RRL wrote:I should have looked back to see specifically which part of the PRS1 behavior you were talking about. I definitely think the probes poking at the sleeper when everything is nice and smooth is perverse. The nurse bit is a great analogy. I guess the idea is that PRS1 is being 'proactive', but it hasn't made any adjustments for me based on the results of probing, so I would think that after some time of fruitless probing the machine would say, "hmm, I guess I'll stop unless I have a reason to start again." Then I thought perhaps there was some association with the VS2 snores I have throughout the night (anywhere from 9 to 22 events lately)....doesn't seem like it: just found a solid 90 minutes of snore-free, event-free time and the machine was poking at me the whole time?!
and
Another thing I noticed: I had 4 OAs and 1 H within 64 minutes. During that time, the machine stopped its persistent probing (except for one small probe) and utilized 6 "pressure pulses" which apparently helped the machine 'decide' they were OAs and Hs, yet no other action was taken - apparently because they were not clustered together enough. Wouldn't continuing to probe right after the first identified OA allow the machine to sense if an improvement could me made via increased pressure, like preventing the next OA? Nope. Found another set of 2 Hs. Machine sits there and "pressure pulses" to confirm they are Hs, no more probing, just waiting for subsequent events. Next, I find 2 OAs over half an hour. This time the machine does the pressure pulses AND the pressure probes throughout the timeframe without any increase in pressure to perhaps prevent the next event from happening. Is this some kind of sick coding prank? What am I not understanding?
I think you have some major misconceptions about both the pressure probing and the pressure pulses and their roles in the PR S1 Auto algorithm going on. You may also have a misconception about what the APAP is actually doing to control your apnea.
Let's start with the most important thing: An APAP is NOT a non-invasive ventilator. The APAP's job is to provide enough additional pressure to make it difficult (but not impossible) for your upper airway to collapse when you are asleep. When you quit breathing due to an apnea, it is NOT the APAP's job to get your breathing started again. Yes, that seems counter intuitive, but APAP is designed to wait for you to wake up enough to open up your airway and start breathing again. Once the event is over, the APAP
may choose to increase the pressure by 1-2 cm if certain conditions are met. In order to increase the pressure after the event is over, one or more of the following typically must be present: (1) the event occurred within 2-3 minutes after the previous event---i.e. there's a cluster of two or more very closely spaced events, or (2) there is evidence of flow limitation in the flow rate curve, which can trigger a pressure increase even without an event, or (3) there is snoring present, which can trigger a pressure increase even without an event. If none of these conditions are met, the machine won't increase the pressure after an event.
Next let's talk about the Pressure Pulses: The Pressure Pulses (PP) are not designed to somehow "bust through" an apnea and open up your airway. Busting through an apnea and forcing your airway open would take more pressure than our CPAP/APAP machines can deliver. The Pressure Pulses are designed to allow the machine to figure out whether a potential apnea is a "Clear airway apnea (CA)" or an "Obstructed airway apnea (OA)". The machine sends the PP out and measures the back pressure. If the back pressure does not increase, the extra air presumably went down to your lungs through a clear airway. If the back pressure does increase, the extra air could not get into your lungs, and the apnea is presumed to be an OA. It's important to determine whether apneas are OAs or CAs because increasing pressure in the presence of CAs is a bad idea: Additional pressure does not prevent more CAs from happening and too much pressure can cause more CAs to happen. PPs are also used by the S1 when it is having a tough time tracking your breathing; this most often happens when there is a very large leak.
Finally lets talk about the pressure probes, which are technically one half of the PR Search Algorithm. The test pressure increases are done when the breathing is relatively stable and the idea is to be
very proactive. The machine sends out a test 2cm pressure increase every 10 minutes or so and it studies the shape of the flow rate curve intensely during the test pressure increase. If subtle improvements are seen, the machine continues to raise the pressure proactively until no further improvements in the shape of the wave flow are seen. If no improvements are seen, the pressure is lowered back to the original setting. The kinds of "improvements" the machine is looking for represent evidence that the lower pressure was allowing very subtle flow limitations to develop, but these subtle flow limitations were far too small to be scored as an official Flow Limitation, Hypopnea, or OA.
After the pressure has been raised (typically because of events, snoring, or FLs), the PR Search Algorithm also has a test pressure
decrease component. Unlike the Resmed machines, the PR S1 does not immediately start decreasing the pressure as soon as the breathing starts to stabilize. Rather, the PR S1 leaves the pressure where it is and then periodically it tests a
decrease in pressure. If any subtle deterioration in the flow rate is detected, the pressure is raised back up to the current level. If no deterioration in the flow rate is detected, the machine continues to decrease the pressure until either the minimum pressure setting is reached OR subtle deterioration is detected, in which case the machine increases the pressure back up to the level it was at with no deterioration.
As for that 64 minute period where you had 4 OAs and 1 H scored with a bunch of pressure pulses and test pressure increases, but no actual pressure increase? My guess is that except for the OAs and the H, your breathing probably looked exceptionally good since the machine was doing the test pressure increase probes. The thing to keep in mind is that many isolated OAs and Hs come with NO warning---in other words they don't occur after snoring or flow limitations have been detected. And because there are no warning signs, there's not much that the machine can do to prevent them. And if they're isolated from each other by more than 2 minutes of normal sleep breathing, the PR S1 is not programmed to increase the pressure in response to them. (And neither is an S9 as I recall.)
I recognize that both PRS1 and S9 help many people and provide effective therapy, but something just doesn't seem right with this PRS1 algorithm to me.
It might help you understand what the PR S1 Auto algorithm is trying to do if you had a better idea of what goes on in a manual titration study. I'll try to find the link to the AASM rules for manual titration studies tomorrow night. There's also an old 2009 article that is a benchmark study of the then state-of-the-art auto algorithms from several different companies. Somethings (such as OA vs. CA detection) have changed a lot since 2009, but some of the basic decisions about when to increase/decrease pressure and how much to increase/decrease the pressure are still in place in the algorithms used by today's machines. I'll also try to find my link to that article if you would like to read it.
Maybe I'm overreacting. Maybe my PRS1 is actually preventing a lot of events and providing optimal therapy. Maybe the S9 wouldn't prevent events as well. I have no idea. I just want to feel better. Maybe I need to stop looking at the data!
My best guess is that your PRS1 is doing a good job of preventing most of your events. My guess is that the S9 would also do a good job at preventing most of your events. Your AHI might even be lower the the S9, but that may or may not make you feel any better. And perhaps you also have a bit of Resmed S9 AutoSet envy. A lot of S9 users think no other machine can possibly be as good as the S9 and they're not shy about saying it around here. (And the S9 is a very good machine. But so is the PR S1.)
Looking at your data is a two-edged sword. If you look at it the first thing in the morning and say to yourself, "Oh shit, I had 5 events last night in a 50 minute period and the machine didn't do a damn thing. And I feel lousy and I bet those events are why I feel lousy.", then looking at the data every day may be negatively influencing you. But if you look at the data and the first thing that goes through your mind is "At least the AHI is below 3.0 today so the machine must be doing something right even though I feel lousy", then the data may be helping you to keep going. (For what it's worth, when I was a newbie, I needed to look at the data every day just to reassure myself that the machine was doing it's job even though I was feeling a whole lot worse than I had before starting CPAP.)
I'll end by sharing something that Morbius said more than once in his many guises: CPAP doesn't fix
bad sleep. CPAP fixes sleep disordered breathing and nothing else. If there's anything else wrong with your sleep, you can still feel lousy even though CPAP has fixed your sleep disorder breathing.
And now it's time for me to get some sleep.