What ASV adjustments affect hypopneas most?
- The Sheikh
- Posts: 165
- Joined: Sun Aug 19, 2012 12:22 pm
What ASV adjustments affect hypopneas most?
Hi Folks -
I'm looking for some suggestions....
I've been successful getting my central and obstructive apnea index down to near zero. However the hypopneas seem difficult to get a handle on. My PR Advanced SV ASV unit shows about 15-25 hypopnea events each night. This is not bad, but I'd like to tweak it in better if possible. Evidently my RESIDUAL Cheyne-Stokes breathing pattern may be causing it, I dunno.
At this point here are my settings - after months of trial and error, they are optimized so that CA, OSA, FL, VS and hypopneas, taken as a group, are at a minimum:
Pressure Max: 20
Epap min: 4
Epap max: 6
PS min: 4 (breathing pulse)
PS max 15
Breathing rate: Auto - averages about 13.8 bpm
Rise time: 2
I am wondering WHAT adjustment(s) may help to decrease the hypopnea events? I have not played with the backup breathing rate yet. I have had it set to "Auto."
I tried increasing the PS min breathing amplitude from 3,4 or 5 attempting to better stablize the hypopneas, but 3 or 4 seems to be best. If I go too high, the CA's start in and hypopneas get more frequent.
What effect will the breathing backup rate have on hypopneas, if any? I average about 13.8 for the last month set at auto. I wonder if running it on manual at a slower breathing rate might help?
Anyone have suggestions using these adjustments or any other method to decrease hypopneas?
BTW, I am sleeping well and my O2 remains above 93% all night, so this is more an exercise in just trying to get the hypopnea numbers to read better - more in line with the other indices.
Thanks.
Tom
I'm looking for some suggestions....
I've been successful getting my central and obstructive apnea index down to near zero. However the hypopneas seem difficult to get a handle on. My PR Advanced SV ASV unit shows about 15-25 hypopnea events each night. This is not bad, but I'd like to tweak it in better if possible. Evidently my RESIDUAL Cheyne-Stokes breathing pattern may be causing it, I dunno.
At this point here are my settings - after months of trial and error, they are optimized so that CA, OSA, FL, VS and hypopneas, taken as a group, are at a minimum:
Pressure Max: 20
Epap min: 4
Epap max: 6
PS min: 4 (breathing pulse)
PS max 15
Breathing rate: Auto - averages about 13.8 bpm
Rise time: 2
I am wondering WHAT adjustment(s) may help to decrease the hypopnea events? I have not played with the backup breathing rate yet. I have had it set to "Auto."
I tried increasing the PS min breathing amplitude from 3,4 or 5 attempting to better stablize the hypopneas, but 3 or 4 seems to be best. If I go too high, the CA's start in and hypopneas get more frequent.
What effect will the breathing backup rate have on hypopneas, if any? I average about 13.8 for the last month set at auto. I wonder if running it on manual at a slower breathing rate might help?
Anyone have suggestions using these adjustments or any other method to decrease hypopneas?
BTW, I am sleeping well and my O2 remains above 93% all night, so this is more an exercise in just trying to get the hypopnea numbers to read better - more in line with the other indices.
Thanks.
Tom
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: Apex Wizard 310 Nasal CPAP Mask |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: ResMed Adapt SV (ASV), PR AutoSV Advanced ASV, with SleepyHead, CMS-55H Oximeter and ZEO sleep monitor |
- JohnBFisher
- Posts: 3821
- Joined: Wed Oct 14, 2009 6:33 am
Re: What ASV adjustments affect hypopneas most?
EPAP min and PS Min makes the most difference with Hypopneas.
I would recommend against switching from AUTO for the Breaths Per Minute (BPM). You can seriously mess up your breathing (your CO2 and oxygen exchange) by getting the values wrong. That is especially true if you then adjust the rise time. Most healthy folks will just end up with a horrible night of sleep. But if you have central apneas you will wake up with one humdinger of a headache. We've had some folks try it here before and found "AUTO" was the best choice.
Hope that helps.
I would recommend against switching from AUTO for the Breaths Per Minute (BPM). You can seriously mess up your breathing (your CO2 and oxygen exchange) by getting the values wrong. That is especially true if you then adjust the rise time. Most healthy folks will just end up with a horrible night of sleep. But if you have central apneas you will wake up with one humdinger of a headache. We've had some folks try it here before and found "AUTO" was the best choice.
Hope that helps.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O |
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
- The Sheikh
- Posts: 165
- Joined: Sun Aug 19, 2012 12:22 pm
Re: What ASV adjustments affect hypopneas most?
