Hello,
I picked up the s9 elite yesterday and used it last night. On my resmed s8 my leak, with the same mask I am using was only 0.2 Now with the s9 it's reading 4l/s. Would have to double check to see if it's L/s, or L/m, but I think seconds. Could my leak be worse with this machine. Mask fit says excellent. Do have the new slim line tubing.
Thanks for any replies.
got my resmed s9 ? about leak
Re: got my resmed s9 ? about leak
It's 4 Liter per Min and it's good b/c it is below 24 L/Min, the allowed maximum.
When I started CPAPing at the beginning of 2011, I had lousy results using my S9 Elite and a full face Fisher & Paykel mask. Looking back at it I can't believe it how bad they were:



Compare those results to my present data with my present machine, new mask, and higher pressure:
{nice results}:




When I started CPAPing at the beginning of 2011, I had lousy results using my S9 Elite and a full face Fisher & Paykel mask. Looking back at it I can't believe it how bad they were:



Compare those results to my present data with my present machine, new mask, and higher pressure:
{nice results}:




_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
Last edited by avi123 on Fri Mar 07, 2014 11:16 am, edited 5 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
Re: got my resmed s9 ? about leak
The S9 reports in liters per minute. Your S8 reported in liters per second.
With the S9 your leak number to avoid is 24 L/min...so your 4 L/min means minimal leaks .
24 L/min is the same as your old number to avoid with the S8 of 0.40...just multiply liters per second by 60 seconds.
With the S9 your leak number to avoid is 24 L/min...so your 4 L/min means minimal leaks .
24 L/min is the same as your old number to avoid with the S8 of 0.40...just multiply liters per second by 60 seconds.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
Last edited by Pugsy on Sat Mar 08, 2014 3:15 am, edited 1 time in total.
I may have to RISE but I refuse to SHINE.
Re: got my resmed s9 ? about leak
Wow, thank you for the replies. Love the data! That's great. So I guess I'm good.!!!
Twink