I am now trialing a BiPAP Auto M and can't figure out why it switches from IPAP to EPAP before I'm ready. I inhale and before I am done it switches to exhale. Is it a matter of changing a bi-flex setting or turning bi-flex off and playing with the Rise Comfort setting?
Re: Question for BiPAP Auto M Experts
I'm wondering if you have any medical issues not mentioned? Have you tried calling your DME?lazyace wrote:I am now trialing a BiPAP Auto M and can't figure out why it switches from IPAP to EPAP before I'm ready. I inhale and before I am done it switches to exhale. Is it a matter of changing a bi-flex setting or turning bi-flex off and playing with the Rise Comfort setting?
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: Question for BiPAP Auto M Experts
Hi,
Is is possible that you are taking very deep breaths. If so, it could be that you are changing the rate of your inhalation/exhalation and triggering the change. I have noticed that this can happen to me if I take unusually large breaths (for me).
Is is possible that you are taking very deep breaths. If so, it could be that you are changing the rate of your inhalation/exhalation and triggering the change. I have noticed that this can happen to me if I take unusually large breaths (for me).
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: Encore Pro, Backup-M Series BiPAP Auto. Set as ABiPAP 15I/9E AHI 0.2. Normally operates @ 12I/9E 1987 OSA diagnosis |
If your never go off on a tangent, you are doomed to live your life going around in circles.
Re: Question for BiPAP Auto M Experts
No medical issues that would impact the settings.GumbyCT wrote:I'm wondering if you have any medical issues not mentioned? Have you tried calling your DME?lazyace wrote:I am now trialing a BiPAP Auto M and can't figure out why it switches from IPAP to EPAP before I'm ready. I inhale and before I am done it switches to exhale. Is it a matter of changing a bi-flex setting or turning bi-flex off and playing with the Rise Comfort setting?
Could be somewhat deep breaths, but I don't think it is abnormal.builta wrote:Hi,
Is is possible that you are taking very deep breaths. If so, it could be that you are changing the rate of your inhalation/exhalation and triggering the change. I have noticed that this can happen to me if I take unusually large breaths (for me).
Re: Question for BiPAP Auto M Experts
Hi,lazyace wrote:I am now trialing a BiPAP Auto M and can't figure out why it switches from IPAP to EPAP before I'm ready. I inhale and before I am done it switches to exhale. Is it a matter of changing a bi-flex setting or turning bi-flex off and playing with the Rise Comfort setting?
I also have the M series BiPAP Auto. I notice that it sometimes cuts me off when I inhale. I have a bit of a tendency to breathe shallow and show, then try to catch up with a deep breath. The machine has made me more aware of my breathing, and I have been learning to breathe at a little more regular rate.
I also notice that I hold my breath from time to time, especially when falling asleep. The machine will detect this and try to force me to breathe. For example, if I hold my breath a bit, then take a deep breath to catch up, the machine will automatically go to high pressure by itself 5 times in a row to try to get me to breathe.
-john-
- momof2inmn
- Posts: 47
- Joined: Mon Sep 07, 2009 8:11 pm
- Location: Minnesota, USA
Re: Question for BiPAP Auto M Experts
Yep, deep breaths is what does it for me, too!
- rested gal
- Posts: 12881
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: Question for BiPAP Auto M Experts
lazyace, the BiPAP (including the BiPAP Auto) will blow the higher IPAP pressure for about 3 seconds. If you've not started to breathe out within that 3 second time period the machine will automatically switch down to the lower EPAP. That can feel like the machine is cutting off your inhalation.
The BiPAP is supposed to stay down at the lower exhale pressure until you start to inhale again. But the BiPAP is not supposed to stay up at the IPAP pressure for longer than about 3 seconds, even if you are still inhaling.
I wish they'd make 4 seconds instead of 3 seconds be the limit for the higher IPAP to last, but... it's 3 seconds. Apparently most people normally inhale for about 1 to 2 seconds.
The BiPAP (including the BiPAP Auto) will stay at the lower EPAP pressure until you start to breathe in again. Inhaling triggers the machine to switch up to the higher IPAP pressure.
