AirSense 10 units won't run on an inverter? Screwed again?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by archangle » Tue Mar 17, 2015 5:45 am

wm_hess wrote:From reading that, it appears that an inverter can be used to run the device?
Let's assume WebSiteCompany (WSC) is honest and knowledgeable. Assume they say on their web site that the A10 works OK on an inverter.

Is WSC going to replace your A10 machine if it is damaged while using an inverter?

If your A10 machine quits working while being used with an inverter, is WSC going to decide they steered you wrong and replace your A10? How do you know the A10 died from the inverter, and not for unrelated reasons?

How do they know it was damaged because of a general problem with using inverters, or could the problem be that the inverter you used was defective?

Are they going to replace your A10 machine if you used an inverter because of their web site, but didn't buy their battery?

How long is it going to take for them to decide that they were at fault, even if they do? Are they going to want you to send them your A10 machine and the inverter so they can test them and evaluate them?

If you use an inverter and your UPS fails for unrelated reasons, will ResMed deny your warranty claim?
wm_hess wrote:If I may ask though why would you want to use an inverter instead of a converter? Doesn't a inverter use more energy than a converter, thus draining a battery faster?
The ResMed converter is expensive. Inverters are easily obtained elsewhere, and can be used for devices other than CPAP machines. Many people already have inverters.

UPS's can power your CPAP full time and automatically fall back to battery usage when the power goes out. UPS's use inverters. If the A10 won't work on inverters, it won't work on a UPS, either. Most home generators use inverters.

ResMed owes it to their customers to tell them what forms of power will and won't safely power their CPAP machines. Especially since they do give out SOME information about running off of battery power.

If an inverter WILL damage the A10, they owe it to their customers to inform them of this. I consider it to be a design flaw.

By the way, Philips Respironics doesn't publish any information about running on battery. I'd be happy to be proved wrong on this, but I haven't been able to find it.

As I said, I expect it's OK, but it's possible that it can damage the machine. ResMed should update their battery power information page to say yes or no for the A10 machines.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by chunkyfrog » Tue Mar 17, 2015 7:36 pm

I figured it made more sense to buy the Devillbiss for when power supply is a problem.
Then, I not only do not risk my good machine, but I have a spare in case something else happens.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by TangledHose » Tue Mar 17, 2015 8:10 pm

From my little experiment I think it works just fine on a standard modified sine wave inverter. I agree with an earlier poster that's it's probably just a case of the available literature from ResMed has not yet caught up to the release of the new machine.

I think if any component was going to fail over time running on an inverter it would be the power converter on the power cord, which probably has a warranty similar to the machine itself.

Arc.....I notice you are currently listing an S9 in your equipment profile, what percentage of the total run hours have been on an inverter with your current machine? Any problems with your current ResMed machine due to running it from an inverter? And lastly, is this an important consideration for a purchase of a new Airsense machine? If so what is your time frame? Could you delay the purchase of an Airsense until you get definitive information from ResMed? Are you making a Mountain out of a Mole Hill?

Just sayin'

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

User avatar
chartle
Posts: 152
Joined: Mon Mar 12, 2018 7:56 am

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by chartle » Mon Jun 18, 2018 10:30 am

Ok bumping up this thread because of the Airsense 10 and it not being listed in the battery guide for use with an inverter.

I just pulled up the June 2018 ver 6 guide and its still not listed. But I don't see anything warning you NOT to use one. :?:

https://www.resmed.com/us/dam/documents ... lo_eng.pdf

Coming back to this thread because I have a new RV and CPAP and boondocking questions pop up a lot.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by palerider » Mon Jun 18, 2018 11:14 am

chartle wrote:
Mon Jun 18, 2018 10:30 am
Ok bumping up this thread because of the Airsense 10 and it not being listed in the battery guide for use with an inverter.

I just pulled up the June 2018 ver 6 guide and its still not listed. But I don't see anything warning you NOT to use one. :?:

https://www.resmed.com/us/dam/documents ... lo_eng.pdf

Coming back to this thread because I have a new RV and CPAP and boondocking questions pop up a lot.
The air10's are powered the same as the S9.

Using an inverter for them is dumb, in respect to battery life. Use the 12v converter.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by prodigyplace » Mon Jun 18, 2018 12:43 pm

chartle wrote:
Mon Jun 18, 2018 10:30 am
Ok bumping up this thread because of the Airsense 10 and it not being listed in the battery guide for use with an inverter.

I just pulled up the June 2018 ver 6 guide and its still not listed. But I don't see anything warning you NOT to use one. :?:

https://www.resmed.com/us/dam/documents ... lo_eng.pdf

Coming back to this thread because I have a new RV and CPAP and boondocking questions pop up a lot.
The 12 volt converter for the AirSense 10 works and is much more efficient than using an inverter. I use the ResMed converter with a 12 volt battery for power outage situations. I have used it for several nights (without heated hose & humidity in my case) and it works fine.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by chunkyfrog » Mon Jun 18, 2018 2:13 pm

Sixty eight bucks for the converter beats the syrup out of hundreds for the proprietary battery pack.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

D.H.
Posts: 3524
Joined: Sun Sep 04, 2016 7:07 pm

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by D.H. » Mon Jun 18, 2018 3:31 pm

I suggest that if you're starting with a DC power supply, you should not convert it to AC. When you use this machine at home, the "brick" on the power cord converts household electricity (which is AC) to DC. This is what the machine runs on. It makes no sense to convert it and then convert it back!

