what are the indications for a bipap (instead of an apap)?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Jeannh
Posts: 296
Joined: Sun Aug 18, 2013 12:08 pm
Location: CA

what are the indications for a bipap (instead of an apap)?

Post by Jeannh » Sat Nov 09, 2013 3:58 pm

What are the indications for a bipap (instead of an apap)?

thanks

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: EPAP 8, IPAP 14, PS 3.6

User avatar
bwexler
Posts: 1565
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: what are the indications for a bipap (instead of an apap)?

Post by bwexler » Sat Nov 09, 2013 4:25 pm

Not an expert but having just changed, it seems uncontrolled central apneas seem to be one indication.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
CarpeNoctum
Posts: 116
Joined: Tue Nov 01, 2011 7:40 pm
Location: Ilwaco WA

Re: what are the indications for a bipap (instead of an apap)?

Post by CarpeNoctum » Sat Nov 09, 2013 4:46 pm

bwexler wrote:Not an expert but having just changed, it seems uncontrolled central apneas seem to be one indication.
Nope, regular BiPap doesn't treat centrals. BiPap ST or ASV type machines will treat centrals.

If you're cpap pressures are so high that it's problematic to breathe against...BiPap will help with that.
CN
Last edited by CarpeNoctum on Sat Nov 09, 2013 4:50 pm, edited 1 time in total.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: what are the indications for a bipap (instead of an apap)?

Post by JDS74 » Sat Nov 09, 2013 4:48 pm

Jeannh wrote:What are the indications for a bipap (instead of an apap)?
thanks
Do you mean a straight bi-level machine rather than the auto-titrating one like you have?

A straight bi-level is used to treat plain obstructive apnea where the patient has difficulty tolerating the single pressure needed to treat the OSA's so the bi-level gives controlled pressure relief.

An APAP, autopap, is a variable pressure machine that can react to the patient's needs. Some APAP's are also bi-level machines that give the best of both regimens.

Neither deals with central apneas. For that you need a ventilator type machine, one of which is the one bwexler has and refers to.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 64943
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: what are the indications for a bipap (instead of an apap)?

Post by Pugsy » Sat Nov 09, 2013 4:58 pm

BiPaps for those of use with no history of central apnea problems and not seeing a truckload of them on our reports...indications might be comfort with inhale and exhale as it is easier especially at higher pressures or maybe someone is having a lot of aerophagia problems and have the difference between inhale and exhale can offer substantial relief from aerophagia issues.
These are what I call "plain" bipaps.

Then there are the BiPaps that are used to treat obstructive and/or central apnea or complex sleep apnea issues where the centrals are numerous enough to present a problem...This would be the ASV type or Timed breath machines. Fancier Bilevel pressure machines.

So there is more than one type of bilevel pressure machine. Not everyone who uses a bilevel (BiPap is Respironcis trade name and VPAP is the ResMed trade name) is treating centrals. I use a "plain" bilevel pressure machine. I don't have central issues. I use it because I sleep better with it. I like the comfort.

So you have to understand what you are wanting to do and which type of bilevel machine you are referring to.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Rick007
Posts: 50
Joined: Sun Oct 20, 2013 4:40 pm
Location: Ontario, Canada

Re: what are the indications for a bipap (instead of an apap)?

Post by Rick007 » Sat Nov 09, 2013 5:14 pm

I am relatively new here, and was just researching the cause and treatment for the high number of flow limitations I see on my Sleepyhead graphs. I came across a web site that seemed to be trustworthy, and the author seemed to be saying that flow limitations are best handled by a BiPAP machine.

I am really confused as to the difference between a BiPAP machine and a machine with the EPR. I looked at the descriptions of the various Resmed models, and it sounds like the purpose of BiPAP is to reduce the exhalation pressure, and yet that is the same purpose of the EPR function in Resmed's other machines like the S9 Autoset. So how is a BiPAP machine different than a machine with EPR.

Rick

User avatar
Pugsy
Posts: 64943
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: what are the indications for a bipap (instead of an apap)?

Post by Pugsy » Sat Nov 09, 2013 5:26 pm

Rick007 wrote:So how is a BiPAP machine different than a machine with EPR.
Not a lot of difference that most people would notice as long as the greatest relief a person might want or need is 3 cm because 3 cm is as high as EPR will go. With a bilevel machine you can have 4 or 5 or 6 or 7 or more cm difference between inhale and exhale.
I did get a chance to test the ResMed VPAP in CPAP mode (12 cm ) and used 3 cm EPR which gave me 9/12 and then I switched over to bipap fixed mode with EPAP at 9 and IPAP at 12 and I could sense a very slight difference in the timing when inhale and exhale would kick in. Using EPR..the machine would end the relief and being the increase whether I was full ready to inhale or not...with the bilevel mode the machine waited for me to being the inhale and then it kicked in.
There is an adjustment to EPR that can be made that might have done the same thing and improved the timing. I didn't think to try it though and I no longer have the S9 VPAP to try it.

As it was though..I actually prefer 4 cm relief. Just something that feels really natural to me...even 5 cm felt better than 3.

On a bilevel machine the difference between EPAP and IPAP is called Pressure support.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Rick007
Posts: 50
Joined: Sun Oct 20, 2013 4:40 pm
Location: Ontario, Canada

Re: what are the indications for a bipap (instead of an apap)?

Post by Rick007 » Sat Nov 09, 2013 5:58 pm

Pugsy wrote:Not a lot of difference that most people would notice as long as the greatest relief a person might want or need is 3 cm because 3 cm is as high as EPR will go. With a bilevel machine you can have 4 or 5 or 6 or 7 or more cm difference between inhale and exhale.
Thanks for that explanation Pugsy. It certainly makes sense. I guess the BiPAP machines have a little more processing power as well as a pump that can increase pressure very quickly.

