPAP 12? Or is there a min that my machine will start at? If PS is 4?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: SLEEPY Head/ PR System One BiPap Auto w/Bi-Flex

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Saw a different DR today, his intrpretation of high CA Index

Post by palerider » Sun Feb 08, 2015 3:30 pm

cwillis wrote:Not sure what you mean pressure is fixed? That means my pressure will always be IPAP 17 and EPAP 12? Or is there a min that my machine will start at? If PS is 4?
pressure SUPPORT (ps) is fixed on the vpap line... the respironics have a minps and maxps setting, the resmeds simply have a ps setting.

within limits, you can duplicate the settings on your respironics by setting minps and maxps to 4.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
cwillis
Posts: 71
Joined: Sun Dec 14, 2014 10:47 pm
Location: SF Bay Area
Contact:

Re: Saw a different DR today, his intrpretation of high CA Index

Post by cwillis » Sun Feb 08, 2015 6:22 pm

palerider wrote:
cwillis wrote:Not sure what you mean pressure is fixed? That means my pressure will always be IPAP 17 and EPAP 12? Or is there a min that my machine will start at? If PS is 4?
pressure SUPPORT (ps) is fixed on the vpap line... the respironics have a minps and maxps setting, the resmeds simply have a ps setting.

within limits, you can duplicate the settings on your respironics by setting minps and maxps to 4.
Thank you Pale Rider so does that mean this new machine will start at IPAP 13 and EPAP 8?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: SLEEPY Head/ PR System One BiPap Auto w/Bi-Flex

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Saw a different DR today, his intrpretation of high CA Index

Post by palerider » Sun Feb 08, 2015 6:39 pm

cwillis wrote:
palerider wrote:
cwillis wrote:Not sure what you mean pressure is fixed? That means my pressure will always be IPAP 17 and EPAP 12? Or is there a min that my machine will start at? If PS is 4?
pressure SUPPORT (ps) is fixed on the vpap line... the respironics have a minps and maxps setting, the resmeds simply have a ps setting.

within limits, you can duplicate the settings on your respironics by setting minps and maxps to 4.
Thank you Pale Rider so does that mean this new machine will start at IPAP 13 and EPAP 8?
the settings you mentioned earlier were minepap 12, maxipap 17, ps 4.

that means the pressures would vary between 16/12 and 17/13 at that .... strange sounding setting.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
cwillis
Posts: 71
Joined: Sun Dec 14, 2014 10:47 pm
Location: SF Bay Area
Contact:

Re: Saw a different DR today, his intrpretation of high CA Index

Post by cwillis » Sun Feb 08, 2015 9:44 pm

palerider wrote:
cwillis wrote:
palerider wrote:
cwillis wrote:Not sure what you mean pressure is fixed? That means my pressure will always be IPAP 17 and EPAP 12? Or is there a min that my machine will start at? If PS is 4?
pressure SUPPORT (ps) is fixed on the vpap line... the respironics have a minps and maxps setting, the resmeds simply have a ps setting.

within limits, you can duplicate the settings on your respironics by setting minps and maxps to 4.
Thank you Pale Rider so does that mean this new machine will start at IPAP 13 and EPAP 8?
the settings you mentioned earlier were minepap 12, maxipap 17, ps 4.

that means the pressures would vary between 16/12 and 17/13 at that .... strange sounding setting.
Pale Rider how would you suggest I start changing my pressures in my current machine to get used to these pressure prescribed, go up on all settings by one?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: SLEEPY Head/ PR System One BiPap Auto w/Bi-Flex

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Saw a different DR today, his intrpretation of high CA Index

Post by palerider » Sun Feb 08, 2015 9:47 pm

cwillis wrote:
palerider wrote: the settings you mentioned earlier were minepap 12, maxipap 17, ps 4.

that means the pressures would vary between 16/12 and 17/13 at that .... strange sounding setting.
Pale Rider how would you suggest I start changing my pressures in my current machine to get used to these pressure prescribed, go up on all settings by one?
what do you have now ,and where do you want to get to?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
cwillis
Posts: 71
Joined: Sun Dec 14, 2014 10:47 pm
Location: SF Bay Area
Contact:

Re: Saw a different DR today, his intrpretation of high CA Index

Post by cwillis » Sun Feb 08, 2015 10:00 pm

I'm at Min EPAP 9 Max 12 Min IPAP 11 Max 14 PS min 2 PS max 4. I guess I want to see how I deal with pressure increase and how it effects my Centrals.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: SLEEPY Head/ PR System One BiPap Auto w/Bi-Flex

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Saw a different DR today, his intrpretation of high CA Index

Post by palerider » Sun Feb 08, 2015 10:18 pm

cwillis wrote:I'm at Min EPAP 9 Max 12 Min IPAP 11 Max 14 PS min 2 PS max 4. I guess I want to see how I deal with pressure increase and how it effects my Centrals.
I'm not trying to advise anything, I'm simply clarifying your confusion about pugsys statement
PS setting on the VPAP is fixed and won't vary so if you are using a PS minimum and maximum setting on the BiPap Auto you will need the PS minimum and maximum to equal each other to mimic how the VPAP works.
I suggest you continue doing what she suggests.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.