Quattro Air mask not venting? Or is it??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
artpenjr
Posts: 20
Joined: Sat Mar 07, 2009 10:33 am
Location: New Jersey Shore

Quattro Air mask not venting? Or is it??

Post by artpenjr » Sat May 17, 2014 8:02 am

I just received my new ResMed "Quattro Air" full face mask. I tried to wear it last night, but when fitting it for leaks I realized it was not venting?

This mask has some new design that is suppose to have a circular vent around the quick-release elbow that holds the tubing. The vent is part of the mask itself, it cannot be removed. I cannot feel hardly any air at all coming out of the vent, even with the mask leak check with high pressure air going to the mask.

My normal masks are the "Mirage Quattro" and the "Ultra Mirage" full face masks. Both of these have much larger, more pronounced vents that I can feel air escaping normally through.

I was afraid to wear the "Quattro Air" last night because I could hardly feel any air at all venting, and felt like I was re-breathing the air I was exhaling.

Does anyone have this new mask? Can you feel any air venting around the front elbow like it should? Thanks in advance for your reply!

Art

User avatar
LSAT
Posts: 13359
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Quattro Air mask not venting? Or is it??

Post by LSAT » Sat May 17, 2014 8:59 am

artpenjr wrote:I just received my new ResMed "Quattro Air" full face mask. I tried to wear it last night, but when fitting it for leaks I realized it was not venting?

This mask has some new design that is suppose to have a circular vent around the quick-release elbow that holds the tubing. The vent is part of the mask itself, it cannot be removed. I cannot feel hardly any air at all coming out of the vent, even with the mask leak check with high pressure air going to the mask.

My normal masks are the "Mirage Quattro" and the "Ultra Mirage" full face masks. Both of these have much larger, more pronounced vents that I can feel air escaping normally through.

I was afraid to wear the "Quattro Air" last night because I could hardly feel any air at all venting, and felt like I was re-breathing the air I was exhaling.

Does anyone have this new mask? Can you feel any air venting around the front elbow like it should? Thanks in advance for your reply!

Art
The advertisment for the Quattro Air says.....Circular Diffused Venting

User avatar
Denial Dave
Posts: 1335
Joined: Wed Sep 19, 2012 6:45 am
Location: Connecticut

Re: Quattro Air mask not venting? Or is it??

Post by Denial Dave » Sat May 17, 2014 9:09 am

I had the same weird sense that it wasnt providing air. It's one of the reason that I quit on it. I could feel the venting but at my high pressure, I felt suffocating. Yes, I had ramp off..

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started at VPAP pressure setting of 20/14.4. I've survived Stage 1 cancer and lost 80+ lbs. Pressure is now 14.5 / 11
if you don't know where you are going... any road will take you there.... George Harrison

User avatar
The Latinist
Posts: 465
Joined: Sat Apr 19, 2014 10:00 pm

Re: Quattro Air mask not venting? Or is it??

Post by The Latinist » Sat May 17, 2014 9:36 am

The Flow Rate curve for the Quattro Air is nearly identical to that of the Mirage Quattro. Any difference you are perceiving is purely psychological because you are used to a mask that blows air all over you. The Quattro Air is just doing exactly what it's supposed to do: dispersing the vented air quietly and diffusely.

Quattro Air:

Image

Source: Quattro Air User Guide

Mirage Quattro:
Image

Source: Mirage Quattro User Guide

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: APAP 12-16 cmH2O, EPR 1. Untreated AHI: 96; treated AHI 2.3.

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Quattro Air mask not venting? Or is it??

Post by library lady » Sat May 17, 2014 10:08 am

I have Quattro Air; the vent does come off, push in on both sides of the larger, grayish vents on either side and the elbow will pop right off - "quick release". Unless you have a faulty vent, I don't see how it can "not vent". Put your pinkie finger or the eraser end of a pencil in and push in from the end that goes onto the hose, and push in the flap inside. If that doesn't move, then it's faulty and you'll need to exchange it for a working mask. That's where the air goes in, and if air can't get in, you won't get any therapy and no air will get in.

