Sleepy Head Data Help, better crop

Post by Upalot » Sat Apr 26, 2014 9:37 pm

I should have listed my stuff. My CPAP is a Resmed R9 which is an auto usually leave it in auto but I think the surge of air wakes me sometimes. In fact probably more than sometimes. My humidifier is the Hfi, my mask is the Mirage Quattro. It’s a full face mask. I am typically a mouth breather, Always have been. Even when I was younger and skinny I snored. I have had sleep issues for probably about 30 years of so. I finally got to see a sleep doc about 2 1/2 years ago. I decided to finally get some help. My wife was telling me I stopped breathing all the time. I took an at home sleep study which showed my breathing was stopping about 65 times an hour. I know that is pretty bad. I don’t ever remember sleeping through the night in my adult life. I used to get up about 2-3 or 3-4 times. But now it’s more like 6-8. I am overweight and I smoke. I am trying to quit smoking. I know it will kill me. It’s been very hard, right now I am trying a vapor pin, and it gives you nicotine but not the problems from the smoke. I am hoping to stop my smoking all together using it. I admit the smoking has been the most addictive thing I have ever done. It’s horrible; I own a business and have a lot of stress so that does not help my sleep either. It has progressively gotten worse over the past few years. Everyone else talks about the benefit of CPAP treatment. I cannot really comment on it, it has never really made me feel better. Just different levels of shitty.

I reattached the sleepy head graphs only showing the ones Pugsly recommended and the side bar showing the AHI and other information. I would greatly appreciate any help I can get. If my AHI gets better will I wake up feeling rested? Do all you folks feel good in the morning? I never do, and half the time right now I sleep some crazy hours, way too long. I think I am starting to become depressed with all this and that is affecting me also. It’s been terribly stressful just trying to get some rest at night. I always wake up tired and do not feel like I have much energy. I hate it. To make matters much worse I am a type a guy and all this has really affected me from a mental health standpoint.

Any help or advice would be awesome. Thanks a bunch

Image

Image

_________________
Humidifier

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Sleepy Head Data Help, better crop

Post by kaiasgram » Sun Apr 27, 2014 1:51 am

Upalot,

I've colored in the acceptable leak level again on your leak graph so you can see how much of the time you're leaking excessively (above 24). Leaks are the first thing to work on. You may need a different mask, or a better fit or size of your Mirage Quattro, and/or the mask cushion needs replacement. You might also need to look into mask liners.
Image

If you haven't seen it yet, please check out "Taming the Mirage Quattro" by forum member Janknitz : http://maskarrayed.wordpress.com/taming ... e-quattro/

For sure we want to get your AHI down below 5 which may involve adjusting your pressure settings, but not yet -- leaks have to get under control first. Once leaks go above that 24 line, the AHI reporting is not necessarily accurate -- the machine basically gets confused about what's going on (I suspect that's why you have so many "Unspecified Apnea" events flagged -- the machine knows something is happening but it can't figure out what).

You may well be sensitive to the pressure changes -- lots of us don't do well with the ups and downs and sometimes wide swings in pressure. When the leak issues get fixed we can then take a closer look at the pressures and see what might help there. Right now too much therapy air is escaping from the leaky mask.

Not everyone feels good even after a considerable period of time on cpap (I speak from firsthand experience here), however in your case we can see what the problem(s) may be, so there's room to improve and every reason to be hopeful. Right now you aren't getting effective therapy, so no wonder you're not sleeping well and not feeling any benefit. But there's a lot you can do, you just have to do it in a stepwise fashion.

So, your mask: How old is it? Have you tried other masks in the past and if so, what issues did you have with them? Where does the mask leak the worst? Have you tried a mask liner or anti-leak strap (leak strap seen here: http://www.padacheek.com/PAC_AntiLeak_Strap.html)?

