Deep Cycle Home Backup Supply
Deep Cycle Home Backup Supply
Someone PMed me and asked for recommendations for a backup power supply. Rather than send it to one person, here's what I suggested.
This is for a Philips PRS1 machine.
There are quite a few options, each with its own advantages and drawbacks. For home backup use, I recommend this:
1) Deep Cycle marine battery about the size of a car battery.
2) Battery box.
3) "Float" charger
4) Adapters for your CPAP
It's sort of big, so find a place to put it next to your CPAP where you won't trip over it or kick it over, and you're safe.
Parts
1) Battery - Get a "Deep cycle" marine battery. Make sure it doesn't say anything about hybrid, starting, dual purpose, etc. It can mention trolling.
Here's one example. http://www.walmart.com/ip/EverStart-Max ... ifications
2) Battery box - Get one that your battery will fit in.
One example: http://www.walmart.com/ip/Snap-Top-Ever ... x/16781380 is one example.
3) Float charger. You need one that can stay hooked up to the battery all the time and not damage it. I've used several types and had several that are supposed to be a "float" charger, but they still ate the battery up over time. The one that's worked for me through several chargers and batteries is the Schumacher SE-1-12s. The last time I bought one, the cheapest and easiest place to get one was AutoZone. http://www.autozone.com/autozone/access ... 513702_0_0_
4) Cables. For the Philips Respironics PRS1 (Not 60 series), the adapter is https://www.cpap.com/productpage/respir ... -cord.html. For 60 series, it's https://www.cpap.com/productpage/dc-pow ... hines.html.
These plug into a cigarette lighter plug. You need a battery to cigarette plug adapter. https://www.cpap.com/productpage/respir ... cable.html You can buy this item locally, but be sure to get one with a fuse near the battery clips.
Assembly
Put the battery in the box. I set the box inside another plastic storage box, just in case I get some kind of leak from both the battery and the battery box. Be sure you won't kick it over or trip on it.
Hook the charger up to the screw terminals on the battery. Clip the cigarette lighter/clip cable to the battery. Get the polarity right. Red to plus, black to negative. Plug the Respironics cable into the cigarette plug.
Check the water level. Note that the water level should not be at the top. It should be at the bottom of the little tube you look through. http://www.wikihow.com/Check-Car-Battery-Water-Levels. Distilled water is good.
Plug the battery charger into the wall. The green light should come on. Watch it for a few days, and the red light should come on when the battery is fully charged. It make take several days if the battery is new.
Put the top of the battery box on and strap it on. The top keeps you from dropping something on the top and shorting the battery out. Strap the cover on.
When power goes out, unplug the AC power adapter from the CPAP and plug in the battery cord.
Check the water level after 1 month. Then about every 6 months or so. More often if you're using a different charger.
When the power goes out, unplug the AC adapter from the CPAP and plug in the adapter from the battery.
There are many other ways to do this with different levels of cost, portability, convenience, etc.
Notes:
[edited 11/11/2012] - This system recharges slowly. This is fine for normal standby use, but don't think you're going to run a generator for a few hours during a long outage and quickly recharge the battery. It may take a few days to recharge after you use the battery for a few nights.
This is for a Philips PRS1 machine.
There are quite a few options, each with its own advantages and drawbacks. For home backup use, I recommend this:
1) Deep Cycle marine battery about the size of a car battery.
2) Battery box.
3) "Float" charger
4) Adapters for your CPAP
It's sort of big, so find a place to put it next to your CPAP where you won't trip over it or kick it over, and you're safe.
Parts
1) Battery - Get a "Deep cycle" marine battery. Make sure it doesn't say anything about hybrid, starting, dual purpose, etc. It can mention trolling.
Here's one example. http://www.walmart.com/ip/EverStart-Max ... ifications
2) Battery box - Get one that your battery will fit in.
One example: http://www.walmart.com/ip/Snap-Top-Ever ... x/16781380 is one example.
3) Float charger. You need one that can stay hooked up to the battery all the time and not damage it. I've used several types and had several that are supposed to be a "float" charger, but they still ate the battery up over time. The one that's worked for me through several chargers and batteries is the Schumacher SE-1-12s. The last time I bought one, the cheapest and easiest place to get one was AutoZone. http://www.autozone.com/autozone/access ... 513702_0_0_
4) Cables. For the Philips Respironics PRS1 (Not 60 series), the adapter is https://www.cpap.com/productpage/respir ... -cord.html. For 60 series, it's https://www.cpap.com/productpage/dc-pow ... hines.html.
These plug into a cigarette lighter plug. You need a battery to cigarette plug adapter. https://www.cpap.com/productpage/respir ... cable.html You can buy this item locally, but be sure to get one with a fuse near the battery clips.
