Hi,
Anyone having a problem with getting and keeping a good seal with the heated humidifier chamber? When full to the mark, there is still a air leak in the unit (loud). No humidifier no noise. If so how did you stop it?
Happy New Year
Muddpa
PR System One Heated Humidifier
PR System One Heated Humidifier
Some were there is a mask for me 

-
- Posts: 184
- Joined: Fri Jun 12, 2009 8:32 pm
- Location: NW Colorado 50 miles from nowhere
- Contact:
Re: PR System One Heated Humidifier
I trust you've checked the soft gaskets where the two parts join, the back of the humidifier to the tank and the tank to the lid? Latched and unlatched the two halves of the tank itself? There's no moving parts to go south in the humidifier so it seems it should be a gasket somewhere.
_________________
Machine: DreamStation Auto CPAP Machine |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: PulseOx CMS-50D+, Papcap, SleepyHead |
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: PR System One Heated Humidifier
Gasket tears are easy to miss; our hosts sell replacement gaskets for many humidifiers.
I even bought a spare awhile back--probably ought to pull it out and look over the old one.
At least I'd know where the spare is, because, as we all know, most failures happen on holidays and weekends.
I even bought a spare awhile back--probably ought to pull it out and look over the old one.
At least I'd know where the spare is, because, as we all know, most failures happen on holidays and weekends.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- 2 B Sleeping Soundly
- Posts: 822
- Joined: Sat Nov 05, 2011 2:24 pm
- Location: Southern California
Re: PR System One Heated Humidifier
Muddpa,
I believe kwikwater and CF are probably right about the problem you are having. I once dislodged the dry box seal while wiping down the inside of the humidifier. You can feel it in the back of the humidifier once the water tank is removed. When the humidifer was turned on it made a lot of noise and the air flow was severely compromised. I used a flashlight to see where to reinstall the seal and after doing so had no issues.
If the seal is torn or damaged the you can get them from our host very cheap https://www.cpap.com/productpage/Dry-Bo ... ifier.html. I bought some as back ups as you never know when you might need something, like what CF alluded to.
John
I believe kwikwater and CF are probably right about the problem you are having. I once dislodged the dry box seal while wiping down the inside of the humidifier. You can feel it in the back of the humidifier once the water tank is removed. When the humidifer was turned on it made a lot of noise and the air flow was severely compromised. I used a flashlight to see where to reinstall the seal and after doing so had no issues.
If the seal is torn or damaged the you can get them from our host very cheap https://www.cpap.com/productpage/Dry-Bo ... ifier.html. I bought some as back ups as you never know when you might need something, like what CF alluded to.
John
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Additional Comments: 11-09-11 Start PAP therapy. Current settings: APAP 13cmH2O - 17cmH2O / Ramp, off / A-Flex, off / Respironics Premium Chin Strap / Sleepyhead Software |
One time a cop pulled me over for running a stop sign. He said, "Didn't you see the stop sign?" I said, "Yeah, but I don't believe everything I read." -- Steven Wright
Re: PR System One Heated Humidifier
Hey Gang,
Thanks for all of the input. Will put on my glasses and flashlight and see what happen.
Thanks for all of the input. Will put on my glasses and flashlight and see what happen.
Some were there is a mask for me 