User Flags

Post by Leadhead » Wed Feb 05, 2014 10:29 am

Last night I registered a "user flag" identified in the event graph as a green line. In looking at it the flow rate graph, it appears to be an apnea with flat line breathing for about 10 seconds. I have seen posts here talking about setting custom flags but I have not done this. Where in Sleepyhead are these flags set...I looked in preferences but no luck.

_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Additional Comments: 9.5/14 Pressure...Sleepyhead

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: User Flags

Post by Pugsy » Wed Feb 05, 2014 11:02 am

It's in Preferences...CPAP graph...on the right side....by default it is off.
Custom event flags let's you tell Sleepyhead to flag something that the machine wouldn't flag because it didn't meet criteria.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LTen
Posts: 44
Joined: Sat Dec 07, 2013 10:24 am
Location: Knoxville,tn
Contact:

Re: User Flags

Post by LTen » Wed Feb 05, 2014 12:06 pm

Wow, learnt something again today. !! Amazing how much there is to learn and how impressive you guys are to provide info.

User avatar
Leadhead
Posts: 44
Joined: Thu Feb 10, 2011 8:04 pm
Location: CO

Re: User Flags

Post by Leadhead » Wed Feb 05, 2014 12:21 pm

Thanks Pugsy
So, my user flag was set at 20% and 10 seconds...I didn't set it but it was there. It says user flags are not included in AHI. Since mine was set, does that mean all my 10 second apneas were not recorded in AHI and therefore distorted my stats?

_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Additional Comments: 9.5/14 Pressure...Sleepyhead

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: User Flags

Post by Pugsy » Wed Feb 05, 2014 12:39 pm

No, having the user flags set to pick up stuff that would normally get ignored doesn't alter the fact that you had 10 sec (or longer) events and your AHI is still your AHI.

Using lower % reduction as a filter doesn't negate the AHI you are seeing. If anything it would flag more events and not less because the reduction needed to get a flag was only 20% instead of the normal 50% for hyponeas or 90% for OAs.

I don't like using the custom settings for newbies who don't understand what they are seeing and are fixating on numbers anyway.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
johantv
Posts: 42
Joined: Wed Oct 15, 2014 1:42 pm
Location: The Netherlands

Re: User Flags

Post by johantv » Fri Oct 24, 2014 2:38 am

Hi,

In sleepyhead v09.08-01 I've defined 2 user flags. Sleepyhead indicates user flag set as 20% works best finding apneus.
Since apneus are normally 10+ seconds I set UF1 as 20% and 5 (till 10) sec, UF2 is set as 20% and 2.5 (till 5) sec. Below is the graph (my first here) which shows these UF's and flow rate. This night UF1# = 10 and UF2# = 31, AHI was 0.3
Are these flatliners less than 10 seconds normal? Anything wrong with my breathing or is this normal?

http://i.imgur.com/ujbf8LR.png

PS.
The graph is for UF1=5s and UF2=2.5s but if UF2 is set 20% and 1.5 second 160# found, If set for 1.0 second 996# are found
Start therapy: 2014/3 | AHI before 43, Now <1 | APAP 6.6 - 8.0cmH20 | EPR 1
No Pressure, no Diamonds

User avatar
johantv
Posts: 42
Joined: Wed Oct 15, 2014 1:42 pm
Location: The Netherlands

Re: User Flags

Post by johantv » Mon Oct 27, 2014 12:20 pm

Anyone? Perhaps it's nonsense to check for these small blockages but in combination with FL's which are always there I like to find out why my AHI is on average 0.7 and I'm still feeling tired.
Start therapy: 2014/3 | AHI before 43, Now <1 | APAP 6.6 - 8.0cmH20 | EPR 1
No Pressure, no Diamonds

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: User Flags

Post by Pugsy » Mon Oct 27, 2014 12:31 pm

johantv wrote:Anyone? Perhaps it's nonsense to check for these small blockages but in combination with FL's which are always there I like to find out why my AHI is on average 0.7 and I'm still feeling tired.
Maybe the answer to your fatigue isn't available in the "numbers" . Using 20% flow reduction criteria...I doubt if anyone would have a totally clean report even those without OSA. Might be useful if you have UARS but the therapy is still the same only you have to go by how you feel and not the numbers.

Otherwise you get the same questions everyone gets when they don't feel as good as the numbers say they should expect to feel.
How long have you been on cpap therapy?
How many hours of sleep are you getting?
Are those hours of sleep fragmented with a lot of wake ups and if so..any idea why?
Do you take any meds of any kind? This one catches a lot of people because often the side effects to various meds are fatigue or drowsiness or insomnia?
Do you have insomnia issues...either falling asleep or staying asleep?
Any issues with mask or bed comfort?
And any other health issues?
Google fatigue causes and you will see that there are many, many potential causes and effective cpap therapy can only fix it if the only cause is sleep apnea.
How's your sleep hygiene? Good or bad? Most of us do some bad sleep hygiene stuff...I know I do and I know better.

Finally...sometimes it just takes time and sometimes a lot of time. Since we have to use the machine anyway...while giving it time use that time to do some detective work to see if there's anything else that is going on that might impact how you feel.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
johantv
Posts: 42
Joined: Wed Oct 15, 2014 1:42 pm
Location: The Netherlands

User Flags

Post by johantv » Mon Oct 27, 2014 4:23 pm

@Pugsy, thanks for your answers and suggestions.

