Hose from cpap gurgles and makes popping noises

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Cinders
Posts: 9
Joined: Sat Nov 16, 2013 9:38 am

Hose from cpap gurgles and makes popping noises

Post by Cinders » Wed Jan 15, 2014 8:07 pm

My hose has been waking me up with noises. I know that sounds odd. I air it out every day so the moisture leaves.
Anybody have this problem? I FINALLY got comfortable wearing my mask everynight. What's with tho hose??

mahalo111
Posts: 8
Joined: Wed Dec 18, 2013 8:34 pm

Re: Hose from cpap gurgles and makes popping noises

Post by mahalo111 » Wed Jan 15, 2014 8:19 pm

Your humidity is set too high. I have to keep mine on one or it fills with water in the middle of the night.

_________________
Mask

mahalo111
Posts: 8
Joined: Wed Dec 18, 2013 8:34 pm

Re: Hose from cpap gurgles and makes popping noises

Post by mahalo111 » Wed Jan 15, 2014 8:20 pm

Also try filling the water resevoir only 1/4 full.

_________________
Mask

User avatar
LSAT
Posts: 13360
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Hose from cpap gurgles and makes popping noises

Post by LSAT » Wed Jan 15, 2014 8:23 pm

The problem is water in the hose that is blocking the air.. Try moving your CPAP to a position lower than your mattress. Possibly near the floor. That will cause any water build up to run down hill to the humidifier. Also check your humidity setting and lower it. You may need to cover your hose. Condensation forms when the room is cool and the air in the hose is warm. That adds to the moisture problem .If you continue to have problems try the "hose buddy".

https://www.cpap.com/productpage/HoseBu ... ystem.html

User avatar
Pugsy
Posts: 65146
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Hose from cpap gurgles and makes popping noises

Post by Pugsy » Wed Jan 15, 2014 8:27 pm

You have the 60 series machine?? Do you by any chance already have the heated hose? If unsure..the heated hose has a black end where it attaches to the humidifier.

Moisture in the hose or mask will make some pretty weird and loud noises. Called rainout or condensation. Very common.

Several options to reduce this annoyance.

Increase the bedroom temperature.
Decrease humidity level chosen if you can do so comfortably. But sometimes lowering the humidity causes issues with sinuses and nasal mucosa.
Use a heated hose.
Put machine quite a bit lower than mattress so that gravity causes the moisture to flow downhill out of the hose.
Hose cozy to help insulate the hose.

See my posts in this thread about rain out in the mask and/or in the hose
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Wed Jan 15, 2014 8:37 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

TheDuke
Posts: 259
Joined: Tue Oct 27, 2009 10:23 pm
Location: Western Kentucky
Contact:

Re: Hose from cpap gurgles and makes popping noises

Post by TheDuke » Wed Jan 15, 2014 8:30 pm

I suggest using a well insulated hose sleeve. I run my humidifier at a high level and sleep in a cool room (65 degree) and never have "rainout". My hose sleeve came from one of the on-line CPAP suppliers, and is made of nylon shell with some sort of fiberfill insulation like an insulated jacket.