Is Resmed ASV more feature limited than Philips?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Overflow
Posts: 94
Joined: Tue Apr 15, 2014 5:23 pm

Is Resmed ASV more feature limited than Philips?

Post by Overflow » Wed May 14, 2014 7:13 pm

Looking at some posts on here and slides like this:

http://www.carolinasleepsociety.org/doc ... of_asv.pdf


It seems the resmed machine has less flexibility in its settings, is this enough of a disadvantage to consider the other machine in practice?

Maybe that is just outdated, as RM info states:

http://www.resmed.com/epn/products/s9_v ... c=patients
Building on ResMed's trusted ASV therapy, PaceWave now automatically treats upper airway obstruction with auto-adjusting EPAP (ASVAuto mode), while treating central breathing disorders with auto-adjusting pressure support.
Anyone here used both of the latest machines and has their own preference?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying S1 550p ASV and wisp mask.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Is Resmed ASV more feature limited than Philips?

Post by JDS74 » Wed May 14, 2014 8:03 pm

Sounds as if they are playing catchup without stepping on Respironics patents.
Its likely that the current difference is minor and choice may well hinge on issues of comfort.
If you are happy with the way a Respironics machine "feels", then that is likely how you will feel with their ASV.
OTOH, if ResMed "feels" good to you, then their ASV will likely be comfortable for you.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65129
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is Resmed ASV more feature limited than Philips?

Post by Pugsy » Wed May 14, 2014 8:28 pm

Forum member JohnBFisher has had the occasion to use both brands...ResMed and Respironics ASV machines.
He reports that there is a little difference in how they handle things and how they feel. I bet he will be along later and share his thoughts.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Overflow
Posts: 94
Joined: Tue Apr 15, 2014 5:23 pm

Re: Is Resmed ASV more feature limited than Philips?

Post by Overflow » Thu May 15, 2014 12:53 pm

Ok still on topic but a slight tangent. Interested in these machines as trying to buy one for OSA only. Meeting fierce resistance from sleep clinic and resmed as not diagnosed with CSA.

Today sleep technician was trying to dissuade me from ASV by saying it only targeted up to 90% of something. Looking at the literature could she mean

'Sets MV target to 90% of the patient's own ventilation' from AUtoASV mode? That seems to imply that the machine can takeover up to 90% of patient's own breating acting as a ventilator, but would surely not preclude it from still effectively treating OSA and other apneas up to only 90% efficiency? My theory was that machine does what the others do for OSA but may do it just slightly better as it can adapt to patient's breathing pattern and can greater flexibility with expiratory pressure than EPR. Is this a fundamentally flawed perspective I wonder, the professionals certainly seem to think it is a waste of time (had no titration study with them though).

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Trying S1 550p ASV and wisp mask.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Is Resmed ASV more feature limited than Philips?

Post by JDS74 » Thu May 15, 2014 1:11 pm

If you have OSA only AND you have a sleep study that confirms that, then getting an ASV machine is a very costly way to go.
At most, get an auto bi-level like the Respironics 760 and you will be treated very well.
Probably just an auto CPAP like the 560 will do just fine.
In any event, get one with full data recording so you can see what is going on.
BTW, full dat recording does not mean just hour of use that the least costly machines provide.

The 90% being referred to is the percentage of folks who have sleep apnea and only have obstructive apneas.
The ASV class of machine is for the less than 1% of patients who have central apneas in a significant number.
Getting an ASV machine is throwing away more than $2,000 for no benefit to you.

Just my thought.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65129
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is Resmed ASV more feature limited than Philips?

Post by Pugsy » Thu May 15, 2014 7:24 pm

Overflow wrote:Ok still on topic but a slight tangent. Interested in these machines as trying to buy one for OSA only. Meeting fierce resistance from sleep clinic and resmed as not diagnosed with CSA.
You might find this thread interesting.
viewtopic.php?f=1&t=94430&p=872479#p872479

Mainstream medical is going to be against what you are thinking but there's one doctor I know of who did (and I assume still does) promote using ASV for a high majority of his patients..even UARS if I remember right.
Dr Barry Krakow. Look him up and read some of his stuff. He definitely thinks outside the box...might suit you and might not.

It's a pretty pricey experiment if insurance isn't behind you and they won't unless you have something besides OSA.
ASV machines aren't known for being easy to adjust to but I took to it like a duck to water from the first night.
I bought my machine privately...only way I would/could ever get one since I don't have enough centrals to fool with. Figured I would try it and move on and likely sell it and get my money back. I love the S9 Adapt SV and I have the 36007 model...and someone would have to pry it out of my cold dead hands to get it away from me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Stevoreno_55
Posts: 369
Joined: Tue Jan 14, 2014 10:28 pm
Location: MS Gulf Coast

Re: Is Resmed ASV more feature limited than Philips?

Post by Stevoreno_55 » Thu May 15, 2014 7:56 pm

Pugsy wrote:
Overflow wrote:Ok still on topic but a slight tangent. Interested in these machines as trying to buy one for OSA only. Meeting fierce resistance from sleep clinic and resmed as not diagnosed with CSA.
You might find this thread interesting.
viewtopic.php?f=1&t=94430&p=872479#p872479

Mainstream medical is going to be against what you are thinking but there's one doctor I know of who did (and I assume still does) promote using ASV for a high majority of his patients..even UARS if I remember right.
Dr Barry Krakow. Look him up and read some of his stuff. He definitely thinks outside the box...might suit you and might not.

It's a pretty pricey experiment if insurance isn't behind you and they won't unless you have something besides OSA.
ASV machines aren't known for being easy to adjust to but I took to it like a duck to water from the first night.
I bought my machine privately...only way I would/could ever get one since I don't have enough centrals to fool with. Figured I would try it and move on and likely sell it and get my money back. I love the S9 Adapt SV and I have the 36007 model...and someone would have to pry it out of my cold dead hands to get it away from me.
I have a question; what does ASV stand for?


Stevoreno_55
MS Gulf Coast
05/15/14

_________________
Mask
Additional Comments: Newly diagnosed pressures as of June 05, 2014 are 22cmIPAP and 18cmEPAP; first diagnosed with OSA in 1999.

User avatar
Pugsy
Posts: 65129
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is Resmed ASV more feature limited than Philips?

Post by Pugsy » Thu May 15, 2014 8:06 pm

Stevoreno_55 wrote:I have a question; what does ASV stand for?
I have a question; what does ASV stand for?

Adaptive Servo Ventilation...a non invasive ventilator type of machine that will push a rapid and strong burst of air in response to a central...sort of like a ventilator that you see on the movies but not exactly like it.


Hey Overflow...
Oh lookie what I found from Dr Krakow...he talks about ASV...just stumble on it while getting correct spelling for something.
https://www.youtube.com/watch?v=7fMki3KFRgg

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Stevoreno_55
Posts: 369
Joined: Tue Jan 14, 2014 10:28 pm
Location: MS Gulf Coast

Re: Is Resmed ASV more feature limited than Philips?

Post by Stevoreno_55 » Thu May 15, 2014 9:17 pm

Pugsy wrote:
Stevoreno_55 wrote:I have a question; what does ASV stand for?
I have a question; what does ASV stand for?

Adaptive Servo Ventilation...a non invasive ventilator type of machine that will push a rapid and strong burst of air in response to a central...sort of like a ventilator that you see on the movies but not exactly like it.


Hey Overflow...
Oh lookie what I found from Dr Krakow...he talks about ASV...just stumble on it while getting correct spelling for something.
https://www.youtube.com/watch?v=7fMki3KFRgg
So ASV is not the same as CPAP? Just curious.


Stevoreno_55
MS Gulf Coast
05/15/14

_________________
Mask
Additional Comments: Newly diagnosed pressures as of June 05, 2014 are 22cmIPAP and 18cmEPAP; first diagnosed with OSA in 1999.

User avatar
Pugsy
Posts: 65129
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is Resmed ASV more feature limited than Philips?

Post by Pugsy » Thu May 15, 2014 9:49 pm

Stevoreno_55 wrote:So ASV is not the same as CPAP? Just curious.
Yes and no

Think of it as a really fancy high dollar cpap (actually it is a high end bilevel pressure) machine that treats obstructive apnea and central apnea. So it can be used for plain OSA but most often it isn't RX'd except for people who either have only centrals or a combination of centrals and obstructive because they are so darned expensive.
CPAPs themselves...only use a single fixed pressure and can't respond to centrals.
ASV machines have some features that allow for rapid and large increases in pressure which are needed when a person has a central apnea. They will actually breathe for you for a period of time until your own body begins breathing like it should.
This is NOT the bilevel pressure machine that I have told you about for your high pressure needs. That bilevel machine doesn't do the fancy stuff the ASV bilevels do.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Stevoreno_55
Posts: 369
Joined: Tue Jan 14, 2014 10:28 pm
Location: MS Gulf Coast

Re: Is Resmed ASV more feature limited than Philips?

Post by Stevoreno_55 » Fri May 16, 2014 5:35 am

Pugsy wrote:
Stevoreno_55 wrote:So ASV is not the same as CPAP? Just curious.
Yes and no

Think of it as a really fancy high dollar cpap (actually it is a high end bilevel pressure) machine that treats obstructive apnea and central apnea. So it can be used for plain OSA but most often it isn't RX'd except for people who either have only centrals or a combination of centrals and obstructive because they are so darned expensive.
CPAPs themselves...only use a single fixed pressure and can't respond to centrals.
ASV machines have some features that allow for rapid and large increases in pressure which are needed when a person has a central apnea. They will actually breathe for you for a period of time until your own body begins breathing like it should.
This is NOT the bilevel pressure machine that I have told you about for your high pressure needs. That bilevel machine doesn't do the fancy stuff the ASV bilevels do.
I don't know if I have centrals; would that information be listed on my last sleep study report from 2007? Are centrals serious like my severe OSA? If my pressure increases after I have my next study which starts next week; let's say it goes from 18cm to 20cm will I more than likely remain on a straight CPAP or have to move to a BIPAP? The RT with my current DME has been letting me use a one year old ResMed S9 Elite CPAP machine and the RT has recommended an "autoset" machine. Are "autoset" machines different from straight CPAP machines and does ResMed make an S9 Elite "autoset" or is a ResMed S9 Elite CPAP machine different from a ResMed S9 "autoset" machine? Your feedback is always appreciated.


Stevoreno_55
MS Gulf Coast
05/16/14

_________________
Mask
Additional Comments: Newly diagnosed pressures as of June 05, 2014 are 22cmIPAP and 18cmEPAP; first diagnosed with OSA in 1999.

User avatar
Pugsy
Posts: 65129
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Is Resmed ASV more feature limited than Philips?

Post by Pugsy » Fri May 16, 2014 5:45 am

Stevoreno_55 wrote:I don't know if I have centrals; would that information be listed on my last sleep study report from 2007? Are centrals serious like my severe OSA? If my pressure increases after I have my next study which starts next week; let's say it goes from 18cm to 20cm will I more than likely remain on a straight CPAP or have to move to a BIPAP? The RT with my current DME has been letting me use a one year old ResMed S9 Elite CPAP machine and the RT has recommended an "autoset" machine. Are "autoset" machines different from straight CPAP machines and does ResMed make an S9 Elite "autoset" or is a ResMed S9 Elite CPAP machine different from a ResMed S9 "autoset" machine? Your feedback is always appreciated.
Centrals should have been documented on your old sleep study...and they will be documented on the new upcoming sleep study.
Yes...centrals can be severe...it all depends on how many, how long they last and do they disrupt sleep...lots of things to consider when deciding if centrals warrant special treatment.

I think you should be on bipap now with 18 cm...for sure if you move to 20 cm since the cpap machine has no where to go in case you need more and lots of people need more than 20.

The S9 AutoSet works just like the S9 Elite except it has a mode of operation that allows for auto adjusting pressures (it also has cpap mode which is what the Elite does) and that auto adjusting (APAP) mode might help you use less pressure some of the night and only increase to 18 or 20 if needed. That's the only difference between the S9 Elite and the S9 Autoset....that APAP mode which might be of great benefit to you if they try to stick you with cpap/apap single pressures of 18 to 20 again instead of going to bilevel pressure machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Is Resmed ASV more feature limited than Philips?

Post by robysue » Fri May 16, 2014 8:26 am

I think the figure 90% is being thrown around in multiple contexts in this thread and it's time to clarify what might be meant by that number.

First:
Overflow wrote: Today sleep technician was trying to dissuade me from ASV by saying it only targeted up to 90% of something. Looking at the literature could she mean

'Sets MV target to 90% of the patient's own ventilation' from AUtoASV mode?
First a definition: MV= minute ventilation is the total amount of air inhaled over a one minute period.

So this 90% is a machine setting on ASV machines: It tells the ASV when to trigger (initiate) breaths by drastically ramping up the IPAP because the patient is not breathing sufficiently well for central reasons, not obstructive ones. Part of the problem in central apnea (which ASV is designed to treat) is a CO2 overshoot/undershoot cycle. During the "overshoot" part of the cycle, too much CO2 is blown off and that reduces the respiratory drive and the breaths become shallower and shallower. When the minute ventilation recorded by the machine drops to 90% (or less) of the so called "target minute ventilation", the machine starts triggering breaths by increasing the IPAP and switching between EPAP and IPAP at fixed time intervals. The idea behind this is:
  • the drop in minute ventilation is the early heads up that a CO2 overshoot has likely happened and that the breathing is becoming unstable in central fashion; hence the patient may have entered the CO2 undershoot part of the cycle.
  • the machine triggered breaths should prevent the CO2 overshoot/undershoot problem spiraling out of control leading to CAs at the nadir of each breathing cycle.
And ASV machine is just as effective as a regular CPAP/APAP/BiPAP/BiPAP Auto when it comes to treating the obstructive events in OSA: If the ASV's EPAP pressure is correctly titrated, it should control the OSA as well as a well titrated CPAP/APAP/BiPAP/BiPAP Auto. If the ASV's EPAP pressure is too low, too many obstructive events can get by the defenses---just like what happens if the pressure of a CPAP (or APAP/BiPAP/BiPAP Auto) is set to low.
Overflow wrote:That seems to imply that the machine can takeover up to 90% of patient's own breating acting as a ventilator,
No. You are misinterpreting stuff here. The ASV machine does not (usually) take over "90% of the patient's own breathing." The "90%" number tells the ASV when to start acting as a ventilator: If the person's minute ventilation drops below 90% of the target ventilation number, the ventilator part of the ASV kicks in and the machine starts triggering breaths. How much of the time the ASV is triggering breaths and how much time the patient is triggering their own breathing varies quite a bit from patient to patient. But in the ASV reports I've seen posted here, the patient triggered breaths percentage is usually in the mid to high 80s or low 90s. In other words, it's rare for a properly titrated ASV machine to need to trigger more than 20% of the breaths.
Overflow wrote:but would surely not preclude it from still effectively treating OSA and other apneas up to only 90% efficiency? My theory was that machine does what the others do for OSA but may do it just slightly better as it can adapt to patient's breathing pattern and can greater flexibility with expiratory pressure than EPR.
Again no.
If the ASV's EPAP pressure is correctly titrated then the ASV should be 100% effective in treating OSA, same as a well titrated CPAP.

But the "adapting to the patient's breathing pattern" is tied to how the machine treats the person's Central Apnea. Part of the ASV algorithm involves allowing the machine to determine a moving MV target value based on the last several minutes of normal breathing instead of using a predetermined fixed MV target. That helps prevent the ventilator-style part of the algorithm from kicking it too often.

As for "greather flexibility with expiratory pressure than EPR", the answer is "maybe, maybe not." On an ASV machine there is a (often fixed) EPAP pressure setting, a min IPAP (or PS) setting, and a max IPAP (or PS) setting. PS and IPAP are intimately tied to each other: IPAP = EPAP + PS, so min IPAP = EPAP + min PS and max IPAP = EPAP + max PS.

When the breathing is stable (from a CENTRAL point of view), the ASV acts like a bilevel machine (an S9 VPAP or a System One BiPAP) where the pressure cycles back and forth between EPAP and min IPAP = EPAP + min PS following the patient's exhalations and inhalations. So if PS is set to 4 or more, then yes, there will be more expiratory relief than an S9 Elite/AutoSet with EPR = 3. But if PS is 0, 1, 2, or 3, there will NOT be any more expiratory relief than an S9 Elite/AutoSet with EPR. In any case, the expiratory relief achieved by the PS setting on the ASV can be obtained on the MUCH cheaper bilevel machines.