As we start to work on these issues, it's a good idea to keep posting in the same thread rather than starting a new one. That way all your info and background is available in the same place.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

Upalot
Posts: 15
Joined: Fri Apr 25, 2014 9:18 pm

Re: Sleepy Head Data Help, better crop

Post by Upalot » Sun Apr 27, 2014 7:20 pm

thanks a lot I will work on the leaks first. And see where it goes. I will also try and post this in the first thread again. I did not have much luck the first time, I could not get the graphs in. I will try again.

_________________
Humidifier

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Sleepy Head Data Help, better crop

Post by kaiasgram » Sun Apr 27, 2014 8:05 pm

Upalot wrote:thanks a lot I will work on the leaks first. And see where it goes. I will also try and post this in the first thread again. I did not have much luck the first time, I could not get the graphs in. I will try again.
I would just let that first thread go and stick with this one. It's more recent and the graphs are working fine here.

Keep us posted about the mask and leaks. I think you have a good chance of getting your therapy dialed in and starting to feel much better.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepy Head Data Help, better crop

Post by Pugsy » Sun Apr 27, 2014 8:08 pm

What is causing the too frequent wake ups (shown by the breaks in therapy line)? Any idea?
That many wake ups all by themselves will make a person feel like crap even if the reports didn't show bad leaks and/or too many events.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Sleepy Head Data Help, better crop

Post by robysue » Sun Apr 27, 2014 10:03 pm

Upalot,

I posted this in the other thread as well. Not sure which one you plan on reading so I decided to repeat my long response here as well.

Your leaks need to be worked on. On the graph that Kaisgram posted, she highlighted the "Large Leak"; my eyes say that your leaks are above that line for significant periods of time. It looks like there are three very significant periods of long, large leaks, each of which is between 45 and 120 minutes long. My eyes say that there's about 3 1/2 to 4 hours of very large leaks. My guess is that you see a lot of Mr. Red Frowny on the Sleep Quality Report on the S9's LCD each morning.

The data that you posted in the last post indicates that the S9 scored a significant number of unspecified apneas; these are scored when the machine cannot tell whether the apnea is an OA or a CA and the most common reason why the machine can't tell is that the leaks are too large.

Another thing that I notice is that during the periods of large leaks, the machine is not responding to apneas, most likely because many of them are scored as "unknown apneas." But it could also be that the machine is trying to NOT cause the leaking to get even worse.

So if this leak data is typical, it could be that your AHI is still too high because of one or both of the following:

1) Loss of therapeutic pressure when the leaks are (well) above 24 L/min. If the machine is having trouble maintaining pressure, you don't get the full benefit and your airway is more prone to collapse.

2) Your machine is having trouble responding appropriately to your events. Two things can contribute to this. First, the large leaks lead to many of the apneas being classified as "unknown" rather than OA, and the machine does not respond to "unknown" apneas. Second, the minimum pressure setting might not be quite high enough. Compounding the min pressure problem is that the machine is reluctant to increase the pressure further once the apneas start during the large leak periods.

Working on the leaks may go a long way towards fixing the AHI problem.
Upalot wrote:I should have listed my stuff. My CPAP is a Resmed R9 which is an auto usually leave it in auto but I think the surge of air wakes me sometimes. In fact probably more than sometimes. My humidifier is the Hfi, my mask is the Mirage Quattro.
Judging from your data, I suspect that the "surges of air" that wake you are very large leaks and NOT pressure increases.

So if you can get the leaks under control, then you might just wake up a lot less frequently.

my mask is the Mirage Quattro. It’s a full face mask. I am typically a mouth breather, Always have been.
Do you mouth breathe in the daytime?

If you are indeed a dedicated mouth breather, then I can understand why you are using a FFM. But you still are experiencing some excessively large leaks that are affecting the efficacy of your therapy. I would suggest that you do one or more of the following:

A) Review the fitting guides for the Quattro carefully. Also review JanKnitz's Taming the Mirage Quattro fitting tips very carefully. It's a common error to over tighten the headgear and straps in an effort to keep this mask from leaking. But that prevents the mask's air cushion from fulling inflating, which in turn makes it more likely to start leaking once the machine starts to increase pressure.

B) Fit the mask at your full pressure and NOT the beginning ramp pressure. Masks are harder to seal at higher pressures, and since you need some pretty significant pressure at times, you need to fit the mask at those higher pressures.

C) Consider using mask liners, a chinstrap, or an anti-leak strap. The large leaks might be triggered by changes in your facial features once you are lying down and asleep. These may help the mask maintain its seal when those facial changes occur.

D) Consider an alternate size mask cushion and/or an alternate full face mask. If the Mirage Quattro is the only mask you've ever tried, it may be time to try another mask.
I used to get up about 2-3 or 3-4 times. But now it’s more like 6-8.
Fragmented sleep is adding to your problems. Of course you don't need to be told that. Question is: How to fix the fragmented sleep. And how to do that depends on what's triggering the wakes. Since you think that surges of air are a factor, there's some reason to be optimistic that if you can get the leaks better under control, that the number of wakes will start to drop.
I am overweight and I smoke. I am trying to quit smoking. I know it will kill me. It’s been very hard, right now I am trying a vapor pin, and it gives you nicotine but not the problems from the smoke. I am hoping to stop my smoking all together using it. I admit the smoking has been the most addictive thing I have ever done. It’s horrible;
Having lost my mother, my brother, and two uncles to smoking related lung cancer, I do wish you the best of luck in kicking the smoking habit. It is indeed insidiously additive and very, very hard for most to quit. However once you are finally off the smokes that too may help you sleep better.
I own a business and have a lot of stress so that does not help my sleep either. It has progressively gotten worse over the past few years.
Stress does aggravate sleep problems. And the combination of having never really gotten your PAP therapy truly optimized and work related stress easily explains why you're not feeling any better.
Everyone else talks about the benefit of CPAP treatment. I cannot really comment on it, it has never really made me feel better. Just different levels of shitty.
You've never really gotten the CPAP therapy genuinely optimized since you say your long term AHI is still about 10 on CPAP. And your leaks are significant. So the CPAP isn't doing its job as well as it should be (probably because of the leaks) AND the leaks are most likely waking you up.

So again, it comes back to Fix the leaks FIRST and see how much that does to address the AHI problem and the too many wakes during the night problem.

I would greatly appreciate any help I can get. If my AHI gets better will I wake up feeling rested? Do all you folks feel good in the morning? I never do, and half the time right now I sleep some crazy hours, way too long. I think I am starting to become depressed with all this and that is affecting me also. It’s been terribly stressful just trying to get some rest at night. I always wake up tired and do not feel like I have much energy. I hate it. To make matters much worse I am a type a guy and all this has really affected me from a mental health standpoint.
It's hard to tell exactly how much better you will feel if the AHI gets better. Once the AHI is consistently below 5, some people start feeling better almost immediately. Others? It takes time before they notice much of a difference. It all depends on how much else is affecting the person's sleep AND how many other health problems the person is dealing with.

Think of it this way:

Good sleep is necessary for feeling good in the day time with enough energy to do what you want to get done every day. But good sleep by itself may not be sufficient to guarantee that you'll feel good during the day with enough energy to do what you want to get done every day. Other health conditions and STRESS can wipe you out even if you get good sleep night after night. So you'll need to deal with anything else that is causing daytime fatigue as well.

And optimal CPAP therapy is necessary for good sleep since you have severe sleep apnea. Once your CPAP therapy is truly optimized (the leaks are under control and the AHI is consistently less than 5), then your sleep will no longer be constantly interrupted by too many apneas and too many Large Leaks. But optimal CPAP therapy by itself may not be sufficient to guarantee that your sleep will be good. If you continue to have problems with fragmented sleep once the leaks and AHI are under control, your sleep may still feel bad and fixing that bad sleep will require more than just fixing the CPAP therapy.

So getting your CPAP therapy optimized is necessary if you want to have a chance of fixing the sleep problems and feeling better in the daytime. And getting your CPAP therapy optimized means fixing the leaks and getting the AHI down to where it needs to be.

Now to focus on just a couple of other things more carefully:
and half the time right now I sleep some crazy hours, way too long. I think I am starting to become depressed with all this and that is affecting me also. It’s been terribly stressful just trying to get some rest at night.
Sleeping crazy hours suggests that you may have some additional sleep issues beyond the inadequately treated OSA.

How would you describe your sleep hygiene? Cleaning up the sleep hygiene may help the sleeping crazy hours problem. And a more normal sleep schedule may be critical in making you feel more rested in the morning after you do get the CPAP therapy optimized.

And stressing out about sleep often adds significantly to the problems of both getting to sleep and staying asleep. Which in the long run makes you feel worse not better. But it's hard not to "worry" when it seems as if everything is going to hell in a handbasket when you're lying in bed trying desperately trying to get to sleep.

So you may need to work on trying to destress about the sleep. (And I say that as someone who has had to spend a large part of the last 3 years learning how to NOT stress out about my own sleep.) It helps to have a nice relaxing bedtime routine. And it's critical to learn how to draw some closure to the day.
To make matters much worse I am a type a guy and all this has really affected me from a mental health standpoint.
Sounds like the combination of bad sleep and stressing about the bad sleep and the Type A personality are all causing a lot of additional stress in your daily life and that it's really taking a toll on you.

What kind of things do you do when you want to relax? You need to take some time each day for conscious relaxation and destressing. You also need to work on making bedtime as unstressful as possible. If you don't have a relaxing bedtime that allows you to put the day behind you AND does not cause you to start focusing on your sleep problems. (And yes, I know that's hard too.)

Best of luck

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

Upalot
Posts: 15
Joined: Fri Apr 25, 2014 9:18 pm

Re: Sleepy Head Data Help, better crop

Post by Upalot » Mon Apr 28, 2014 2:49 pm

You are all very kind to take so much time to reply to my problems. I will work on my leaks immediatly. Do you think my preasure it too high. I do think the preasure wakes me up. Beucase you say that I remember sometime waking up and reaching over and turning the machine off then back on to reset the preasure to the minimum and not the high preasure that cuased me to waken. Thanks agian.

_________________
Humidifier

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepy Head Data Help, better crop

Post by Pugsy » Mon Apr 28, 2014 3:03 pm

It's impossible to accurately evaluate pressure needs in the presence of so much leak.
You do appear to spend a lot of time up in the higher range though which could of course be a disruptive factor to your overall sleep.

From the looks of things (assuming leaks were NOT a factor) it appears you might need a higher minimum pressure which might help reduce the need for the near maximum to come into play as often.

What could be tried... and it is far from an ideal solution ...would be to limit the maximum (that isn't doing your leaks any favors) and see if it helps with the wake ups and maybe work up to a more ideal pressure if it was needed. Some therapy is better than no therapy and your fragmented sleep is totally killing your sleep quality...so maybe some sleep consolidation is better than so many wake ups and totally messed up sleep quality. I stress that this may not be ideal if someone reigns in that maximum if you are indeed needing it to hold the airway open.
So it would be undertaken with the idea of getting leaks and wake ups under more control with the understanding that some OAs might well slip past the defenses and an increase in pressure could very well be needed once leaks and wake ups (hopefully) are better managed.
Of course if we allow some OAs to slip past the defenses then one runs the risk of the OAs causing wakeups or arousals...so damned if you do and damned if you don't sometimes.

Something like 12 minimum and 14 maximum maybe...just to see what happens with leaks and wakeups. I don't think 12 minimum will do a great job but you might get lucky in that if the leaks are better managed then those higher pressures wouldn't be needed.
It's so hard to say with any degree of certainty what might or might not work when faced with leaks in the number and duration that you are experiencing. But hey...worth a try because right now things are way in the toilet as it is.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Sleepy Head Data Help, better crop

Post by kaiasgram » Mon Apr 28, 2014 3:29 pm

Upalot, hi -- I just finished typing up a reply and I see Pugsy covered what I was going to say so I won't repeat.

