lowering the air pressure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
rhimell2012
Posts: 1
Joined: Wed Dec 04, 2013 11:28 pm

lowering the air pressure

Post by rhimell2012 » Wed Dec 04, 2013 11:31 pm

I have a Remstar Auto with C-flex M # 1017453 How do I lower the air pressure. It was set at 7 and now it is blowing much higher around 14. It is discontinued and I need it as a back up to my newer machine.

User avatar
Pugsy
Posts: 65301
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: lowering the air pressure

Post by Pugsy » Thu Dec 05, 2013 8:55 am

The Legacy series machine?? That looks like this one.
http://www.cpap-supply.com/REMstar-Auto ... 017453.htm

You can request the provider manual here or there are short cut directions as to how to get to clinical setup here at this link.
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

The Legacy model was pre M series which was released in 2007.....
Respironics REMstar Auto (older pre-2007 standard unit, not "M" series, not the PR System One): you can enter the setup menu by holding down the Ramp button and both of the right and left arrow buttons under the display window, while plugging in the CPAP. The 3 buttons must be held down until the setup screen starts

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.