I do not understand why my centrals are increasing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
red1
Posts: 90
Joined: Sat Oct 22, 2011 12:13 am

I do not understand why my centrals are increasing

Post by red1 » Sat Mar 24, 2012 11:39 pm

Over the last year, my centrals have been kept at about .05. Over the past few months, I began to get the odd day, maybe twice a month, where I would have between a central index of 3 & 6.

Lately, meaning the 8 days, they have been getting higher. Yesterday I had a central index of 11.5 & today, I had a central index of 24.9.

It is really beginning to scare me. I spoke to my sleep doc & respirologist about this a copule months back, she said its nothing & you certainly do not need a vpap adapt sv. Thats when she gave me the Vpap Auto. During these centrals, the vpap auto pressure does not change at all. I wore my oxymeter last night, the flow of the oxygen levels on the oxymeter report looked just like the respiration rate flow of the ResScan report, where I had the events.

I wish I knew what what causing them. Someone mentioned to me a couple months back that the waveform looked like cheyne stokes. I do not have any heart condition that I am aware of. I just went through a stress test, a 48 hour halter monitor, and a echocardiogram & it was all fine. I have a little swelling in my right lower leg, just above the anckle. They did a altrasound & took blood pressures from my legs & it was fine. I do know I my c02 blood work sits at the high end of normal.

Yesterday, the 11.5 was the highest central index I had, but now, the 24.9, wow, thats more then double. I really am at a loss here. I do not know what do do or what to think. Here are a couple screen shots from last night. I broke the centrals up into two pages, so I could keep them large enough to read.
My Spntaneous cycled breathing was way down to 95%. My I & E median ratio was up to 1:2.04 from my average median of 1:1.64

Thank you for your help. Nota much luch posting these images. I will have to try the link method.

Image

Image

The images come out smaller from photo bucket.
Last edited by red1 on Sun Mar 25, 2012 5:02 pm, edited 7 times in total.

User avatar
Pugsy
Posts: 65242
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I do not understand why my centrals are increasing

Post by Pugsy » Sun Mar 25, 2012 8:05 am

Use the IMG brackets on the Editor (next to Url brackets) and paste the image location inside of the brackets and you get these. Someone is likely going to ask you to zoom in closer on the section of time where the centrals are flagged one after another. So can you zoom in closer on the time between 2: 50 or so..maybe 10 minute segment or less. See the breath form up close.

Image

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: I do not understand why my centrals are increasing

Post by robysue » Sun Mar 25, 2012 8:44 am

red1,

First the "easy" stuff: To get the image to show up directly in the post, you need to put the link inside an
[ img ] ... [ /img ] command where there are no blanks between the []'s and the img and \img "words". If you Quote this message you'll see how I did it for the links you provided.

So here are red1's graphs:
Image

Image

Now for the harder stuff.

You write:
Over the last year, my centrals have been kept at about .05. Over the past few months, I began to get the odd day, maybe twice a month, where I would have between a central index of 3 & 6.

Lately, meaning the 8 days, they have been getting higher. Yesterday I had a central index of 11.5 & today, I had a central index of 24.9.

It is really beginning to scare me. I spoke to my sleep doc & respirologist about this a copule months back, she said its nothing & you certainly do not need a vpap adapt sv. Thats when she gave me the Vpap Auto.

.....

Yesterday, the 11.5 was the highest central index I had, but now, the 24.9, wow, thats more then double. I really am at a loss here. I do not know what do do or what to think.
Sounds like a couple of months back you were seeing the odd day here and there with a CAI between 3 and 6. And the sleep doc was probably right at the time: It's nothing to get too worried about if you see one index slightly elevated now and then. Particularly if you can tie the bad numbers to a bad night for sleep in some other sense.

But---something's changed and now you're seeing very high numbers on a frequent basis. I'd call the sleep doc back on Monday and give a report of the last 7 or 8 days of numbers along with a short synopsis of how you're feeling during the day. Exhausted? Fog brained? OK, but not great? Same as always? Whatever it is, the description of how you're feeling is important for the doc to be able to evaluate the high CAI in context.
During these centrals, the vpap auto pressure does not change at all. I wore my oxymeter last night, the flow of the oxygen levels on the oxymeter report looked just like the respiration rate flow of the ResScan report, where I had the events.
The machine is intentionally designed to NOT change the pressure in response to clear airway events. The rationale: Clear airway events occur with a unobstructed, non-collapsed airway and are presumed to be central apneas---i.e. the problem is presumed to be caused by the brain failing to send the signal to breath instead of the airway collapsing. Since the airway is already open, there's no physical reason to increase the air pressure (increased air pressure is supposed to make it more difficult for the airway to collapse.) And since some people are prone to develop central apneas at higher pressures, it makes sense to record the CAs, but not respond to them by increasing the pressure since increased pressure may make the situation worse, not better.
I wish I knew what what causing them. Someone mentioned to me a couple months back that the waveform looked like cheyne stokes. I do not have any heart condition that I am aware of. I just went through a stress test, a 48 hour halter monitor, and a echocardiogram & it was all fine. I have a little swelling in my right lower leg, just above the anckle. They did a altrasound & took blood pressures from my legs & it was fine. I do know I my c02 blood work sits at the high end of normal.
How long has the CO2 blood work been at the high end of normal? That could be evidence pointing towards a problem with your body's system to regulate CO2 and that could make it more difficult for your body's automatic nervous system to use CO2 blood levels to trigger breathing in the normal way for sleep breathing. Is the sleep doc aware that you've been tested for heart disease? Is the sleep doc aware of your CO2 blood work? One sad thing about our medical system here in the US is that specialists seldom actually talk to each other. And it's not even clear that specialist A reads what specialist B has written---if specialist B even sends specialist A a copy of the report.