Holding your breath, or the normal pause before inhaling again, makes the machine stay at the lower EPAP pressure until you actually start to inhale again.
The slightest indication of breathing in or breathing out triggers the machine one way or another. That "slight indication" might even be from aerophagia (air getting accidentally pushed down past the esophageal sphincter) which could make the machine think there's been a change in airflow from you. A sudden little leak could also make the machine think there's been a change in your airflow.
In the following thread I wrote about "4 seconds" being the limit for IPAP to blow, but it's 3 seconds, I believe:
viewtopic.php?p=233928#p233928
The BiPAP is supposed to stay down at the lower exhale pressure until you start to inhale again. But the BiPAP is not supposed to stay up at the IPAP pressure for longer than about 3 seconds, even if you are still inhaling.
I wish they'd make 4 seconds instead of 3 seconds be the limit for the higher IPAP to last, but... it's 3 seconds. Apparently most people normally inhale for about 1 to 2 seconds.
The BiPAP (including the BiPAP Auto) will stay at the lower EPAP pressure until you start to breathe in again. Inhaling triggers the machine to switch up to the higher IPAP pressure.
Holding your breath, or the normal pause before inhaling again, makes the machine stay at the lower EPAP pressure until you actually start to inhale again.
The slightest indication of breathing in or breathing out triggers the machine one way or another. That "slight indication" might even be from aerophagia (air getting accidentally pushed down past the esophageal sphincter) which could make the machine think there's been a change in airflow from you. A sudden little leak could also make the machine think there's been a change in your airflow.
In the following thread I wrote about "4 seconds" being the limit for IPAP to blow, but it's 3 seconds, I believe:
viewtopic.php?p=233928#p233928
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: Question for BiPAP Auto M Experts
Same thing happened to me. Change the rise setting on your machine. Mine was set at 2, I upped it to 6 and I sync much better with the machine now
Re: Question for BiPAP Auto M Experts
Oops, forgot. I also extended my "ramp" time from 15 to 30 minutes. It allowed me to fall asleep and stay asleep longer so when the machine reached full pressure, I didn't wake up.
- rested gal
- Posts: 12881
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: Question for BiPAP Auto M Experts
Are you using a ResMed machine, ctmerchant?ctmerchant wrote:Same thing happened to me. Change the rise setting on your machine. Mine was set at 2, I upped it to 6 and I sync much better with the machine now
I think lazyace is using the Respironics BiPAP Auto M series machine, which does not have a "rise" setting of "6", as far as I know.
From the Provider manual of the BiPAP Auto M series:

I don't think the rise setting -- called "comfort rise" in the Respironics BiPAP Auto M series -- has the adjustments in milliseconds (200 ms - 900 ms) that ResMed VPAP (and VPAP Auto) machines have.
From page 8 of the VPAP Auto provider manual:
Rise Time Adjustment
(S mode) Rise Time is the time taken for the pressure to increase from EPAP up
to IPAP. Rise Time can be set to MIN (the fastest Rise Time) and then in 50-unit
increments from 200 to 900. This will control the rate of pressure increase when
the VPAP Auto switches from EPAP to IPAP.
At any rate, "rise" settings (the rate of speed the machine takes to reach the full IPAP pressure once an inhalation has started) are not the same as length of time the IPAP pressure can last during an inhalation. Length of time the IPAP pressure can last is what lazyace was asking about, I think.
But perhaps prolonging the amount of time the machine takes to reach full IPAP pressure might get added onto the built-in 3 seconds (or whatever amount of time) duration IPAP lasts? If so (and I don't know that it would be so) that could give a leeeetle more time for inhalations that last longer than usual. Like when taking an unusually slow, deep breath.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: Question for BiPAP Auto M Experts
Thank you for the information I forgot how to change settings on my M-series. (I know read the owners manual) that's too easy.rested gal wrote:Are you using a ResMed machine, ctmerchant?ctmerchant wrote:Same thing happened to me. Change the rise setting on your machine. Mine was set at 2, I upped it to 6 and I sync much better with the machine now
I think lazyace is using the Respironics BiPAP Auto M series machine, which does not have a "rise" setting of "6", as far as I know.