Assembly
Put the battery in the box. I set the box inside another plastic storage box, just in case I get some kind of leak from both the battery and the battery box. Be sure you won't kick it over or trip on it.
Hook the charger up to the screw terminals on the battery. Clip the cigarette lighter/clip cable to the battery. Get the polarity right. Red to plus, black to negative. Plug the Respironics cable into the cigarette plug.
Check the water level. Note that the water level should not be at the top. It should be at the bottom of the little tube you look through. http://www.wikihow.com/Check-Car-Battery-Water-Levels. Distilled water is good.
Plug the battery charger into the wall. The green light should come on. Watch it for a few days, and the red light should come on when the battery is fully charged. It make take several days if the battery is new.
Put the top of the battery box on and strap it on. The top keeps you from dropping something on the top and shorting the battery out. Strap the cover on.
When power goes out, unplug the AC power adapter from the CPAP and plug in the battery cord.
Check the water level after 1 month. Then about every 6 months or so. More often if you're using a different charger.
When the power goes out, unplug the AC adapter from the CPAP and plug in the adapter from the battery.
There are many other ways to do this with different levels of cost, portability, convenience, etc.
Notes:
[edited 11/11/2012] - This system recharges slowly. This is fine for normal standby use, but don't think you're going to run a generator for a few hours during a long outage and quickly recharge the battery. It may take a few days to recharge after you use the battery for a few nights.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Last edited by archangle on Tue Nov 13, 2012 12:21 pm, edited 1 time in total.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Deep Cycle Home Backup Supply
This is EXACTLY what I was looking for: step-by-step directions and links to everything one should get for a battery backup. I can't thank you enough, Archangel. And you even included stuff for 60 series machines, which is what I have. You're a blessing.archangle wrote:Someone PMed me and asked for recommendations for a backup power supply. Rather than send it to one person, here's what I suggested.
Avoid tooth extractions (including wisdom teeth) & train-track braces; find a functional orthodontist at http://iaortho.org/.
Re: Deep Cycle Home Backup Supply
This is the one I built and it's a lot like the one mentioned by archangle. This is the one by John Fisher viewtopic.php?t=49115
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1 |
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV
Re: Deep Cycle Home Backup Supply
This is certainly a good way to go. Great post. Something that some may still have time to put together in a short period of time if you can start first thing in the morning. Generators are sold out in many areas by us and as far north as Rochester, NY from a visit that we had a little while ago. Our Tractor Supply was sold out by 10:am Saturday Morning, and BJ's wholesale before Friday night. Wanted to bump this Post up!
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: OSA, Fibromyalgia, Rheumatoid Arthritis, Osteoarthritis, Gout, PTSD, & Just a few others! |
Tim
Finger Lakes Region, NY
S9 AutoSet 4cm-12cm
Finger Lakes Region, NY
S9 AutoSet 4cm-12cm
- JohnBFisher
- Posts: 3821
- Joined: Wed Oct 14, 2009 6:33 am
Re: Deep Cycle Home Backup Supply
I previously posted instructions in the Post "Battery Backup Design" .. It's a fairly simple approach to use a deep cycle battery to act as a backup battery. I've used this approach for about 10 years at this point. I am on my second battery. It should last another few years. I test the system about once a month. Even if you don't implement it fully, there is a lot of information that everyone contributed into that post.
Battery Backup Design
viewtopic.php?t=49115
Battery Backup Design
viewtopic.php?t=49115
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O |
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
Re: Deep Cycle Home Backup Supply
Just curious -- how many nights of dc power will your setup provide?archangle wrote:Someone PMed me and asked for recommendations for a backup power supply. Rather than send it to one person, here's what I suggested.
This is for a Philips PRS1 machine.
For those folks who live in the Northeast, power outages from Sandy could extend out one or two weeks. Unless you have a generator (in which case you wouldn't need the battery backup), there is no way to re-charge ANY kind of battery.
................21+ years of restorative, apnea-free sleep.
-
Guest
Re: Deep Cycle Home Backup Supply
you can use your car to charge batteriesSloop wrote:Unless you have a generator (in which case you wouldn't need the battery backup), there is no way to re-charge ANY kind of battery.
a neighbors if they have a generator
or at a nearby shelter as i am sure they will have power
battery power should be enuff to get you thru 2 or 3 nites
note that batteries take longer to charge then to discharge
get 2 batteries so you can charge 1 while using the other
Re: Deep Cycle Home Backup Supply
I use the deep cycle marine battery set up and have for over 2 years. i bought a small foldable trolly at Costco for less that $20 and put my battery set up on it. Now I can wheel it around wherever I want it, since winter is almost upon us I have wheeled it in to my bedroom and put it in a safe corner of the room. I have used it for 4 nights at one time, without the humidifier, and got 8 hours each night. Hope that helps you gals out....if an old lady like me can set this up..so can you
Stay safe my frfiends
Cheers
Nan
Stay safe my frfiends
Cheers
Nan
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Started cpap in 2010.. still at it with great results.