And for what it's worth, to this patient's eyeballs, yes, your breathing pattern does seem to resemble periodic breathing and possibly Cheyne Stokes.
My Spntaneous cycled breathing was way down to 95%. My I & E median ratio was up to 1:2.04 from my average median of 1:1.64
I'm not sure what you are talking about here. I didn't think the S9 VPAP Auto had a mode where it measured spontaneous (patient-triggered/initiated) breaths vs. non-spontaneous (machine-triggered?) breaths since the VPAP Auto is just an Auto Bi-level that does not have the ability to trigger a patient to breathe---i.e. the VPAP Auto does not act like a non-invasive ventilator as near as I know.

Or do you have a VPAP Auto ST machine set with a back-up breaths per minute rate? That's a different machine altogether and much closer to the VPAP Adapt SV machine.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
red1
Posts: 90
Joined: Sat Oct 22, 2011 12:13 am

Re: I do not understand why my centrals are increasing

Post by red1 » Sun Mar 25, 2012 3:44 pm

Thanks for your comments & help Pugsy & Rubysue.

THis next image is has been set to 5 minutes, so you can see everything better. I do have the vpap auto. It shows I & E ratios & the percent of spontaneous breaths. Normally I am at 99 percent. I do not understand why they show it, but they do.

I also noticed in the above images, that breathing carried on after the centrals for over 15 minutes into the next string of centrals.

Thank you again.

Image


Image

This oximeter page below is not part of the other graphs. I did not wear it that night. This section below is from the day before & it shows when I was having centrals. I noticed how much they looked like respiratory rate from the ResScan program. Showing just how my o2 dropped with each central I had.


Image
Last edited by red1 on Sun Mar 25, 2012 6:03 pm, edited 3 times in total.
Air Curve 10 VAuto
Swift FX Nasal

User avatar
Pugsy
Posts: 65242
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I do not understand why my centrals are increasing

Post by Pugsy » Sun Mar 25, 2012 4:19 pm

There is indeed a subsection in the Respiratory Rate regarding patient triggered breathing on the VPAP Auto.
Here is mine from last night. I think I probably should wait for the "central" gurus to chime in on your problem. My original questions was meant more to gather more details for them than me.
Another question? Any change in meds of any kind? Even OTC.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
red1
Posts: 90
Joined: Sat Oct 22, 2011 12:13 am

Re: I do not understand why my centrals are increasing

Post by red1 » Sun Mar 25, 2012 4:37 pm

There has been no change in my medications.

There has been no change in anything, that I can think of. I did have surgery on March 21 for 2 hernias. Even then, I was still having the odd raised numbers in centrals before that. Nothing near what they have reached now.

Could the general anaesthetic of done something?
Air Curve 10 VAuto
Swift FX Nasal

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: I do not understand why my centrals are increasing

Post by robysue » Mon Mar 26, 2012 8:12 am

red1 wrote:There has been no change in my medications.

There has been no change in anything, that I can think of. I did have surgery on March 21 for 2 hernias. Even then, I was still having the odd raised numbers in centrals before that. Nothing near what they have reached now.

Could the general anaesthetic of done something?
That's a question for a doctor. Or more likely---the doctors plural. As in that's a question I'd ask both the sleep doc and the anaesthesislogist.

And are you in any residual pain from the surgery? Are you taking any prescription pain meds for the surgery?

And something's changed in that the AHI has gone up and the CAI has gone way, way up. It's not triggered by any changes in your lifestyle that you are aware of, but the change is still there. That's why I think you should report this to your sleep doc. Maybe the doc can come up with some plausible explanations and, if you're lucky, offer some reasonable advice on what to do.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
red1
Posts: 90
Joined: Sat Oct 22, 2011 12:13 am

Re: I do not understand why my centrals are increasing

Post by red1 » Mon Mar 26, 2012 12:27 pm

I never took any pain meds after the surgery, as much as I really needed it. I did not take anything for pain, just to avoid any problems, like the one that has happened.
Air Curve 10 VAuto
Swift FX Nasal

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: I do not understand why my centrals are increasing

Post by robysue » Mon Mar 26, 2012 1:16 pm

red1 wrote:I never took any pain meds after the surgery, as much as I really needed it. I did not take anything for pain, just to avoid any problems, like the one that has happened.
Could the pain itself be adversely affecting your sleep and making it far more difficult to get to sleep? In other words, some folks have problems with centrals when they are transitioning to/from sleep and it's not clear at all how many of those centrals are "real" in the sense of whether they'd be scored as a central or simply ignored on a PSG with the EEG data available to the tech.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
red1
Posts: 90
Joined: Sat Oct 22, 2011 12:13 am

Re: I do not understand why my centrals are increasing

Post by red1 » Mon Mar 26, 2012 4:49 pm

The pain was only severe the first 3 days after surgery & then only when I moved, caughed or sneezed. It's been gone for some time now.
Air Curve 10 VAuto
Swift FX Nasal