Should I increase or decrease pressure?

Post by robot_head » Thu Jul 24, 2014 12:06 pm

I keep waking up multiple times a night (4-6) and feeling not well rested in the morning. I am headachey and not super happy right now. My sleepyhead log from last night:
http://i.imgur.com/4RLINQY.png

In picture form:
Image

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: Should I increase or decrease pressure?

Post by bwexler » Thu Jul 24, 2014 12:25 pm

It looks like you are doing quite well with your therapy.
It would help us if you would remove the AHI graph and include the Evans flag graph instead.

The first thing I would do is increase the starting pressure on the ramp to 5, or maybe 6.
Then you could bump up the max by .5 cm and see what it looks like after a week or so.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
justinjustin
Posts: 119
Joined: Sun Oct 06, 2013 11:38 pm

Re: Should I increase or decrease pressure?

Post by justinjustin » Thu Jul 24, 2014 12:26 pm

robot_head wrote:I keep waking up multiple times a night (4-6) and feeling not well rested in the morning. I am headachey and not super happy right now.

It seems like your range is set too wide and the APAP wants to be up around 8cm. Your minimum pressure could probably be bumped up.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Complex Sleep Apnea, mainly CSA, with UARS. RDI of 30 w/o xPAP.

robot_head
Posts: 9
Joined: Thu Jul 24, 2014 12:05 pm

Re: Should I increase or decrease pressure?

Post by robot_head » Thu Jul 24, 2014 12:31 pm

bwexler wrote:It looks like you are doing quite well with your therapy.
It would help us if you would remove the AHI graph and include the Evans flag graph instead.
Here you go:
http://i.imgur.com/TSaGFWg.png
Image
bwexler wrote: The first thing I would do is increase the starting pressure on the ramp to 5, or maybe 6.
Then you could bump up the max by .5 cm and see what it looks like after a week or so.
None of the apnea events occur during the ramp, why do you think this would help?

robot_head
Posts: 9
Joined: Thu Jul 24, 2014 12:05 pm

Re: Should I increase or decrease pressure?

Post by robot_head » Thu Jul 24, 2014 12:32 pm

justinjustin wrote:
robot_head wrote:I keep waking up multiple times a night (4-6) and feeling not well rested in the morning. I am headachey and not super happy right now.

It seems like your range is set too wide and the APAP wants to be up around 8cm. Your minimum pressure could probably be bumped up.
Hmm, I'm not using APAP. I'm using CPAP mode of 10.5cm, but with EPRx3

Mary Z
Posts: 1493
Joined: Wed Oct 28, 2009 9:55 am

Re: Should I increase or decrease pressure?

Post by Mary Z » Thu Jul 24, 2014 1:24 pm

With that AHI I would not make any changes, though the pressure changes suggested seemed sound.
A lot of people don't like meds, but I think a mild sleeping or antianxiety pill would help you sleep through the night. A non prescription idea is any of the PM meds that contain Benadryl (diphenhydramine HCL). I don't find these change my AHI, but help me sleep all night. An alternative is melatonin.
Any idea what is waking you up?

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: PR System One Remstar BiPap Auto AS Advanced.
Dog is my copilot

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Should I increase or decrease pressure?

Post by robysue » Thu Jul 24, 2014 2:15 pm

robot_head,

You ask: Should I increase or decrease pressure?

Until you know what is causing the problem you are trying to fix, you won't have an answer to that question. And, most likely the answer to your question right now is Leave things alone for now.


Here's the thing: You write
robot_head wrote:I keep waking up multiple times a night (4-6) and feeling not well rested in the morning. I am headachey and not super happy right now.
The issue you need to address are the multiple wakes. Those wakes are enough to explain why you are not feeling rested in the morning and are feeling headacey and "not super happy right now."

Your AHI is low enough and there's only one small cluster of events at around 5:20 in the data you posted. So the thing that is driving the excess number of wakes is NOT undertreated OSA. And your leaks are technically under control as well. Meaning the data can be trusted: You are not having large numbers of events that the machine is failing to score because of large leak problems. In other words, your machine is doing its job: You are no longer experiencing a large number of obstructive apneans and hypopneas when you are asleep. That doesn't mean that you are not feeling bad; it does mean that the reason you are feeling bad is bad sleep that's caused by something other than untreated or undertreated OSA.

Important questions for you to think about and answer before you start dial winging the pressure settings:

1) What do you think is causing the wakes in the middle of the night? Is this a new disrupting sleep pattern? Or have you had problems with middle of the night wakes before you started PAP therapy? If the pattern is not new, how long have you had problems with middle of the night wakes? And pre-CPAP, what did you tend to do when you woke up in the middle of the night? And how effective was it?

2) Are you physically uncomfortable when you are trying to get to sleep at the beginning of the night with your mask on? If so, can you describe the kind of discomfort you are experiencing in words?

3) How long does it take you to fall asleep at the beginning of the night with your mask on? 5-10 minutes? 10-20 minutes? More than 20 minutes? More than an hour? And if you are not asleep within about 20 minutes of going to bed, what starts going through your mind?

