I think that is all you can conclude from results so far. I think when you set BPM=Off there is no backup mode or it is disabled. Since she has periods of central dysregulation is that a good thing to do?dsm wrote: Good starting point.
1) One value we need is Bev's typical BPM without there being any machine influencing factors such as PS or backup rate or auto-adjustment rate.
We could get that by setting BPM=off
BPM can be easily gotten off the LCD in Monitor mode, it will give you the average of last 6 breaths, I suspect a BPM avg is still there. Same for Rise time and Inspiration time. BPM on her first night on the SV avg'd 14.6, if protocol is followed you use spontaneous -2 (i.e. 14.6-2=12.6). Inspiration time ranges from .5 to 3 seconds. Machine will never let IT get over 3 seconds or it will cycle.
Actually last night was set to CPAP=14 for EPAP. Compared to Aflex, it was the same starting pressure but limited to 18 cm Max. That bumping into IPAP Max is what resulted in HI being seen. Since we started IPAP Min also at 14 cm, it was CPAP +SV mode.dsm wrote: 2) Set the machine in CPAP mode as already suggested and stick with the CMs set to 14 based on it delivering consistenly better results - then we
are looking at Bev's baseline reaction using the SV's better data reporting capability
She doesn't appear to do well with PS either that she cannot tolerate the rapid movement of IPAP working pressure. Wonder how she would do with 18/14 fixed.dsm wrote: 3) Don't use PS yet until we have a night or two on this same machine, in CPAP mode & no timed mode
Bev indicated she is 5' 7" and from what I can tell ideal weight since her PSG shows a BMI=20, it don't get no better than that.dsm wrote:
Add to this what we think we should be seeing in the data for Bev based on what we know of her weight height etc:
a) Peak Average Volume
b) Tidal Flow
c) breathing rate
Agree on wht meds Bev should be on
DSM
AVAPS Chart guide says:
Since there is NO target tidal volume on this machine, we have to compare where she should be for her given height to the Encore avg.
Set the target tidal volume, either to 110% of the displayed patient
tidal volume when ventilated on S/T mode or to 8 ml/kg of ideal weight.
Adjust depending on patient tolerance and clinical outcomes.
Target Vte may be set from 200 ml to 1500 ml.
If I look up her height on the chart of 67" she should be at 530ml tidal volume, at 395 she is a bit low yet. But that is under SV control, not backup settings. Again, I think the IPAP Max limited the machine's response in correcting that, but it doesn't have anywhere you can input height to know what the target volume should be. It is my understanding it only targets peak inspiration volume so to get tidal up you have to get Peak up.
Ramp: If she is to continue with the SV, I think setting up a Ramp for her should help make it easier for her to fall asleep. Ramp only impacts EPAP setting, any IPAP Min will remain as IPAP pressure until Ramp expires.
Giving her a Ramp from 9 to 14 over 30 minutes might help her a lot in falling asleep.