But because you are concerned about the impact of higher pressure on your sleep, I would just say for down the road -- IF it turns out that you do in fact need a very high pressure to treat your apnea once leaks are controlled, you might find you do better with a bilevel machine (in the ResMed lineup it's called VPAP). Bilevel allows you to set separate inhale and exhale pressures, and folks who require high pressures often sleep more comfortably with these machines. Currently you are getting a little exhale relief (that's your EPR x 3), but with bilevel you can have a much greater difference between inhale and exhale pressures. I bring this up only to reassure you that there are still a lot of possibilities for you to be able to sleep more comfortably and have your apnea successfully treated. Stay with us.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Sleepy Head Data Help, better crop

Post by robysue » Mon Apr 28, 2014 3:35 pm

Upalot wrote:You are all very kind to take so much time to reply to my problems. I will work on my leaks immediatly. Do you think my preasure it too high.
As pugsy says, it's impossible to tell until you get the leaks under control.
I do think the preasure wakes me up. Beucase you say that I remember sometime waking up and reaching over and turning the machine off then back on to reset the preasure to the minimum and not the high preasure that cuased me to waken. Thanks agian.
There's probably some psychology at work here. You wake up (for whatever reason) and then notice the pressure (which may very well be higher than it was when you first went to bed). So you turn the machine OFF and back ON because that's the easiest way to know for sure that the pressure is now down where you know you can get to sleep.

There's nothing wrong with doing this. I do it myself all the time. But at this point I often look at my data the next morning and I see wakes (i.e. OFF/ON cycles) that I have no memories of because the total time I was awake was so short. (It typically takes a wake of 5 minutes or more to actually be remembered in the morning.) I've learned not to worry too much about these wakes I don't remember. If there aren't too many of them and I don't remember them, they are NOT overly disruptive to my overall sleep quality.

Another thing I notice in my own data is this phenomenon: I'll sometimes remember turning the machine OFF and ON (usually because I'm worried about aerophagia) and the data shows that I was actually at minimum pressure when I turned the machine OFF and back ON. It's psychological: Before I turn the machine off, I'm aware of the pressure and I'm worried it might be at or close to its max and so I turn the machine OFF and ON so that I can reassure myself that the pressure is back down where I know it won't trigger aerophagia before I get back to sleep, and consequently I have no trouble getting to sleep. By turning the machine OFF and ON, it gives me one less thing to worry about when I'm going back to sleep.

It's actually pretty critical when you are dealing with multiple wakes of unknown cause to consider carefully your own reaction to the wake. Stressing out about finding yourself awake in the middle of the night and worrying about the fact that you are awake is a good way of making the wakes last long enough for you to remember the next morning. And it may also cause you to continue to sleep rather lightly and hence notice the next wake more than you should.

But the less you stress out about it, the more likely you are to get back to sleep quickly, and (in the long run) the fewer middle of the night wakes you will have.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

Upalot
Posts: 15
Joined: Fri Apr 25, 2014 9:18 pm

Re: Sleepy Head Data Help, better crop

Post by Upalot » Tue Apr 29, 2014 9:18 pm

I adjusted my mask and it seems to be helping a lot. I will post the results when I have a few days data. Do you guys change your own pressure? I do not know how, my sleep doc has always done that. But I have wondered if i could change it myself to see how things go. Or is that a bad idea. I have also thought about using straight CPAP mode and not the auto. I think the auto ramps up and blows my mask off. Does that happen to anyone else?

I have not found any way to keep my mask on at the higher preaures period. I don't have a clue how anyone can keep their mask on at 18-20 its like a wind storm.

My mental state of mind is much better now that i have received so much support. Its always nice to know you are not alone in your struggles. Because to be honest sometimes I think"Damn why has God done this to me?" When I should be thinking " thanks God for allowing me to find out about my OSA and getting me some help with CPAP, so I will not die. Because that is exactly what would have happened had I not gotten on CPAP". So no matter how bad you think it might be, God is there and other people are there to support you, thats why everyone stepping up and taking their own private time to write such detailed answers to my questions makes me feal all warm and cuddly inside. I mean it thanks to all. I will keep trying and try to stay more positive and work through my issues.

My mooring after the mask adjustments was much better, still not great but where I was. Coming off the bottom is easy. I must have adjusted my mask, to allow the nose adjustment on my Mirage Quattro mask all the way out on the nose adjustment know on your forehead to allow me more room for my big snoot. But in the process of doing that it weakened the masks top seal and was coming open fairly easily. I had not noticed it until someone on this board mentioned I need to look at a mask refit. I thought heck before I do that I will cover the basics myself. The nose piece was not fitting me properly. I changed it and now it touches my nose a bit and gives me a little clostro but that will pass with time. The mask fits alow better.

I do not know what a leak liner is. Is that something that goes with your mask liner. I also think my mask liner breaks down way before my DME sends me out a new kit in the mail. I have oily skin and it seems to wear on the mask especially hard. I also struggle greatly with the strap. I have a huge hog head and i can never seem to get the strap to adjust so that when my head is tilted forward the strap does not dig into the back of my neck and wake me up. Also why are all the resmed mask straps so darn skinny in the back. Do others have trouble with the mask straps digging into their heads. Oh yes I shave my head daily, forgot to say that. So I do not have any hair back there to help cushion the strap. Its skinny strap on bare bald head.

I am finally recognizing the fact that I will have to work through every one of these little issues so I can sleep better. I would get so frustrated with having to deal with this stuff at bed time. I would just want to go to sleep. But I am getting better at that and realizing that if I handle it better I will not be so grouchy in the morning.I also do it away from bedtime so I am not so messed up in the head from sleep aids.

That is my last question. I do not like the sleep aids at all. I take silenor 6mg. And break an ambian in half or 5mg or ambian each night. The issue is neither seem to keep me asleep. I still wake up, but I feel some of that is having to pee at least twice a night, and the darn pressure waking me up. So its a two part question. I do not feel my body "wakes up" from those drugs the way they are advertised and the way my sleep doc talks. I think it takes say more like 10-11 hours for my body to wake up. I am physically away after about 8 1/2 but my brain is not firing properly until at least 10-11 hours after taking the sleep aids. Does anyone else have these problems. and does anyone take or have any success taking something to help you not pee. Aside form drinking less liquid near bedtime which i have started to monitor more closely. I am a liquid head and figure liquid in liquid out. So I will need to drink less in the evening.

Thanks agin for everyones support. I feel bad always asking and not being able to hep more. Maybe someone will have a woodworking question I can failed, make me feel more needed:)

_________________
Humidifier

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Sleepy Head Data Help, better crop

Post by kaiasgram » Tue Apr 29, 2014 9:39 pm

I think you are already helping by posting all your questions -- because other newbies will come along and see that they are not the only ones struggling. And if you stick around you will soon be in a position to "pay it forward." That's what keeps this forum going.

The anti-leak strap I mentioned is not the same as a mask liner. You can see what it looks like here:
http://www.padacheek.com/PAC_AntiLeak_Strap.html

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepy Head Data Help, better crop

Post by Pugsy » Tue Apr 29, 2014 9:44 pm

The problem with most sleep aids is that they are geared more to help you fall asleep and not so much staying asleep.
That's why they came out with Ambien CR (Controlled Release) to help maintain blood levels a little more consistently throughout the night so lessen the wake ups.

How to get into the ultra secret clinical setup menu can be found in the provider manual...see this pdf
http://www.apneuvereniging.nl/forum/pdf ... manual.pdf

Nocturia is a symptom of sleep apnea and the fact you still are experiencing it leads one to think your therapy is sub optimal and your report above points to more apnea events than we would like to see...and if you still have more apnea events even with leaks under control there's a good chance if your pressure was doing a better job preventing the apneas then the nocturia should reduce or go away. Of course there are other conditions that can cause nocturia but when we see a lot of apnea events...OSA moves to the top of the list real fast.