Re: Deep Cycle Home Backup Supply
It's hard to tell for sure, power usage varies between CPAPs, varies with settings, leak rates, batteries vary, batteries age, etc. My CPAP draws around 2 amps without humidifier. 2 Amps would be about 16 Amp Hours per night. That's around 5 nights. Maybe 4 nights with some safety margin.Sloop wrote:Just curious -- how many nights of dc power will your setup provide?
For those folks who live in the Northeast, power outages from Sandy could extend out one or two weeks. Unless you have a generator (in which case you wouldn't need the battery backup), there is no way to re-charge ANY kind of battery.
Your results may vary widely.
It is a good point that any battery system does eventually run out of power and you have to do something. Exceed that limit and you'll need to move out, get power some other way, or figure out how to charge from your car or other source.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Deep Cycle Home Backup Supply
What an extra GREAT tip!! You all are such great helpnanwilson wrote: i bought a small foldable trolly at Costco for less that $20 and put my battery set up on it. Now I can wheel it around wherever I want it, since winter is almost upon us I have wheeled it in to my bedroom and put it in a safe corner of the room.
Avoid tooth extractions (including wisdom teeth) & train-track braces; find a functional orthodontist at http://iaortho.org/.
- DiverCTHunter
- Posts: 484
- Joined: Thu Jan 05, 2012 11:48 am
- Location: Cleveland, TN
Re: Deep Cycle Home Backup Supply
Archangel:
What are your thoughts on inserting a voltage regulator between the battery and the machine's DC-DC adaptor?
I suspect both PR and ResMed are being conservative about their machine's voltage sensitivity, but since both recommend against using the DC-DC adapter while the alternator is running, I'm tempted to err on their side. My PRS1's power brick gives between 12.21 and 12.24 VDC under no load, so I'm not willing to subject it to the 13.5VDC or 14.2VDC that the float charger puts out in Float and Bulk charge modes unless others have done it with no ill effects.
Here's the setup I'm considering for home. If it works as planned, then the 60-watt PR power brick will go in the travel bag:
PSU:
300-watt AC->DC PSU from old PC @ +12VDC (could also use the -12VDC power rail for ground reference), but my DC Theory always references to earth ground & I doubt the common buss is isolated on the combination voltage regulator / charge controller)
Combo Charge Controller / Voltage Regulator:
Minibox Corp OpenUPS DC-DC voltage regulator and mixed-chemistry charge controller
The pics show 3 buck/boost transformers, so I should be OK using +12VDC in instead of +14VDC in. I can always swap out the PSU for a better one at a later date.
Battery:
28 Ah Sealed Lead-Acid bank (2x load-tested good 14Ah batteries from discarded UPSs, wired in parallel) Battery Cutoff voltage programmed for 70% max discharge gives ~20 Ah usable which should be plenty for 1 night with 8-11.5 cm/H2O APAP.
Charge controller can handle single parallel banks at up to 24V battery bank (30V bulk charge) at 3A max and can also be configured for up to 6 single cells batteries (2V or 4V cells) in a modified serial config, so I can always slip a flooded deep-cycle or AGM trolling battery in there later.
What are your thoughts on inserting a voltage regulator between the battery and the machine's DC-DC adaptor?
I suspect both PR and ResMed are being conservative about their machine's voltage sensitivity, but since both recommend against using the DC-DC adapter while the alternator is running, I'm tempted to err on their side. My PRS1's power brick gives between 12.21 and 12.24 VDC under no load, so I'm not willing to subject it to the 13.5VDC or 14.2VDC that the float charger puts out in Float and Bulk charge modes unless others have done it with no ill effects.
Here's the setup I'm considering for home. If it works as planned, then the 60-watt PR power brick will go in the travel bag:
PSU:
300-watt AC->DC PSU from old PC @ +12VDC (could also use the -12VDC power rail for ground reference), but my DC Theory always references to earth ground & I doubt the common buss is isolated on the combination voltage regulator / charge controller)
Combo Charge Controller / Voltage Regulator:
Minibox Corp OpenUPS DC-DC voltage regulator and mixed-chemistry charge controller
The pics show 3 buck/boost transformers, so I should be OK using +12VDC in instead of +14VDC in. I can always swap out the PSU for a better one at a later date.