If the flap works, then as a test, take a "nap" with the mask on and the unit running ... hold a piece of paper over the vent, about a half inch away in all directions from the elbow. This will give you idea of the air that's coming out of it.

Edit: I can feel air blowing through the venting ... even the hole that is on the mask itself at the 12 o'clock position above the hole the elbow fits into. That is not detachable, but if I put my hand in front of it, I can feel the air coming from it... most of the air is vented through the elbow, and not through the mask itself. Don't worry about the ridge around the vent opening of the mask; I don't believe any air is supposed to come out there... it's just support for the elbow.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

User avatar
The Latinist
Posts: 465
Joined: Sat Apr 19, 2014 10:00 pm

Re: Quattro Air mask not venting? Or is it??

Post by The Latinist » Sat May 17, 2014 10:37 am

library lady wrote:I have Quattro Air; the vent does come off, push in on both sides of the larger, grayish vents on either side and the elbow will pop right off - "quick release".
That actually only removes the elbow. The vent is the circular ring of holes around the connector that the elbow attaches to: it surrounds the elbow, but is not part of it.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: APAP 12-16 cmH2O, EPR 1. Untreated AHI: 96; treated AHI 2.3.

Janknitz
Posts: 8513
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Quattro Air mask not venting? Or is it??

Post by Janknitz » Sat May 17, 2014 10:50 am

It vents fine, it's just very diffuse, soft, and quiet. Very nice, actually. I wore mine for the first time last night. It's a good mask except the headgear caused it to constantly ride up toward my nose and leak at the chin. VERY annoying, and sadly I think that (unless I can jury rig a fix) this mask is toast.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Quattro Air mask not venting? Or is it??

Post by library lady » Sat May 17, 2014 10:56 am

The Latinist wrote:
library lady wrote:I have Quattro Air; the vent does come off, push in on both sides of the larger, grayish vents on either side and the elbow will pop right off - "quick release".
That actually only removes the elbow. The vent is the circular ring of holes around the connector that the elbow attaches to: it surrounds the elbow, but is not part of it.
That ring doesn't have holes in it; I tested them with a pin... nothing gets through them except for the hole above the circle ... the vents are all in the elbow, as you will note if you cover half of them with tape, and try the blower briefly with the mask on as I did. The flap opens and the air goes out there. I don't think this mask would be very effective otherwise, the flap helps "diffuse" the air as needed. Maybe another Quattro Air user can help us out here...

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Quattro Air mask not venting? Or is it??

Post by library lady » Sat May 17, 2014 11:26 am

And what are the holes in the elbow for if not for venting? I realize they are just on the sides of it, but they go in different directions... I feel lots of air blowing on me every night from the elbow up toward my face, but maybe I'm wrong about the circle... perhaps it also vents air; I just don't see how if I can't stick the point of a pin into it.. it just rests on surface, and when I'm washing it water doesn't go through the circle either.

@Janknitz, I had trouble at first with the mask coming up on my chin, but I finally got the headgear adjusted so that it doesn't do that anymore. Couldn't tell you exactly how I did that, it was a lot of trial and error.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

old dude
Posts: 643
Joined: Sat Jul 27, 2013 4:08 pm

Re: Quattro Air mask not venting? Or is it??

Post by old dude » Sat May 17, 2014 2:13 pm

Janknitz wrote:It vents fine, it's just very diffuse, soft, and quiet. Very nice, actually. I wore mine for the first time last night. It's a good mask except the headgear caused it to constantly ride up toward my nose and leak at the chin. VERY annoying, and sadly I think that (unless I can jury rig a fix) this mask is toast.
Janknitz, here is the answer to your problem: https://www.cpap.com/productpage/headge ... sk-fp.html

It was pure serendipity that I discovered this; until I did my mask would also ride up and I'd awake "eating the mask" every night with the bottom of the mask clenched between my teeth. The problem was that the stock top strap in non-adjustable and too small. This headgear turned the Air into the best FFM I've ever used. It's a perfect fit and as a side benefit it's about three times nicer and more durably constructed than the ResMed junky headgear that comes with the mask.

You might want to give it a try.

Janknitz
Posts: 8513
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Quattro Air mask not venting? Or is it??

Post by Janknitz » Sat May 17, 2014 2:30 pm

And what are the holes in the elbow for if not for venting? I realize they are just on the sides of it, but they go in different directions... I feel lots of air blowing on me every night from the elbow up toward my face, but maybe I'm wrong about the circle... perhaps it also vents air; I just don't see how if I can't stick the point of a pin into it.. it just rests on surface, and when I'm washing it water doesn't go through the circle either.
The holes on the elbow are for the anti asphyxiation valve. They let fresh air into the mask if the no air is coming through the hose. A little silicone flap covers those holes when air is coming in the hose.

Thanks, Old Dude, for the video. I'll take a look. My problem is the headgear is too big. But I like everything else about the mask and even invested in a Padacheek liner for it.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
The Latinist
Posts: 465
Joined: Sat Apr 19, 2014 10:00 pm

Re: Quattro Air mask not venting? Or is it??

Post by The Latinist » Sat May 17, 2014 2:34 pm

library lady wrote:And what are the holes in the elbow for if not for venting?
Those are the anti-asphyxia valves, not the vent. Check your user guide: the vent and the valves are clearly marked. And the reason you can't stick a pin through the vent holes is that they do not lead straight through: the air is forced to take an indirect route to the outside so that it will exit in a diffuse way rather than in a stream.
old dude wrote:
Janknitz wrote:Janknitz, here is the answer to your problem: https://www.cpap.com/productpage/headge ... sk-fp.html
Thanks for this. I will definitely give this a try. I have the feeling that the Air could be a great mask if I could just fit it correctly, and I had already come to the conclusion that the headgear is the issue but had no idea what headgear to try in its place.

I am going to trial the Airfit P10 next week, but I intend to keep my Quattro Air as a backup for cold and allergy season even if the P10 works for me.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: APAP 12-16 cmH2O, EPR 1. Untreated AHI: 96; treated AHI 2.3.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Quattro Air mask not venting? Or is it??

Post by chunkyfrog » Sat May 17, 2014 2:42 pm

If it's any help, it took me awhile to get used to the extremely diffused venting from my Airfit P10.
It is hard to feel the airflow; licking a finger helps.
Breathing feels the same as with Swift FX, but is quieter.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

library lady
Posts: 1116
Joined: Wed Feb 05, 2014 11:10 am
Location: Rochester, MN

Re: Quattro Air mask not venting? Or is it??

Post by library lady » Sat May 17, 2014 3:52 pm

The Latinist wrote:
library lady wrote:And what are the holes in the elbow for if not for venting?
Those are the anti-asphyxia valves, not the vent. Check your user guide: the vent and the valves are clearly marked. And the reason you can't stick a pin through the vent holes is that they do not lead straight through: the air is forced to take an indirect route to the outside so that it will exit in a diffuse way rather than in a stream.
OK, my bad... thanks for the explanation.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments:  Sleepyhead
Now using AirFit F10 mask; Quattro Air is backup mask. RemZzzz mask liners with both.

artpenjr
Posts: 20
Joined: Sat Mar 07, 2009 10:33 am
Location: New Jersey Shore

Re: Quattro Air mask not venting? Or is it??

Post by artpenjr » Sat May 17, 2014 4:01 pm

Thank you all for your replies!

I think the steady blast of air from my Quattro Mirage and Ultra Mirage vents has led me to think I need that much air moving through the mask to expel the air I breathe out, so I am not re-breathing all night.

The Quattro Air has so little air coming out of the circular vent, I can't imagine it expelling all the air I breathe out from the mask between each breath. I'm a big guy!

Without a pulse oximeter I doubt if I would ever know if I am re-breathing, other than waking up with a headache maybe!

I just hope this mask isn't defective and the microscopic holes in this vent aren't suppose to be larger. Sounds like you good folks are having the same experience with the quiet and slow air flow though, so it must be normal. Call me paranoid