I plan on doing some more research into my Flow Limitations and will probably ask some questions in a dedicated thread, but early research shows at least one doctor recommends a BiPAP machine for treatment of flow limitations.

Rick

User avatar
CarpeNoctum
Posts: 116
Joined: Tue Nov 01, 2011 7:40 pm
Location: Ilwaco WA

Re: what are the indications for a bipap (instead of an apap)?

Post by CarpeNoctum » Sat Nov 09, 2013 6:28 pm

Rick007 wrote: Thanks for that explanation Pugsy. It certainly makes sense. I guess the BiPAP machines have a little more processing power as well as a pump that can increase pressure very quickly.

I plan on doing some more research into my Flow Limitations and will probably ask some questions in a dedicated thread, but early research shows at least one doctor recommends a BiPAP machine for treatment of flow limitations.
Rick, it's my understanding that "flow limitations" are the precursor of hypops and apneas. They are what an auto machine senses that allows it to increase pressure to stave off the subsequent event.

So really the machine doesn't treat flow limitations per se, but used them in it's efforts to treat other more serious events (hypops and apneas).

An auto cpap is just as capable as a BiPap to prevent obstructive apneas...UNLESS they require very high pressures. In which case a Bilevel machine is indicated. Many bilevels are capable of 25-30 cmH20 (depending on make and model)

I think your research will bear this out.
CN

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: what are the indications for a bipap (instead of an apap)?

Post by chunkyfrog » Sat Nov 09, 2013 6:34 pm

In some cases, aerophagia may be helped by using a bipap.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: what are the indications for a bipap (instead of an apap)?

Post by DoriC » Sat Nov 09, 2013 7:05 pm

We use Bipap for comfort and better numbers, pressure needs are average. It's also my understanding that Pressure Support differs from EPR in that PS actually maintains pressure on exhale while EPR suspends it until you inhale again. I'd like to have a definitive answer to that statement?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

johnthomasmacdonald
Posts: 977
Joined: Sun Sep 02, 2012 1:48 pm
Location: northwest new jersey and Boston Mass

Re: what are the indications for a bipap (instead of an apap)?

Post by johnthomasmacdonald » Sat Nov 09, 2013 7:07 pm

comfort was the main issue for me, i felt like i was suffocating with trying to exhale against such high pressure, i needed relief. Switching to bipap allowed me to be able to use the machine through the night. Without it i would have given up.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Resmed ASV adapt enhanced - epap=8; minPS=3; maxPS=17 ave. pressure =10

purple
Posts: 837
Joined: Sat Feb 28, 2009 9:06 pm

Re: what are the indications for a bipap (instead of an apap)?

Post by purple » Sat Nov 09, 2013 7:27 pm

Curious. Why do you use APAP rather than CPAP. Some use APAP because their correct titration pressures have not yet been determined. I think there are some breathing issues for which APAP works better. In my case, my breathing patterns do not work well with APAP.

I use a Bi Level machine because I have severe Aerophagia, as well as it works better. Severe Aerophagia also means I am not exhaling against pressure, because I am swallowing air.

In my opinion. The lower pressure is what I call the stent pressure, the pressure which keeps my airway open. When I inhale, I suspect, because my airway is so narrow, I pull air faster than the machine provides at stent pressure. So the inhalation pressure is higher than my exhale.

Seems my machine does not start the inhale, until it senses I am already inhaling. Perhaps it is possible for some Bi Level machines to kick on to the inhale without my initiating it. The machine starting the inhale, perhaps at the higher pressure, without me seems to be the area of a Vent, which might be a non-invasive vent. There is another topic, in that, I know when I am awake, I have a long hesitation between inhaling and exhaling. This is what I think makes a machine set to auto to keep increasing pressure when I really do not need it.

A person with some health problems, like MS will get at least a Bi Level machine as their first machine. Eventually I would suppose they will go onto a vent.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: what are the indications for a bipap (instead of an apap)?

Post by BlackSpinner » Sat Nov 09, 2013 8:50 pm

Curious. Why do you use APAP rather than CPAP. Some use APAP because their correct titration pressures have not yet been determined. I think there are some breathing issues for which APAP works better. In my case, my breathing patterns do not work well with APAP.
Some people need different pressures during REM.
I have mine set 8.5 - 12. It doesn't often get that high but it is nice to know it can adjust. My cpap pressure was 9 but when I got the apap I had it open (7-20) it was often hanging around 10 so I adjusted it.

I really didn't notice the difference once I got used to the different sound of the machine.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
Jeannh
Posts: 296
Joined: Sun Aug 18, 2013 12:08 pm
Location: CA

Re: what are the indications for a bipap (instead of an apap)?

Post by Jeannh » Sat Nov 09, 2013 9:41 pm

wow- thank you all for so much info! This forum always amazes me.
Curious. Why do you use APAP rather than CPAP.
i wanted the flexibility, and my doctor was amenable. I discovered that I flex between 10-12, so set it to 9-14. I've had a lot of dry mouth problems, even with taping that is still secure in the morning, and what I thought was aerophagia. After some experimentation, I think the bloating and gas was from sorbitol (in sugarless gum), so now I'm back to dry mouth, which I am hoping to solve by lowering a medication, but it's been tough. Sleep doc asked me to lower the high limit to 12, so I'm on my second night of that.

Just trying to sleep through the night. Jean

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: EPAP 8, IPAP 14, PS 3.6