Battery:
28 Ah Sealed Lead-Acid bank (2x load-tested good 14Ah batteries from discarded UPSs, wired in parallel) Battery Cutoff voltage programmed for 70% max discharge gives ~20 Ah usable which should be plenty for 1 night with 8-11.5 cm/H2O APAP.
Charge controller can handle single parallel banks at up to 24V battery bank (30V bulk charge) at 3A max and can also be configured for up to 6 single cells batteries (2V or 4V cells) in a modified serial config, so I can always slip a flooded deep-cycle or AGM trolling battery in there later.
Last edited by DiverCTHunter on Mon Oct 29, 2012 12:39 pm, edited 2 times in total.
When in doubt, open the case. Remember: If you can't open it, you don't own it!
Prescribed APAP range - 6-10 cm/H2O, titrated at 8.
Current range - 9.0-11.5 cm/H2O - still searching for the magic "zero night" but averaging 2.2 AHI
Prescribed APAP range - 6-10 cm/H2O, titrated at 8.
Current range - 9.0-11.5 cm/H2O - still searching for the magic "zero night" but averaging 2.2 AHI
- Lizistired
- Posts: 2835
- Joined: Tue Dec 14, 2010 10:47 pm
- Location: Indiana
Re: Deep Cycle Home Backup Supply
2Tim, I hope you are talking about John Fisher's setup. Arch's is for PRS1 machines. S9's need the Resmed DC adapter, unless there has been another solution I haven't read about.2Tim215 wrote:This is certainly a good way to go. Great post. Something that some may still have time to put together in a short period of time if you can start first thing in the morning. Generators are sold out in many areas by us and as far north as Rochester, NY from a visit that we had a little while ago. Our Tractor Supply was sold out by 10:am Saturday Morning, and BJ's wholesale before Friday night. Wanted to bump this Post up!
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better. |
ResScan: http://www.resmed.com/int/assets/html/s ... c=patients
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
Re: Deep Cycle Home Backup Supply
I only tested my rig to the equivalent of two 8-hour nights. But as I said in another post, the meter barely budged off of full charge, so I guess it is possible I too might get around 4 nights. I can say this -- I hope I never have to find out!archangle wrote:It's hard to tell for sure, power usage varies between CPAPs, varies with settings, leak rates, batteries vary, batteries age, etc. My CPAP draws around 2 amps without humidifier. 2 Amps would be about 16 Amp Hours per night. That's around 5 nights. Maybe 4 nights with some safety margin.Sloop wrote:Just curious -- how many nights of dc power will your setup provide?
For those folks who live in the Northeast, power outages from Sandy could extend out one or two weeks. Unless you have a generator (in which case you wouldn't need the battery backup), there is no way to re-charge ANY kind of battery.
................21+ years of restorative, apnea-free sleep.
Re: Deep Cycle Home Backup Supply
You might get more if your current draw is low. I haven't done any long term tests of my CPAP with the Kill a Watt power meter and no humidifier, so the average might be lower than I think. Maybe when it gets humid, I'll run without humidifier overnight and get better numbers.Sloop wrote:archangle wrote:
I only tested my rig to the equivalent of two 8-hour nights. But as I said in another post, the meter barely budged off of full charge, so I guess it is possible I too might get around 4 nights. I can say this -- I hope I never have to find out!
Don't put much faith in the power meter. The way lead acid batteries discharge, the voltage drops very slowly, then sort of falls off a cliff. It's very difficult to make a cheap "remaining power" meter for rechargeable batteries.
Here's an example:

In practice, building a "remaining power" battery meter is even tougher than that diagram would indicate.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
- DiverCTHunter
- Posts: 484
- Joined: Thu Jan 05, 2012 11:48 am
- Location: Cleveland, TN
Re: Deep Cycle Home Backup Supply
Yup, that's why I was originally going to use the "true sine wave" APC UPS I pulled out of the pile at the recycling center. Unfortunately, the board and both of the 12V batteries in the 24V bank were bad.archangle wrote: In practice, building a "remaining power" battery meter is even tougher than that diagram would indicate.
The board I'm considering for DC-DC UPS operation has a "columb counter" feature, so theoretically I can use the management software to set the charge curve and approx Amp Hours of my batteries, then it'll give an "accurate" charge meter.
Oh well, at least the APC got separated into Lead, Ferrous and Copper/Ferrous here instead of in a fire at some slum in Africa.
When in doubt, open the case. Remember: If you can't open it, you don't own it!
Prescribed APAP range - 6-10 cm/H2O, titrated at 8.
Current range - 9.0-11.5 cm/H2O - still searching for the magic "zero night" but averaging 2.2 AHI
Prescribed APAP range - 6-10 cm/H2O, titrated at 8.
Current range - 9.0-11.5 cm/H2O - still searching for the magic "zero night" but averaging 2.2 AHI