From the Provider manual of the BiPAP Auto M series:
I don't think the rise setting -- called "comfort rise" in the Respironics BiPAP Auto M series -- has the adjustments in milliseconds (200 ms - 900 ms) that ResMed VPAP (and VPAP Auto) machines have.
From page 8 of the VPAP Auto provider manual:
Rise Time Adjustment
(S mode) Rise Time is the time taken for the pressure to increase from EPAP up
to IPAP. Rise Time can be set to MIN (the fastest Rise Time) and then in 50-unit
increments from 200 to 900. This will control the rate of pressure increase when
the VPAP Auto switches from EPAP to IPAP.
At any rate, "rise" settings (the rate of speed the machine takes to reach the full IPAP pressure once an inhalation has started) are not the same as length of time the IPAP pressure can last during an inhalation. Length of time the IPAP pressure can last is what lazyace was asking about, I think.
But perhaps prolonging the amount of time the machine takes to reach full IPAP pressure might get added onto the built-in 3 seconds (or whatever amount of time) duration IPAP lasts? If so (and I don't know that it would be so) that could give a leeeetle more time for inhalations that last longer than usual. Like when taking an unusually slow, deep breath.
Oh here's lookin' at you kid.
Don't Bend or Squash, My Aluminum Hat,it keeps them from knowing what I am thinking!
I need more Coffee&Old Bushmills!
"Without Truckdrivers America Stops!"
I'm not always wrong,but I'm not always right!
"Semper Fi"
I need more Coffee&Old Bushmills!
"Without Truckdrivers America Stops!"
I'm not always wrong,but I'm not always right!
"Semper Fi"
Re: Question for BiPAP Auto M Experts
RG,rested gal wrote:lazyace, the BiPAP (including the BiPAP Auto) will blow the higher IPAP pressure for about 3 seconds. If you've not started to breathe out within that 3 second time period the machine will automatically switch down to the lower EPAP. That can feel like the machine is cutting off your inhalation.
The BiPAP is supposed to stay down at the lower exhale pressure until you start to inhale again. But the BiPAP is not supposed to stay up at the IPAP pressure for longer than about 3 seconds, even if you are still inhaling.
I wish they'd make 4 seconds instead of 3 seconds be the limit for the higher IPAP to last, but... it's 3 seconds. Apparently most people normally inhale for about 1 to 2 seconds.
The BiPAP (including the BiPAP Auto) will stay at the lower EPAP pressure until you start to breathe in again. Inhaling triggers the machine to switch up to the higher IPAP pressure.
Holding your breath, or the normal pause before inhaling again, makes the machine stay at the lower EPAP pressure until you actually start to inhale again.
The slightest indication of breathing in or breathing out triggers the machine one way or another. That "slight indication" might even be from aerophagia (air getting accidentally pushed down past the esophageal sphincter) which could make the machine think there's been a change in airflow from you. A sudden little leak could also make the machine think there's been a change in your airflow.
In the following thread I wrote about "4 seconds" being the limit for IPAP to blow, but it's 3 seconds, I believe:
viewtopic.php?p=233928#p233928
I have come to the conclusion in just two nights that unless there is a way to lengthen the time between inhale and exhale, I simply cannot tolerate this machine. I feel like I am almost hyperventilating because I can't keep pace with the cycles. Do you know whether there is another machine that offers the user the ability to lengthen that span? I like the BiPAP better than the APAP for the aerophagia. But now, the breathing is more of a nuisance than the aerophagia was. What is the difference between the Respironics BiPAP and a VPAP? Does ResMed make a comparable product? If so, what do you know about it?
Thanks!
Re: Question for BiPAP Auto M Experts
Do you know what the settings currently are?
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: Question for BiPAP Auto M Experts
I've got biflex turned off because when I played with those settings it didn't seem to have much impact on the spread. I have the rise set on 3. IPAP is set to 7 and EPAP is 5.GumbyCT wrote:Do you know what the settings currently are?