3) Are you physically uncomfortable when you wake up in the middle of the night with your mask on? If so, can you describe the kind of discomfort you are experiencing in words? Are you feeling overwhelmed by the air pressure when you wake up in the night?

4) You are using a pretty low starting ramp pressure of 4cm. How do you feel right at the beginning of the night and every time you turn the machine off and back on? Are you having trouble catching your breath comfortably at the beginning of the ramp period?

5) What do you do when you wake up? Do you lie in bed for a very long time fighting to get back to sleep? Or are you able to return to sleep relatively quickly after most of the middle of the night wakes?

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Should I increase or decrease pressure?

Post by palerider » Thu Jul 24, 2014 2:30 pm

robot_head wrote: None of the apnea events occur during the ramp, why do you think this would help?
I could be wrong, but I don't *believe* that the machine flags, or responds to events, during ramp, which may be why the suggestion was made. note, I'm not suggesting a change, the only use for ramp is a comfort feature, I personally don't use it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

robot_head
Posts: 9
Joined: Thu Jul 24, 2014 12:05 pm

Re: Should I increase or decrease pressure?

Post by robot_head » Thu Jul 24, 2014 4:11 pm

robysue wrote:robot_head,

You ask: Should I increase or decrease pressure?

Until you know what is causing the problem you are trying to fix, you won't have an answer to that question. And, most likely the answer to your question right now is Leave things alone for now.


Here's the thing: You write
robot_head wrote:I keep waking up multiple times a night (4-6) and feeling not well rested in the morning. I am headachey and not super happy right now.
The issue you need to address are the multiple wakes. Those wakes are enough to explain why you are not feeling rested in the morning and are feeling headacey and "not super happy right now."

Your AHI is low enough and there's only one small cluster of events at around 5:20 in the data you posted. So the thing that is driving the excess number of wakes is NOT undertreated OSA. And your leaks are technically under control as well. Meaning the data can be trusted: You are not having large numbers of events that the machine is failing to score because of large leak problems. In other words, your machine is doing its job: You are no longer experiencing a large number of obstructive apneans and hypopneas when you are asleep. That doesn't mean that you are not feeling bad; it does mean that the reason you are feeling bad is bad sleep that's caused by something other than untreated or undertreated OSA.

Important questions for you to think about and answer before you start dial winging the pressure settings:

1) What do you think is causing the wakes in the middle of the night? Is this a new disrupting sleep pattern? Or have you had problems with middle of the night wakes before you started PAP therapy? If the pattern is not new, how long have you had problems with middle of the night wakes? And pre-CPAP, what did you tend to do when you woke up in the middle of the night? And how effective was it?
This is a new sleep pattern for me, and may be related to quitting medical pot concurrently. I'm undertaking this to keep myself from eating too much and causing gut trouble at night, and dry mouth.
robysue wrote: 2) Are you physically uncomfortable when you are trying to get to sleep at the beginning of the night with your mask on? If so, can you describe the kind of discomfort you are experiencing in words?
I have the mask strapped to my face very tightly to avoid leaks. I would say I'm a bit uncomfortable at the beginning of the night
robysue wrote: 3) How long does it take you to fall asleep at the beginning of the night with your mask on? 5-10 minutes? 10-20 minutes? More than 20 minutes? More than an hour? And if you are not asleep within about 20 minutes of going to bed, what starts going through your mind?
I'd say 10-20 minutes. I take Valerian root before bed and that seems to put me out nicely
robysue wrote: 3) Are you physically uncomfortable when you wake up in the middle of the night with your mask on? If so, can you describe the kind of discomfort you are experiencing in words? Are you feeling overwhelmed by the air pressure when you wake up in the night?
I am waking up feeling sweaty and hot, and usually right after a vivid dream. I do not feel overwhelmed by air pressure no (I have EPRx3 turned on).
robysue wrote: 4) You are using a pretty low starting ramp pressure of 4cm. How do you feel right at the beginning of the night and every time you turn the machine off and back on? Are you having trouble catching your breath comfortably at the beginning of the ramp period?
No trouble, but I could increase that
robysue wrote: 5) What do you do when you wake up? Do you lie in bed for a very long time fighting to get back to sleep? Or are you able to return to sleep relatively quickly after most of the middle of the night wakes?
Usually I go right back to bed, but sometimes I lie in bed for what feels like a long time fighting to get back to sleep. Usually after 3 small wakeups when I know morning is near I'll get anxious and have trouble getting back to sleep

robot_head
Posts: 9
Joined: Thu Jul 24, 2014 12:05 pm

Re: Should I increase or decrease pressure?

Post by robot_head » Thu Jul 24, 2014 4:14 pm

I was unaware that events are not recorded during ramp, thanks!

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: Should I increase or decrease pressure?

Post by Tatooed Lady » Thu Jul 24, 2014 4:20 pm

when I was dealing with headaches related to APAP treatment, it was due to humidity. Possibly you need to adjust it, either higher or lower.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

robot_head
Posts: 9
Joined: Thu Jul 24, 2014 12:05 pm

Re: Should I increase or decrease pressure?

Post by robot_head » Fri Jul 25, 2014 11:18 am

Woke up again 4 times last night, feel awful today, almost zero obstructive apnea events last night: