euipment with larger humidifier chamber

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
NWgirl
Posts: 6
Joined: Sat Mar 08, 2014 10:15 am

euipment with larger humidifier chamber

Post by NWgirl » Mon Mar 17, 2014 10:47 am

I am using a PR System One Remstar 60 series auto machine with heated tubing. The heated tubing improved things a whole lot for comfort. It hasn't put a dent in the running out of water problem. It happens twice a night and with no regularity. I even tried setting an alarm clock so I could refill the tank before it runs out. Now that I have tried all ways and sideways to solve the problem with no luck I am looking for a cpap that has a large capacity humidifier. Even 4 ounces of extra water could give me at least 6 hours of straight sleep. That would be wonderful! Cost is not an issue because calm sleep is a key to my health issues.
Thank you to anyone who has info out there.
NWgirl

User avatar
sleeplessinaz
Posts: 1067
Joined: Mon Oct 29, 2007 10:49 am
Location: Mesa, Arizona

Re: euipment with larger humidifier chamber

Post by sleeplessinaz » Mon Mar 17, 2014 10:59 am

I used the Resmed 9 And I never run out of water and I have mine set on 4.5. Might want to fill in your equipment in your profile as we can see what mask and machine you are using.

_________________
MachineMask
Additional Comments: new Airsense auto set 12/08/14. Cpap pressure is
Start Date 08/30/07
APAP setting is 6 to 12
HH 2.5
Side Sleeper
HypoThyroidism & Diabetes
New Airsense autoset 12/08/14

nanwilson
Posts: 3463
Joined: Tue Apr 13, 2010 10:35 am
Location: Southern Alberta

Re: euipment with larger humidifier chamber

Post by nanwilson » Mon Mar 17, 2014 11:03 am

We need to know what machine you are using and where you live.... arid Arizona, wet Florida?????? For starters you could take a look at cpap.com website, they have a stand alone humidifier that may have a bigger capacity than the one you are using.
Started cpap in 2010.. still at it with great results.

User avatar
Pugsy
Posts: 65014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: euipment with larger humidifier chamber

Post by Pugsy » Mon Mar 17, 2014 11:16 am

Water consumption can be affected by any of these or combinations of these
Ambient humidity where you live
Settings on the machine for humidity delivery (using a heated hose with the PR S1 setting choices are 1, 2 or 3 with 3 being 90%)
Leaks...the greater the leaks the greater the consumption
Pressures used..the greater the pressure the greater the consumption.

I max my PR S1 setting out and never have it use all the water in one night. I might get close with 8 hours of use but never had it run dry unless I simply forgot to fill it before bedtime. My pressures often go to 20 cm. Leaks are rarely large and prolonged though.

So what is your setting on the humidifier for humidity delivery?
What is your pressure setting?
How do your leaks look?
Not much we can do about ambient humidity unless we add a room humidifier to put more moisture in the air.

It is not normal to use 2 tanks of water each night during an 8 hour period.
It's not normal to use an entire tank of water in 8 hours unless there is something else going on to increase water consumption.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: euipment with larger humidifier chamber

Post by chunkyfrog » Mon Mar 17, 2014 11:25 am

My tank runs dry once in awhile-usually if I sleep late, or the room is very dry.
There have been no ill effects to tank, machine or me,
(--unless you count the unpleasant odor that tells me I'm overdue for a tank cleaning.)
My leak rate tells me my mouth leakage is minimal--so you might check your leak numbers.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Wulfman...

Re: euipment with larger humidifier chamber

Post by Wulfman... » Mon Mar 17, 2014 11:54 am

NWgirl wrote:I am using a PR System One Remstar 60 series auto machine with heated tubing. The heated tubing improved things a whole lot for comfort. It hasn't put a dent in the running out of water problem. It happens twice a night and with no regularity. I even tried setting an alarm clock so I could refill the tank before it runs out. Now that I have tried all ways and sideways to solve the problem with no luck I am looking for a cpap that has a large capacity humidifier. Even 4 ounces of extra water could give me at least 6 hours of straight sleep. That would be wonderful! Cost is not an issue because calm sleep is a key to my health issues.
Thank you to anyone who has info out there.
NWgirl
You seem to have problems following instructions...........
Every time you've posted, people have asked you to fill out your equipment profile to provide information. YOU HAVEN'T!
Your "running out of water" problem is most likely due to your mouth falling open while you sleep and the increased airflow is blowing all the water out your mouth. Too much humidification can actually cause more nasal congestion and contribute to nasal congestion......which causes mouth-breathing or mouth-leaking.

Reading through your previous posts, I read you live in the Pacific Northwest, so it's not like you're in an arid climate.
Are you using a nasal mask or a full face mask? I'm guessing a nasal mask.


Den

.

NWgirl
Posts: 6
Joined: Sat Mar 08, 2014 10:15 am

Re: euipment with larger humidifier chamber

Post by NWgirl » Mon Mar 17, 2014 12:10 pm

My setting on the Remstar Auto Aflex system one heated humidifier with heated tubing is a 3. I changed it from a 2 last night because the dryness wakes me up. The pressure setting is 9.5. No significant leaks. I was just in to the medical equipment guy and he checked it all out and said I was in compliance and all was well. I don't mouth breath. If I open my mouth I wake up. Thanks for all the info. I really don't understand this and am hoping there is something that has been overlooked by the support people to solve this soon.

Wulfman...

Re: euipment with larger humidifier chamber

Post by Wulfman... » Mon Mar 17, 2014 12:17 pm

NWgirl wrote:My setting on the Remstar Auto Aflex system one heated humidifier with heated tubing is a 3. I changed it from a 2 last night because the dryness wakes me up. The pressure setting is 9.5. No significant leaks. I was just in to the medical equipment guy and he checked it all out and said I was in compliance and all was well. I don't mouth breath. If I open my mouth I wake up. Thanks for all the info. I really don't understand this and am hoping there is something that has been overlooked by the support people to solve this soon.
We STILL don't know what mask you use.

Do you use software and download your data? What are your "leak rate" numbers? I've seen lots of people swear they don't mouth-leak or mouth-breathe only to find out they really DO.

Something doesn't make sense.


Den

.

User avatar
jencat824
Posts: 1442
Joined: Sun Oct 14, 2012 5:06 pm
Location: Jeffersonville, Indiana

Re: euipment with larger humidifier chamber

Post by jencat824 » Mon Mar 17, 2014 10:05 pm

It helps us to better help one another by going to your user profile & filling in all your equipment:
Machine
Mask
Humidifier
And in your signature line add your prescribed pressure.

This info will then appear in blue below your post, every time you post.

Sounds to me like you have a lot of leaking going on. My hubby has always used what I felt like was too much water each night. He has been struggling with leaks too. He just got a new mask & not only did the leaks just about disappear, he is using almost no water. He has the same machine as I do & is now using the Full Life FFM, an older mask but quite effective for him, obviously.

Jen

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure Auto 12-20cm CPAP compliant since 2000
Other masks I've tried: *=liked, #= no way
Piliaro, SleepWeaver Elan*, Swift FX w&w/o Bella Loops#, OpitLife#,Simplicity*, Mirage Vista*, Go Life for Her#, IQ (original hg only)*, Quattro FX (barely)###, Wisp*, Nuance#, Swift LT for her**

User avatar
munkyBeatz
Posts: 14
Joined: Sat Aug 25, 2012 3:32 pm

Re: euipment with larger humidifier chamber

Post by munkyBeatz » Tue Mar 18, 2014 2:17 am

It doesn't matter what machine the original poster is using. If you are running out of water completely that frequently, you need to have your machine checked. Either a seal is broken, or leak, or the heating element is acting up. Your DME can check for codes on the unit as well, they'll indicate if there is an issue related to that.

User avatar
digitalepiphany
Posts: 255
Joined: Thu Sep 05, 2013 12:31 am
Location: Dallas, TX

Re: euipment with larger humidifier chamber

Post by digitalepiphany » Tue Mar 18, 2014 2:55 am

The machine is not necessarily the problem. I was having the same issue with running out of water. Changing masks, and reducing my leak rate, has solved that problem. Now, I only run out of water if I sleep too long, which still happens more often than I'd like.

_________________
Mask: Amara Full Face CPAP Mask with Gel & Silicone Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: It's a VPAP ST. IPAP:16 EPAP: 12

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: euipment with larger humidifier chamber

Post by Sludge » Tue Mar 18, 2014 3:56 am

NWgirl wrote:No significant leaks. I was just in to the medical equipment guy and he checked it all out and said I was in compliance and all was well.
He's a LSOS.

Upload the machine data and you'll see what the problem is.
You Kids Have Fun!!

User avatar
Sheriff Buford
Posts: 4110
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: euipment with larger humidifier chamber

Post by Sheriff Buford » Tue Mar 18, 2014 4:58 am

Your machine should not run out of water (even TWICE a night). There is something definitely wrong with your humidifier. If the medical place where you got your equipment will not repair it or they say it's ok, ask to borrow a used humidifier for a few nights and see if the borrowed humidifier runs dry. I'll bet it won't.

Sheriff