So if additional expiratory relief is the primary issue rather than central apnea, the machine of choice would be a plain old bilevel machine, like the the S9 VPAP S, the S9 VPAP Auto, the System One BiPAP Pro, and the System One BiPAP Auto.

We also have
JDS74 wrote:The 90% being referred to is the percentage of folks who have sleep apnea and only have obstructive apneas.
The ASV class of machine is for the less than 1% of patients who have central apneas in a significant number.
The usually cited numbers are that about 10-15% of newly diagnosed OSA patients who are put on CPAP will develop clinically significant problems with pressure induced central apneas and may meet the clinical definition of Complex Sleep Apnea. For some of these patients, it may be possible to find a sweet spot for the pressure settings (possibly through APAP or BiPAP or BiPAP Auto) where the pressure used is high enough to decently manage the OSA but low enough to keep the central part of CompSA under control. But for many (possibly most) of the people in this unlucky group of PAPers with CompSA, a good compromise pressure cannot be found, and these folks really need a well tuned ASV machine. The EPAP controls the OSA and the ASV algorithm controls the central part of the CompSA.

Finally I have to agree with JDS74: If you are not having any problems with central apneas, it seems like an ASV would be an awfully expensive way of addressing your problems. If the primary problem with CPAP/APAP is that you're using pressure(s) near the top of the range and/or EPR = 3 is not providing enough expiratory pressure relief, then the appropriate machines to be looking at are the bilevels. On the Resmed side, that means the S9 VPAP S and the S9 VPAP Auto.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Is Resmed ASV more feature limited than Philips?

Post by archangle » Sat May 17, 2014 3:51 am

Overflow wrote:Ok still on topic but a slight tangent. Interested in these machines as trying to buy one for OSA only. Meeting fierce resistance from sleep clinic and resmed as not diagnosed with CSA.

Today sleep technician was trying to dissuade me from ASV by saying it only targeted up to 90% of something. Looking at the literature could she mean

'Sets MV target to 90% of the patient's own ventilation' from AUtoASV mode? That seems to imply that the machine can takeover up to 90% of patient's own breating acting as a ventilator, but would surely not preclude it from still effectively treating OSA and other apneas up to only 90% efficiency? My theory was that machine does what the others do for OSA but may do it just slightly better as it can adapt to patient's breathing pattern and can greater flexibility with expiratory pressure than EPR. Is this a fundamentally flawed perspective I wonder, the professionals certainly seem to think it is a waste of time (had no titration study with them though).
Many ASV users find it to be uncomfortable to live with, at least at first, because it wants you to breathe its way, not your way.

You might think it wouldn't do "ASV things" unless you have central apnea, but I'm not sure about that. It might actually give you poorer results than a manual CPAP.

Most ASV machines can be set to do manual CPAP if necessary, but that's a big waste of money.

Unless you have central apnea, an ASV machine won't necessarily give you better therapy than a manual CPAP. An Auto CPAP might actually give you better therapy than an ASV machine if you don't need ASV.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Is Resmed ASV more feature limited than Philips?

Post by Sir NoddinOff » Sat May 17, 2014 12:05 pm

I have both machines. I think most of your questions got answered quite capably, I would only add that I happen to really love the the PR autoSV 'rise time' feature which can be found in the clinician's menu (rise time is present in both the 950s and 960s). It allows one to set the inhale time period a bit longer, which is nice if your nose is occasionally stuffed up or maybe you have a deviated septum resulting in longer inhale requirements... this feature seems it gives you an adjustable second or two longer to suck in the needed air into your lungs and doesn't cut you off like the ResMed models. That's just my experience though I also like the bigger humidifier tank in the PR, plus the optional 'classic mode' humidification feature which REALLY cranks up the humidity for us occasional mouth breathers (a life saver in my case). Here's more on 'Classic Mode':

viewtopic/t86605/Philips-Respironics-Sy ... stion.html

viewtopic/t86835/Humidifier-questions-f ... -user.html

viewtopic/t97055/PR-Series-60-Humidifier-Question.html

Final note: I it have from a reliable member that you can run any PR autoSV in straight CPAP, BiPAP and ASV, something you can't do with any one ResMed Adapt.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.