Hi John,JohnBFisher wrote:EPAP min and PS Min makes the most difference with Hypopneas.
Thanks for the info.
Could you give me an idea of how higher or lower Epap min and PS min might affect the hypopneas?
From trial and error it appears that higher epap mins cause more hypopneas - but Epap min needs to be at a certain level to handle the OSA and flow restrictions.
Though, I am not clear on how the PS min breathing pulses affect everything. So far I find a 3-4 cm/h20 level is the highest I can run before the overall indices start to deteriorate. I would think a stronger PS min pulse (above 3-4) would help stabilize the breathing amplitude, but it seems to make the hypopnea index worse at the same time.
BTW, does the rise time setting (1-5) work when in BPM auto mode or just when in manual mode?
Tom
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: Apex Wizard 310 Nasal CPAP Mask |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: ResMed Adapt SV (ASV), PR AutoSV Advanced ASV, with SleepyHead, CMS-55H Oximeter and ZEO sleep monitor |
-
StillAnotherGuess
- Posts: 132
- Joined: Sat Apr 23, 2011 4:26 pm
Re: What ASV adjustments affect hypopneas most?
Given your current EPAP settings your best chance to reduce hypopneas is to incrementally increase PS Min. And disable Rise Time. At EPAP Min 4, Rise Time is useless.The Sheikh wrote: I am wondering WHAT adjustment(s) may help to decrease the hypopnea events?
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21 |
- JohnBFisher
- Posts: 3821
- Joined: Wed Oct 14, 2009 6:33 am
Re: What ASV adjustments affect hypopneas most?
Well, you've already found that increasing your EPAP does not help. So, I would go with an incremental increase of PS min to bump up the inhalation pressure. This should help keep your airway open a little bit better. But as you've already noted, you may have a very limited room for change.The Sheikh wrote:... Could you give me an idea of how higher or lower Epap min and PS min might affect the hypopneas? ...
Unfortunately, we all have to tinker with the settings (make one change .. measure the results for a week or so .. make another change .. repeat measurements .. etc ...). This is a clear case where one size does not fit all. In fact, with ASV therapy there is no one right answer. The answer is "whatever works best for you!".
I've used this approach several times and clearly end up with a result that can be easily defended when I discuss it with my doctor. He has no problems with this approach to determining what works best for me.
The Rise Time is only available if you do not have BPM set to AUTO. As I noted, changing that can make the unit more difficult to use and possibly result in very poor sleep. I would just leave your BPM setting in AUTO.
Hope that helps.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O |
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
-
johnthomasmacdonald
- Posts: 977
- Joined: Sun Sep 02, 2012 1:48 pm
- Location: northwest new jersey and Boston Mass
Re: What ASV adjustments affect hypopneas most?
how does asv therapy differ from cpap/bipap? - Both machines are shooting air aren't they? what's the fundamental difference?
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: Resmed ASV adapt enhanced - epap=8; minPS=3; maxPS=17 ave. pressure =10 |
Re: What ASV adjustments affect hypopneas most?
The difference is ASV adjusts the pressures on each breath to maintain volumes and rates. BiPAP and auto-bipap will adjust pressures too, but they do it in response to events.johnthomasmacdonald wrote:how does asv therapy differ from cpap/bipap? - Both machines are shooting air aren't they? what's the fundamental difference?
The ASV is a much more advanced machine
Also, an ASV will trigger a breath for you if you don't breathe, at some automatic rate or numerical rate, as a backup. In this, it functions more like a BiPAP/ST
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP |
Re: What ASV adjustments affect hypopneas most?
I have similar issues. I finally have my average AHI below 5 by going to the BiPap ASV. My central apneas and obstructed apneas are now very low, near 0 most nights, but Hypopneas have gone up rendering my AHI of 3 - 5. I found in some Philips literature that said to have pressure support at 4 or above if hypopneas become dominant. I have tried many configurations, but I am prone to leakage at higher pressures and then central apneas again go up. It definitely takes trial and error. I am currently at PS of 5 and slightly lowered EPAP min to 8. Hypopneas have decreased but I now have a few more central apneas. AHI still 3-5. May be as good as it gets for me. I agree to leave BPM to auto. I tried figuring my natural bpm multiple times and different rise times were tried also. Always rendered poor sleep. I believe my bpm may change as I go into deep sleep or REM, so auto is the way to go. I will continue to slightly adjust epap min and ps. Maybe in 0.5 increments instead of whole units as I have. Then I must be patient and try the new settings at least a week before readjusting again.
Re: What ASV adjustments affect hypopneas most?
it's better to start your own thread than to tag onto an older one. you'll likely get some response that way.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg