Late Sleep Cycle Cluster Events: Strategies?
- DavidCarolina
- Posts: 477
- Joined: Wed Feb 23, 2011 6:23 pm
Late Sleep Cycle Cluster Events: Strategies?
I love this board, as its the first place i learned
1. Most people here know more about apnea than their doctors do
2. Its not the "average AHI" but rather how you FEEL next day and preventing major clusters
In my case , I can take a nap during the day one hour without machine (NOT RECOMMENDED! but i "can") and have few problems. My sleepyhead data always shows the first hour of sleeping is event free more or less.
As I approach maybe the 4-6 hour point of a 7 hour sleep sequence my "lower than 1 AHI" numbers can skyrocket into something like 6, and my periodic breathing can do the same.
So Im thinking: Alarm clock before it happens? Increase the pressure after 4 hours sleep? Or simply dont sleep more than 4 hours at a shot, split it up?
Have no idea, anyone help? Because if i go to cleveland clinic sleep center my 2.0 "average" AHI looks great to them. They generally just dont care about "EVENTS".
1. Most people here know more about apnea than their doctors do
2. Its not the "average AHI" but rather how you FEEL next day and preventing major clusters
In my case , I can take a nap during the day one hour without machine (NOT RECOMMENDED! but i "can") and have few problems. My sleepyhead data always shows the first hour of sleeping is event free more or less.
As I approach maybe the 4-6 hour point of a 7 hour sleep sequence my "lower than 1 AHI" numbers can skyrocket into something like 6, and my periodic breathing can do the same.
So Im thinking: Alarm clock before it happens? Increase the pressure after 4 hours sleep? Or simply dont sleep more than 4 hours at a shot, split it up?
Have no idea, anyone help? Because if i go to cleveland clinic sleep center my 2.0 "average" AHI looks great to them. They generally just dont care about "EVENTS".
- ChicagoGranny
- Posts: 15446
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Late Sleep Cycle Cluster Events: Strategies?
That sounds like a crybaby to me, but I will play along.skyrocket into something like 6
How long are those events? How is your energy level the next day?
ChicagoGramps
Last edited by ChicagoGranny on Tue Jan 21, 2014 8:04 pm, edited 1 time in total.
Re: Late Sleep Cycle Cluster Events: Strategies?
Resmed does not use the word VSNORE but only SNORE. S9 Autoset knows who to eliminate it almost completely.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png
- SleepyBobR
- Posts: 312
- Joined: Tue Jan 19, 2010 4:42 pm
- Location: Toronto, Canada
Re: Late Sleep Cycle Cluster Events: Strategies?
By "skyrocket" to 6 I assume you mean the average for the night goes from 1 after 6 hours of sleep and moves up to 6 after 7 hours. If that's what you mean then your AHI during the last hour is close to 40 which is quite bad. Not sure what to suggest but I first want to confirm that that's what you meant. Otherwise, I agree with Chicago.
_________________
| Mask: Simplus Full Face CPAP Mask with Headgear |
| Additional Comments: AirSense 10 AutoSet For Her @ 7-10 |
-
Wulfman...
Re: Late Sleep Cycle Cluster Events: Strategies?
For what it's worth, you might want to update the equipment in your profile.DavidCarolina wrote:I love this board, as its the first place i learned
1. Most people here know more about apnea than their doctors do
2. Its not the "average AHI" but rather how you FEEL next day and preventing major clusters
In my case , I can take a nap during the day one hour without machine (NOT RECOMMENDED! but i "can") and have few problems. My sleepyhead data always shows the first hour of sleeping is event free more or less.
As I approach maybe the 4-6 hour point of a 7 hour sleep sequence my "lower than 1 AHI" numbers can skyrocket into something like 6, and my periodic breathing can do the same.
So Im thinking: Alarm clock before it happens? Increase the pressure after 4 hours sleep? Or simply dont sleep more than 4 hours at a shot, split it up?
Have no idea, anyone help? Because if i go to cleveland clinic sleep center my 2.0 "average" AHI looks great to them. They generally just dont care about "EVENTS".
Beyond that, what is the breakdown of "events" in the period that they're higher........more apneas or hypopneas?
Could be increased REM at that time........but, could it be positional? (only you know what your sleep position could be)
You should be able to see a pattern in your reports. You need to do a little sleuthing.
Den
.
Re: Late Sleep Cycle Cluster Events: Strategies?
I have the same issues. My last hour is almost always in the 10-15 range. Mostly Centrals, with Periodics as well. I attribute them to my waking up. Often I will wake up for a minute or two about an hour before getting up, and if I do, I turn the machune off for a few seconds and then back on. That way I can turn the last hour off in Sleepyhead and get a better idea of my true AHI.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: AHI 58 Pressure 9cm O2 sat 75% |
- SleepyBobR
- Posts: 312
- Joined: Tue Jan 19, 2010 4:42 pm
- Location: Toronto, Canada
Re: Late Sleep Cycle Cluster Events: Strategies?
Seems to me your true AHI for the night should contain the last hour, bad or good. Why screen it out?
_________________
| Mask: Simplus Full Face CPAP Mask with Headgear |
| Additional Comments: AirSense 10 AutoSet For Her @ 7-10 |
Re: Late Sleep Cycle Cluster Events: Strategies?
I like to see what it would be without it. I have a lot of centrals as I am waking up. It has nothing to do with OSA, for which I am being treated. Actually, my AHI really doesn't matter a bit. What matters to me is that I don't fall asleep during important conversations.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: AHI 58 Pressure 9cm O2 sat 75% |
Re: Late Sleep Cycle Cluster Events: Strategies?
If you are having centrals during REM time I would not use an alarm clock to interfere with REM as it is very important to your well being to have REM. I would ask your sleep specialist what to do if your AHI on a daily basis is >5.
Re: Late Sleep Cycle Cluster Events: Strategies?
I don't use an alarm clock at all. I don't do anything to intentionally interfere with sleep! However, when I wake up I usually turn the thing off and back on to document when I woke up.
As an example, last night I slept from midnight until 605 with an AHI of 5.1.
Between 605 and 716 my AHI was 21.2 because I was waking up. So it works out to a 7.22 for the night. I just like to be able to break it down. My OHA for the night was .59 and for the last session, was 0.0.
As an example, last night I slept from midnight until 605 with an AHI of 5.1.
Between 605 and 716 my AHI was 21.2 because I was waking up. So it works out to a 7.22 for the night. I just like to be able to break it down. My OHA for the night was .59 and for the last session, was 0.0.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: AHI 58 Pressure 9cm O2 sat 75% |
- ChicagoGranny
- Posts: 15446
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Late Sleep Cycle Cluster Events: Strategies?
How many As and how many Hs? How long were the As?Between 605 and 716 my AHI was 21.2 because I was waking up.
ChicagoGramps
Re: Late Sleep Cycle Cluster Events: Strategies?
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: AHI 58 Pressure 9cm O2 sat 75% |
-
ReadyforRest
- Posts: 136
- Joined: Thu Jul 26, 2012 12:43 pm
- Location: Ontario, Canada
Re: Late Sleep Cycle Cluster Events: Strategies?
This thread seems to have diverted away from the original poster and his needs/issues. May I suggest that, Salton, you start your own thread to discuss your issues, so that DavidCarolina's can be addressed here?
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: SleepyHead v0.9.1, Pressure: 9 - 14 |
Re: Late Sleep Cycle Cluster Events: Strategies?
Hi David!DavidCarolina wrote:I love this board, as its the first place i learned
1. Most people here know more about apnea than their doctors do
2. Its not the "average AHI" but rather how you FEEL next day and preventing major clusters
In my case , I can take a nap during the day one hour without machine (NOT RECOMMENDED! but i "can") and have few problems. My sleepyhead data always shows the first hour of sleeping is event free more or less.
As I approach maybe the 4-6 hour point of a 7 hour sleep sequence my "lower than 1 AHI" numbers can skyrocket into something like 6, and my periodic breathing can do the same.
So Im thinking: Alarm clock before it happens? Increase the pressure after 4 hours sleep? Or simply dont sleep more than 4 hours at a shot, split it up?
Have no idea, anyone help? Because if i go to cleveland clinic sleep center my 2.0 "average" AHI looks great to them. They generally just dont care about "EVENTS".
I had the same issue when I was in REM I was having clusters of OA's. The 2 to 3 hour period just looked terrible on the sleep reports. I was feeling OK, but thought I could fine tune it a bit. I ended up bumping up my EPAP setting up from 17 to 19.
The amount of episodes dropped and my AHI is now mostly under 1 where previously it was alway 2 to 3.
I wish you good luck!
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Settings are IPap 23 EPap 19 |






