Fired My Sleep Doctor

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 12:10 pm

robysue wrote:1) Do you feel as though you are doing a lot of tossing and turning on some nights? Is restlessness during the night a problem? Are you waking up frequently or having trouble getting back to sleep when you do wake up? If so, did this start around the time your AHI started to creep upwards?
No. I go to sleep and sleep soundly all night. Occasionally, I will wake up to go to the bathroom. I fall back to sleep quickly after getting back in bed.

2) Has anything changed in your life or health other than the deteriorating OSA problem? Are you under more stress at work or in your private life? Any new medical diagnoses? Any changes to medication?[/quote]

The only change to medication was with the medication pump. The pump was replaced in December. The dose was changed.
Finally, have you had any luck in getting the sleep doc to look at the flow rate data your machine records?
No. I don't think he looks at much of anything and doesn't want me to look at anything, either.

I will get the graphs you asked for posted later today.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 12:18 pm

robysue wrote:paper_nanny,

Another thing that really jumps out in your data is the amount of time being flagged as "timed breaths". On four of the five nights you posted, you spent almost the entire night in "timed breaths". The only night without a lot of "timed breaths" is the night with the lowest AHI. All of that points to the hypopneas being central in nature rather than obstructive.
Yes... That happened when the BPM was changed from 5 to Auto. Would it be worth setting it back to 5?

I also noticed that my Patient Triggered Breaths decreased when the BPM was changed to Auto. From past struggles with getting my machine adjusted, I know that I do worse on nights with lower Patient Trigged Breaths. Since I almost never have a problem with large leaks on my mask, I wonder if moving the Patient Triggered Breaths graph up would give people more information to work with or if leaving the leak graph there is a more useful option.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Fired My Sleep Doctor

Post by robysue » Tue Apr 25, 2017 1:16 pm

Paper_Nanny wrote:
robysue wrote:paper_nanny,

Another thing that really jumps out in your data is the amount of time being flagged as "timed breaths". On four of the five nights you posted, you spent almost the entire night in "timed breaths". The only night without a lot of "timed breaths" is the night with the lowest AHI. All of that points to the hypopneas being central in nature rather than obstructive.
Yes... That happened when the BPM was changed from 5 to Auto. Would it be worth setting it back to 5?
That's a question for Pale_Rider and others who use ASVs. I don't have any experience using them.
I also noticed that my Patient Triggered Breaths decreased when the BPM was changed to Auto. From past struggles with getting my machine adjusted, I know that I do worse on nights with lower Patient Trigged Breaths.
Worse in what sense? In the AHI data? In how you actually feel upon waking? Or both?
Since I almost never have a problem with large leaks on my mask, I wonder if moving the Patient Triggered Breaths graph up would give people more information to work with or if leaving the leak graph there is a more useful option.
Given that you are on an ASV with a diagnosed problem with central apnea AND you say that you do worse on nights with lower Patient Triggered Breaths, I think the Patient Triggered Breaths graph would be much more useful for us.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: Fired My Sleep Doctor

Post by Uncle_Bob » Tue Apr 25, 2017 1:38 pm

ChicagoGranny wrote: People that are giving out bad advice should be knocked back into their gopher holes. It can be very difficult for newbies to separate the chaff from the grain.
I'm sure the newbies will be fine. It seems you are having more of a hard time in not being able to moderate someone else's internet forum

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fired My Sleep Doctor

Post by palerider » Tue Apr 25, 2017 1:59 pm

Paper_Nanny wrote:Image
could you re-do that screenshot with the left panel pulled up, to see what the event totals are?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 2:16 pm

robysue wrote:So I have a request: Can you go through your data and find a stretch of breathing on a night when your CAI was above 10 and zoom in on a 5-10 minute stretch of time so that we can see the individual breaths with the CAs flagged in them? Then zoom in on breathing for the data from 4/17 focusing on the time frame between 10:00 and 11:00. Post a couple of 10-minute stretches from that bad hour. The data has to be zoomed in enough to see the individual breaths. If there is not much similarity in the breathing pattern from the night wrere the events were labeled as CAs, then that points to it being reasonable to assume that your current hypopnea problem has a higher chance of being obstructive in nature. If the breathing pattern with the Hs looks almost identical to the breathing pattern with CAs, that raises the distinct possibility that your current problem is with central hypopneas rather than obstructive ones.
Here are some graphs from nights where my CAI was over 10:

Image

Image

Image

Image

And from 4/17:

Image

Image

Image

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Fired My Sleep Doctor

Post by robysue » Tue Apr 25, 2017 2:16 pm

Uncle_Bob wrote:
ChicagoGranny wrote: People that are giving out bad advice should be knocked back into their gopher holes. It can be very difficult for newbies to separate the chaff from the grain.
I'm sure the newbies will be fine. It seems you are having more of a hard time in not being able to moderate someone else's internet forum
+1

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 2:19 pm

robysue wrote:
Paper_Nanny wrote:
robysue wrote:
I also noticed that my Patient Triggered Breaths decreased when the BPM was changed to Auto. From past struggles with getting my machine adjusted, I know that I do worse on nights with lower Patient Trigged Breaths.
Worse in what sense? In the AHI data? In how you actually feel upon waking? Or both?
As far as I remember, it was both.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Fired My Sleep Doctor

Post by robysue » Tue Apr 25, 2017 2:21 pm

Paper_Nanny wrote:
robysue wrote:So I have a request: Can you go through your data and find a stretch of breathing on a night when your CAI was above 10 and zoom in on a 5-10 minute stretch of time so that we can see the individual breaths with the CAs flagged in them? Then zoom in on breathing for the data from 4/17 focusing on the time frame between 10:00 and 11:00. Post a couple of 10-minute stretches from that bad hour. The data has to be zoomed in enough to see the individual breaths. If there is not much similarity in the breathing pattern from the night wrere the events were labeled as CAs, then that points to it being reasonable to assume that your current hypopnea problem has a higher chance of being obstructive in nature. If the breathing pattern with the Hs looks almost identical to the breathing pattern with CAs, that raises the distinct possibility that your current problem is with central hypopneas rather than obstructive ones.
Here are some graphs from nights where my CAI was over 10:
I sure hope PR looks at these graphs. He's got a much better understanding of central apnea stuff than I do.

Having said that, I will say that those Hs look like they may be central Hs to me. If I were you, I'd take PR's advice and open up your ASV machine settings to see if it can do a better job of preventing the breathing instability that leads to the long chains of central events in the first place.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 2:35 pm

palerider wrote:could you re-do that screenshot with the left panel pulled up, to see what the event totals are?

Image

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Fired My Sleep Doctor

Post by robysue » Tue Apr 25, 2017 4:02 pm

Question for Pale Rider about Paper_Nanny's last data:

Why is her machine NOT raising the IPAP pressure very much between 4:45 and 5:45 even though there are gobs of hypopneas and the machine is also triggering something like 2/3's of Paper_Nanny's breaths during that stretch of time? Would a higher min EPAP help at all?

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fired My Sleep Doctor

Post by palerider » Tue Apr 25, 2017 5:10 pm

Paper_Nanny wrote:
palerider wrote:could you re-do that screenshot with the left panel pulled up, to see what the event totals are?
thank you... could I ask you to do another thing, zip up your entire sd card, so, all the files and directories, into one zip file, and upload it to
:
https://www.dropbox.com/request/0LalWvEsijRIgX8sdM7H

that way I can just look through the different days and hopefully get a better idea, instead of asking you for a bunch of screenshots.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 10:48 pm

robysue wrote:Having said that, I will say that those Hs look like they may be central Hs to me. If I were you, I'd take PR's advice and open up your ASV machine settings to see if it can do a better job of preventing the breathing instability that leads to the long chains of central events in the first place.
They look like centrals to me, too. There don't seem to be a lot of vibratory snore or flow limitations in there when the hypopneas are occurring. It seems like they would be there if it were obstructive. Maybe not, but it seems logical.

If they are centrals, will sleep position make any difference? I do sleep on my back sometimes. But I have only heard about that being a factor if they are obstructive events.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Fired My Sleep Doctor

Post by Paper_Nanny » Tue Apr 25, 2017 10:55 pm

palerider wrote:could I ask you to do another thing, zip up your entire sd card, so, all the files and directories, into one zip file, and upload
Let me know if it didn't work out the way you wanted.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fired My Sleep Doctor

Post by palerider » Tue Apr 25, 2017 11:09 pm

Paper_Nanny wrote:
palerider wrote:could I ask you to do another thing, zip up your entire sd card, so, all the files and directories, into one zip file, and upload
Let me know if it didn't work out the way you wanted.
you did fine... I'm looking at it now.

if you feel like trying it, set the minps to either 2 or 3 tonight, and see if that makes a difference. I'll let you know if I can figure out anything more from the old data... though it might be better to just go ahead.

please clarify something for me, you felt better before, or after the backup rate was changed to auto? and when it was changed to auto, the number of timed breaths increased greatly?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.