Let's see what the AHI and pressures are looking like with the leaks better managed and then you can decide if you want to change anything.
And yes..many of us do change our own pressure. It isn't hard to do and it isn't rocket science but I don't advise anyone doing it if they aren't feeling comfortable about doing it and understand why and what to expect. Changes need to be based on education and not willy nilly changing stuff just because we can.

Mask liners...
http://www.padacheek.com/PAC_Maskliners.html
Easy to make out of an old Tshirt if you want to just try one.
They help with leaks and comfort.
They will also help extend the life of the mask cushion because they will absorb a lot of the facial oils that tend to cause the silicon to get tired.
How often do you wash your mask cushion? With what? and how often are you allowed a new cushion?

I suspect you need more minimum pressure...how much...I don't know and it's hard to have an idea until we see some reports with excessive leaks removed from the picture.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Upalot
Posts: 15
Joined: Fri Apr 25, 2014 9:18 pm

Re: Sleepy Head Data Help, better crop

Post by Upalot » Mon May 12, 2014 7:37 pm

I think i figured out some of my leaking issues, its coming out of my eye sockets. I had not realized this until a few nights ago. I think my pressure it too high and it blows my mask off and then comes out of my eyes.

How do i deal with this one?

This stuff sucks, I use this think every single night and its not working at all for me. Its very depressing. I never feel good. I should say I never feel rested, never. Its a terrible feeling, I just kind of float around all day. Go to bed get a shitty nights sleep and than get up and do it all over again. This kind of stuff is what makes people wacky. Im sure this is why I have depression issues. I feel like I can not get away from it.

Any advice would be great. Thanks

_________________
Humidifier

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Sleepy Head Data Help, better crop

Post by robysue » Mon May 12, 2014 9:45 pm

Upalot wrote:I think i figured out some of my leaking issues, its coming out of my eye sockets. I had not realized this until a few nights ago. I think my pressure it too high and it blows my mask off and then comes out of my eyes.

How do i deal with this one?
You need to report this problem to the sleep doctor ASAP. Seriously. It's not a common problem with PAPing, but it's not unheard of either. And if the air is really getting into your eyes through your tear ducts, the sleep doc needs to be made aware of it. Sometimes the doc will authorize a slight reduction in pressure. Using an eye mask may also help. But if the problem is very severe, you may need to work with both a sleep doc and an eye doc.

This stuff sucks, I use this think every single night and its not working at all for me. Its very depressing. I never feel good. I should say I never feel rested, never. Its a terrible feeling, I just kind of float around all day. Go to bed get a shitty nights sleep and than get up and do it all over again. This kind of stuff is what makes people wacky. Im sure this is why I have depression issues. I feel like I can not get away from it.
Again, you need to report these problems to the sleep doc. And insist on a call back from the doc or his nurse or PA to talk about just how awful you feel during the day and how bad your sleep is in spite of (because of?) the PAP. You need to get the sleep doc involved in helping you figure out a way of making this work. It won't be easy, but sometimes you really do simply need to keep complaining so that they'll see you and you can speak to the doc, nurse, or PA face to face. Given how high your pressures run, it's also worth asking about whether a switch to a bilevel machine might be useful.

In my own very difficult adjustment to PAP therapy it took several doctor authorized changes in pressure, changes from CPAP to APAP to BiPAP to BiPAP Auto, and serious cognitive behavior therapy for the CPAP-induced insomnia before I started to feel even remotely comfortable sleeping with this dang mask on my nose each night. I was muddling through as well as I could, but hubby was very good at insisting that I call the sleep doc's office and let them know just how bad things were getting. On more than one occasion, hubby made the phone call himself because he was so alarmed at the sudden and drastic decline in my ability to function in the daytime.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5