I suggest you use the product on the link below.


Link to converter ===> https://www.cpap.com/productpage/resmed ... hines.html

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by prodigyplace » Mon Jun 18, 2018 5:08 pm

D.H. wrote:
Mon Jun 18, 2018 3:31 pm
I suggest that if you're starting with a DC power supply, you should not convert it to AC. When you use this machine at home, the "brick" on the power cord converts household electricity (which is AC) to DC. This is what the machine runs on. It makes no sense to convert it and then convert it back!

I suggest you use the product on the link below.


Link to converter ===> https://www.cpap.com/productpage/resmed ... hines.html
What did you do with the REAL D.H. ?? This post is actually reasonably accurate & useful!

The DC converter actually converts from 12 volts DC to 24 volts DC though. There is still some conversion, but very efficient.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by Goofproof » Mon Jun 18, 2018 5:22 pm

prodigyplace wrote:
Mon Jun 18, 2018 5:08 pm
D.H. wrote:
Mon Jun 18, 2018 3:31 pm
I suggest that if you're starting with a DC power supply, you should not convert it to AC. When you use this machine at home, the "brick" on the power cord converts household electricity (which is AC) to DC. This is what the machine runs on. It makes no sense to convert it and then convert it back!

I suggest you use the product on the link below.


Link to converter ===> https://www.cpap.com/productpage/resmed ... hines.html
What did you do with the REAL D.H. ?? This post is actually reasonably accurate & useful!

The DC converter actually converts from 12 volts DC to 24 volts DC though. There is still some conversion, but very efficient.
He may have learned to read, it been posted over 6 times, repetition helps. Jim

There has to be conversion because they they weren't smart enough to make a XPAP that ran on 24 volts or didn't require a extra circuit, so that can keep price gouging the customer.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by prodigyplace » Mon Jun 18, 2018 5:45 pm

Goofproof wrote:
Mon Jun 18, 2018 5:22 pm
prodigyplace wrote:
Mon Jun 18, 2018 5:08 pm
D.H. wrote:
Mon Jun 18, 2018 3:31 pm
I suggest that if you're starting with a DC power supply, you should not convert it to AC. When you use this machine at home, the "brick" on the power cord converts household electricity (which is AC) to DC. This is what the machine runs on. It makes no sense to convert it and then convert it back!

I suggest you use the product on the link below.


Link to converter ===> https://www.cpap.com/productpage/resmed ... hines.html
What did you do with the REAL D.H. ?? This post is actually reasonably accurate & useful!

The DC converter actually converts from 12 volts DC to 24 volts DC though. There is still some conversion, but very efficient.
He may have learned to read, it been posted over 6 times, repetition helps. Jim

There has to be conversion because they they weren't smart enough to make a XPAP that ran on 24 volts or didn't require a extra circuit, so that can keep price gouging the customer.
Also, a broken clock is correct twice a day ;)

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense 10 units won't run o n an inverter? Screwed again?

Post by chunkyfrog » Mon Jun 18, 2018 6:02 pm

Hmmmm, makes you wonder what D.H. is short for.
Disturbing hallucination?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by palerider » Mon Jun 18, 2018 6:47 pm

. wrote:
Mon Jun 18, 2018 5:08 pm
. wrote:
Mon Jun 18, 2018 3:31 pm
I suggest that if you're starting with a DC power supply, you should not convert it to AC. When you use this machine at home, the "brick" on the power cord converts household electricity (which is AC) to DC. This is what the machine runs on. It makes no sense to convert it and then convert it back!

I suggest you use the product on the link below.


Link to converter ===> https://www.cpap.com/productpage/resmed ... hines.html
What did you do with the REAL . ?? This post is actually reasonably accurate & useful!
Like I say, DH isn't always wrong... but like a stopped clock, it's just not worth the effort of trying to find the absurdly rare nuggets of correctness among the flood of drivel.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by Goofproof » Mon Jun 18, 2018 7:04 pm

prodigyplace wrote:
Mon Jun 18, 2018 5:45 pm
Goofproof wrote:
Mon Jun 18, 2018 5:22 pm
prodigyplace wrote:
Mon Jun 18, 2018 5:08 pm
D.H. wrote:
Mon Jun 18, 2018 3:31 pm
I suggest that if you're starting with a DC power supply, you should not convert it to AC. When you use this machine at home, the "brick" on the power cord converts household electricity (which is AC) to DC. This is what the machine runs on. It makes no sense to convert it and then convert it back!

I suggest you use the product on the link below.


Link to converter ===> https://www.cpap.com/productpage/resmed ... hines.html


What did you do with the REAL D.H. ?? This post is actually reasonably accurate & useful!

The DC converter actually converts from 12 volts DC to 24 volts DC though. There is still some conversion, but very efficient.
He may have learned to read, it been posted over 6 times, repetition helps. Jim

There has to be conversion because they they weren't smart enough to make a XPAP that ran on 24 volts or didn't require a extra circuit, so that can keep price gouging the customer.
Also, a broken clock is correct twice a day ;)
If you don't mess with the hands. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chartle
Posts: 152
Joined: Mon Mar 12, 2018 7:56 am

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by chartle » Tue Jun 19, 2018 8:13 pm

I guess my question was really if there was some issue that you never want to run an Airsence 10 on an inverter.

The person with the issue had only one option at the time and it was to use an inverter they already had for their fridge. They didn't plan on boondocking for one of the nights on the road and wanted to use it.