Apap Help Resmed S9

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Gottop69
Posts: 14
Joined: Wed Dec 04, 2013 9:02 pm

Apap Help Resmed S9

Post by Gottop69 » Wed Dec 04, 2013 9:06 pm

I am new to the forums


I have the Resmed S9 with, climateline, H5I and ffm Quattro Fx.

I am waking up really stopped up and uncomfortable. I have been on it for 2 months. I have tried every setting on H5i and no luck. I live in Dallas tx. My pressure was set to 5-20 my 95% is 8.8. My ahi is 1.3.

Any suggestions or help would be awesome.

User avatar
LSAT
Posts: 13411
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Apap Help Resmed S9

Post by LSAT » Wed Dec 04, 2013 9:36 pm

Gottop69 wrote:I have the Resmed S9 with, climateline, H5I and ffm Quattro Fx.

I am waking up really stopped up and uncomfortable. I have been on it for 2 months. I have tried every setting on H5i and no luck. I live in Dallas tx. My pressure was set to 5-20 my 95% is 8.8. My ahi is 1.3.

Any suggestions or help would be awesome.


If you are "stopped up" ? You may need a Laxative

User avatar
Gottop69
Posts: 14
Joined: Wed Dec 04, 2013 9:02 pm

Re: Apap Help Resmed S9

Post by Gottop69 » Wed Dec 04, 2013 9:52 pm

Lol mean in my sinuses.

Mattbnr
Posts: 16
Joined: Tue Nov 26, 2013 8:17 pm

Re: Apap Help Resmed S9

Post by Mattbnr » Wed Dec 04, 2013 10:35 pm

Have you tried saline solution before going to sleep? That ayr stuff is supposed to be pretty good stuff.

_________________
Humidifier
S9 autoset diagnosed cpap pressure 14
using APAP 13-17 pressure
Swift LT nasal pillows

User avatar
Nozzelnut
Posts: 188
Joined: Thu Sep 19, 2013 2:29 pm
Location: Western, NY

Re: Apap Help Resmed S9

Post by Nozzelnut » Wed Dec 04, 2013 11:07 pm

Have you replaced the filter on the machine?

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PRS1 60 series for back-up; Sleepyhead and Rescan 4.3

User avatar
Gottop69
Posts: 14
Joined: Wed Dec 04, 2013 9:02 pm

Re: Apap Help Resmed S9

Post by Gottop69 » Thu Dec 05, 2013 9:30 am

Yes tried saline and have replaced filter. Last night left humidifier on 1 and hose on 74. Not as bad today.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Apap Help Resmed S9

Post by robysue » Thu Dec 05, 2013 12:55 pm

Gottop69 wrote:Yes tried saline and have replaced filter. Last night left humidifier on 1 and hose on 74. Not as bad today.
Have you tried turning the humidifier completely OFF? If there's water in the tank, but no heat, the air will still pick up a wee bit of humidity. That's all some people need to be comfortable. There are even a few folks whose noses are happier with NO humidifier at all.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Gottop69
Posts: 14
Joined: Wed Dec 04, 2013 9:02 pm

Re: Apap Help Resmed S9

Post by Gottop69 » Thu Dec 05, 2013 1:35 pm

I will try that tonight. I have the climate line tubing. Will the air still be warm? I was told I may even have asthma

User avatar
Tino2You
Posts: 635
Joined: Tue May 01, 2012 7:18 am
Location: Wilmington DE

Re: Apap Help Resmed S9

Post by Tino2You » Thu Dec 05, 2013 1:50 pm

what is your setting for the climateline?

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Autoset Pressure 9-15.6, EPR 1, no ramp
Tino

User avatar
Tino2You
Posts: 635
Joined: Tue May 01, 2012 7:18 am
Location: Wilmington DE

Re: Apap Help Resmed S9

Post by Tino2You » Thu Dec 05, 2013 1:52 pm

went back through and saw where you have it set to 1 and 74. I let mine autoadjust the humidity and keep the temp to 78. Works for me. I use about 3/4 a tank a night. More often than not, congested sinuses are a resule of too little humidity rather than too much.

-tino

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Autoset Pressure 9-15.6, EPR 1, no ramp
Tino

User avatar
StuUnderPressure
Posts: 1378
Joined: Mon Jun 18, 2012 10:34 am
Location: USA

Re: Apap Help Resmed S9

Post by StuUnderPressure » Thu Dec 05, 2013 3:38 pm

Tino2You wrote:went back through and saw where you have it set to 1 and 74. I let mine autoadjust the humidity and keep the temp to 78. Works for me. I use about 3/4 a tank a night. More often than not, congested sinuses are a resule of too little humidity rather than too much.

-tino
My opinion is also that "Auto" should be used for the ClimateLine hose & then only the temperature would have to be set.

That maintains a constant 80% humidity no matter what temperature you set.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02

SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html

User avatar
Gottop69
Posts: 14
Joined: Wed Dec 04, 2013 9:02 pm

Re: Apap Help Resmed S9

Post by Gottop69 » Thu Dec 05, 2013 6:09 pm

Well I just left ent. I have to have sinus surgery. And have my septum rebuilt. Seems fun. They want to do it before the first of the year.