Pressure pulse?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
old dude
Posts: 643
Joined: Sat Jul 27, 2013 4:08 pm

Pressure pulse?

Post by old dude » Thu Apr 03, 2014 1:47 pm

I've started to see something called a "pressure pulse" appear in SH from time to time. Can anyone tell me just what that is, and where does it fit in the general scheme of things with CAs, OAs, etc? Is it something to be concerned about? Evidently last night I had three of them.

User avatar
Bama Rambler
Posts: 189
Joined: Wed Feb 19, 2014 10:25 am

Re: Pressure pulse?

Post by Bama Rambler » Thu Apr 03, 2014 1:50 pm

The Pressure Pulse is a small burst of pressure that the machine uses when it's trying to determine whether an apnea is obstructive or central (clear Airway).

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: CPAP 15cmH2O

djhall
Posts: 323
Joined: Fri Jun 21, 2013 10:47 am
Location: Vacaville, CA

Re: Pressure pulse?

Post by djhall » Thu Apr 03, 2014 1:52 pm

By analyzing the flow response to device initiated pressure pulses during an apnea, the machine is able to distinguish an obstructed airway apnea from a clear airway apnea.

User avatar
Pugsy
Posts: 65294
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pressure pulse?

Post by Pugsy » Thu Apr 03, 2014 1:56 pm

Not something to be concerned about at all.
It's just the machine deciding it needs a little help figuring out if what it senses is related to open or closed airway...central vs obstructive in nature.
Tiny puff of air that won't even show up on the pressure line as an increase in pressure because it's so brief and small.
Pressure pulses don't always go along with flagged events either. I have seen them with no flagged event because the event was maybe 9 seconds and didn't last long enough to meet the 10 second minimum duration.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Pressure pulse?

Post by JDS74 » Thu Apr 03, 2014 2:07 pm

PR System one machines (I don't know about ResMed and others) tries to differentiate between an obstructive apnea and what they call a clear airway apnea. We refer to those as central apneas. In both cases, the air flow ceases. After a short period of time with no air flow, the PR machine creates a short duration increase in pressure referred to as a pressure pulse. If the machine detects an increase in pressure as a result, it determines that the airway is closed, i.e. obstructive apnea. If no corresponding increase is detected, it concludes that a clear airway apnea has occurred.

The difference exists because the air volume in the hose and mask is very much smaller than the air volume including your lungs. Depending on your size, the lung volume will be in multiples of liters while that of the hose/mask will be a fraction of one liter. A small and temporary increase in pressure has a much greater effect on the smaller volume so the two conditions are distinguishable.

Edit: if you want to generate one, just mask up, turn on the machine, breathe normally for a minute or two, and then just stop breathing for 20 seconds or so. Some of the folks here are very sensitive and don't like the feeling. I'm not so sensitive so I can't feel them when they are happening. In any event, as Pugsy says, occasional pressure pulses are likely of no significance at all. We all stop breathing now and then at night when we roll over, etc.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Last edited by JDS74 on Fri Apr 04, 2014 3:48 am, edited 1 time in total.
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

old dude
Posts: 643
Joined: Sat Jul 27, 2013 4:08 pm

Re: Pressure pulse?

Post by old dude » Thu Apr 03, 2014 6:06 pm

Thanks for the help everybody.

At least I don't have something new to think or worry about.