Snoredog wrote:Do you have IFL1 disabled in the above graphs?
It looks like it is enabled to me based upon the pressure trends shown.
IFL1 is truly OFF. The machine display says 0, and the software settings display for the past 96 hours of data is as shown below:
Now, IFL2 is ON, but not IFL1. According to the SilverLining manual:
IFL1 - Command on Flow Limitation Run
IFL2 - Command on Flow Limitation Run combined with an amplitude decrease
I really wasn't even aware of IFL2 until recently. (Doesn't mean I didn't inadvertently change it earlier though. )
Snoredog wrote:Careful reading too much into the older 418"A" manual, it is a totally different animal to the 418"P".
The 418"P" and the 420E are functionally identical but the 418A is not.
Nevertheless, the 420E manual has the same exact definitions which ozij quoted. Additionally, the 420E manual states:
GoodKnight 420E Manual wrote:The GoodKnight Evolution uses the same event detection technology as does the GoodKnight 418A.
Snoredog wrote:In Bill's case it appears he has IFL1 checked (based upon my view of the pressure response seen). That is why FL runs correspond with pressure increases on his graph.
LOL. Well, despite appearances IFL1 is unchecked and definitely OFF. Hmm, the mystery deepens ... Here's last night's data (along with the night before) which seems different mainly in that the machine went up to maximum pressure and stayed there for a while. There was a short period without flow-limitation "Runs" as well.
ozij wrote:Bill,
Did you ever compare your flow limitation runs no. (or index) with what you got on the Respironics, when you used both machines at about the same time of the year?
I have always found it amazing that people like Rested Gal and myself have about 30% flow limited breath per night. And that doesn't seem to change much, no matter what the pressure.
Well, the data is so different I never did a direct comparison before, Ozij. I do, however, have data from one year ago from the Remstar:
Somehow, the machine came up with an FLI of 1.1 above. In some of the other charts I looked at briefly, the FLI was simply 0. Obviously, there's a difference between how Respironics and PB calculate flow limitation. It also appears that my own breathing pattern doesn't seem to fit either one particularly well.
Regards,
Bill