0) 20% seems nothing, but the flatliners look the same als apneus. So I thought a lot of short < 10secs apneus could perhaps make you feel bad so was interested how others scored....?
1) 7 months on cpap
2) 7/8 Hours a night
3) 3/4 awakenings every night, don't know why. Back to sleep in a few minutes
4) Yes, using statines against cholesterol since 4 months. Througt of that myself so stopped temporarily (one week now). Agreed with my docter to quit for two weeks and than evaluate. After one week stopping medicine still feeling the same though.
5) no insomnia or other issues
6) no mask issues
7) except overweight, no sports....the rest is ok
don't know what you mean, bed is cleaned every 2 weeks I think
9) thanks, investigating a lot, but feeling worse now than when I started therapy. Before therapy I fell a sleep on the couch every night, but during daytime no mayor issues. Now more fatigued during the day. Best thing is no naps at evening (not allowing myself) and not sleeping apart from the wife (no snoring anymore).

Last months experimented with pressure since AHI is always low <1 and having some eyeproblems since start of therapy. So slept 4 months APAP 5-8.6, than 1 month CPAP 5 (very tired, Many FL's), than gradually increasing pressure till now 1 month around 6-12. So perhaps changing pressure a lot wasn't consistent so perhaps a stable pressure should improve my situation?!
Start therapy: 2014/3 | AHI before 43, Now <1 | APAP 6.6 - 8.0cmH20 | EPR 1
No Pressure, no Diamonds

tan
Posts: 565
Joined: Sat Aug 30, 2014 11:58 am

Re: User Flags

Post by tan » Mon Oct 27, 2014 4:28 pm

Not sleeping (like 3-4 awakenings a night) IS insomnia.

and statins are known to cause insomnia and depression for that matter. Are you numbers that bad that you have to take statins? What are your TRG, HDL, HDL and total CHOL? If LipoProfile test, what is the particle number?

User avatar
johantv
Posts: 42
Joined: Wed Oct 15, 2014 1:42 pm
Location: The Netherlands

User Flags

Post by johantv » Mon Oct 27, 2014 4:58 pm

@tan, my idea was that I woke up because of incidental leakages due to higher pressure ot the higher pressure as reason itself. As said, after waking up I fall a sleep again right away.

The docter subscribed them since the ldl/hdl index was too high (don't know the exact figures but think it was 4 or 6). Have a BMI of 34, smoked, no sport, so dangerously enough to prescribe statins. Stopped smoking now for 3 months.

Will evaluate after 2 weeks without statins.
(Wish I was like Homer)
Start therapy: 2014/3 | AHI before 43, Now <1 | APAP 6.6 - 8.0cmH20 | EPR 1
No Pressure, no Diamonds

tan
Posts: 565
Joined: Sat Aug 30, 2014 11:58 am

Re: User Flags

Post by tan » Mon Oct 27, 2014 5:02 pm

johantv wrote:@tan, my idea was that I woke up because of incidental leakages due to higher pressure ot the higher pressure as reason itself. As said, after waking up I fall a sleep again right away.
Insomnia, or sleeplessness, is a sleep disorder in which there is an inability to fall asleep or to stay asleep as long as desired
The docter subscribed them since the ldl/hdl index was too high (don't know the exact figures but think it was 4 or 6). Have a BMI of 34, smoked, no sport, so dangerously enough to prescribe statins. Stopped smoking now for 3 months.

Will evaluate after 2 weeks without statins.
Hard to comment here without knowning numbers, but your background suggests that you are at a higher risk, though I am not sure that statins fix that. But I aint a doctor.

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: User Flags

Post by Pugsy » Mon Oct 27, 2014 10:48 pm

johantv wrote:don't know what you mean, bed is cleaned every 2 weeks I think
Bed comfort as is your bed comfortable for you or is it old and tired and causing aches and pains and your bed pillow..does it work well with your mask.
Aches and pains don't let us sleep so great...that one I know for sure.
johantv wrote:) no insomnia or other issues
3) 3/4 awakenings every night, don't know why. Back to sleep in a few minutes
waking frequently during the night is called sleep maintenance insomnia
there's also the typical sleep onset insomnia where we have trouble falling asleep ...people can have or both.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
johantv
Posts: 42
Joined: Wed Oct 15, 2014 1:42 pm
Location: The Netherlands

User Flags

Post by johantv » Tue Oct 28, 2014 12:06 pm

The bed is fine. A boxspring and a good, firm pillow.
Perhaps taking statins (simvastatine) is the reason for this insomnia (not falling asleep but wakening up to often). The last 2 months I raised my pressure so I blamed that for some dizzyness at daytime and wakening up more often then Usual. Anyway, I have to continu my quest and hopefully find the right balance...
Start therapy: 2014/3 | AHI before 43, Now <1 | APAP 6.6 - 8.0cmH20 | EPR 1
No Pressure, no Diamonds

User avatar
johantv
Posts: 42
Joined: Wed Oct 15, 2014 1:42 pm
Location: The Netherlands

Re: User Flags

Post by johantv » Tue Dec 02, 2014 10:27 am

Good news, my sleep consultant wants me to have a new sleep study end of december to find out why I'm wakening up a few times every night.
Start therapy: 2014/3 | AHI before 43, Now <1 | APAP 6.6 - 8.0cmH20 | EPR 1