You have a full face mask with an anti-asphyxia vent, so you shouldn't have an asphyxiation problem anyway.Roger2 wrote:Could that happen if you have external O2 bleeding in at the connection where the airhose attaches to the cpap machine? I have 4 ltrs comming in at that point.
If you have an oxygen concentrator, it should shut down too if it's a power failure. Are you on bottled oxygen? I guess the CPAP could quit for other reasons than a power outage and the concentrator keeps going.
Assuming your AA vent was blocked, and the O2 continued to flow, I presume 4 liters/minute would give you an air quality that's pretty good. Heck, without the CPAP machine blowing "competing" air into your mask, you might actually have higher O2 than you get with the CPAP running. You wouldn't have any CPAP pressure, so your apnea would be back.
I know that too much oxygen can be a medical problem. I don't know how bad the problem could get if your CPAP quits and O2 keeps going. Your CPAP/O2 setting is designed to give you the right mix of O2 when the machine is blowing, not when the O2 is going and the machine isn't.
Now to throw in even more doom and gloom, there are all kinds of "death and destruction" warnings about hooking O2 up to a CPAP machine. You're warned to turn the O2 off before turning off the CPAP. Repironics says this in one of their manuals:
"When using oxygen with this system, turn the device on before turning on the oxygen. Turn the oxygen off before turning the device off. This will prevent oxygen accumulation in the device. Explanation of the Warning: When the device is not in operation and the oxygen flow is left on, oxygen delivered into the tubing may accumulate within the device’s enclosure. Oxygen accumulated in the device enclosure will create a risk of fire."
"When using oxygen with this system, a Respironics Pressure Valve must be placed in-line with the patient circuit between the device and the oxygen source. The pressure valve helps prevent the backflow of oxygen from the patient circuit into the device when the unit is off. Failure to use the pressure valve could result in a fire hazard."
I presume the "pressure valve" is a safety thing to keep O2 from backing up into the machine if the machine goes off and the O2 is left on. This probably changes the physics of the situation if your anti-asphyxia valve fails.
In your case, Roger2, I think too many things have to fail at once for a real problem to happen. I would be sure to turn off the O2 if the machine ever quits, and to not use the mask if you ever decide the anti-asphyxia vent has failed.