Looking for some advice here....
Reading other post, I decided to go from APAP to straight CPAP....just to see if the pressure change is causing any issues.
I have been using this for 14 days..below is last nites numbers.
My AHI average for last two weeks has been 2.93. Prior to that, two week avg was 2.65
I still feel fatigued in morning whether its the same as before , its hard to say.
Should I continue w/CPAP or go back to Auto pressure?
thxs

-
Wulfman...
Post
by Wulfman... » Fri Nov 01, 2013 3:20 pm
Maybe you need a little more pressure.
Without seeing previous reports and what your pressures were doing with the range you were using, it's hard to determine what you need in CPAP mode......because when you went straight pressure, you picked the minimum pressure from your range.
Den
.
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Fri Nov 01, 2013 3:43 pm
Wulfman... wrote:Maybe you need a little more pressure.
Without seeing previous reports and what your pressures were doing with the range you were using, it's hard to determine what you need in CPAP mode......because when you went straight pressure, you picked the minimum pressure from your range.
Den
.
It still in my sig....after my 30 day eval, my Sleep Doc upped it from 8 to 9.....
-
Wulfman...
Post
by Wulfman... » Fri Nov 01, 2013 4:50 pm
darco wrote:Wulfman... wrote:Maybe you need a little more pressure.
Without seeing previous reports and what your pressures were doing with the range you were using, it's hard to determine what you need in CPAP mode......because when you went straight pressure, you picked the minimum pressure from your range.
Den
.
It still in my sig....after my 30 day eval, my Sleep Doc upped it from 8 to
9.....
I know......I read that. That's what I'm talking about. Maybe you need a little more than 9 cm. (what you picked for your fixed pressure trial which is indicated in the report you posted). Maybe try 10 cm. for a week or so. I'm sure your pressure went up to or over that when you were using a range of pressures.
Den
.
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Sun Nov 03, 2013 9:29 am
Wulfman... wrote:darco wrote:Wulfman... wrote:Maybe you need a little more pressure.
Without seeing previous reports and what your pressures were doing with the range you were using, it's hard to determine what you need in CPAP mode......because when you went straight pressure, you picked the minimum pressure from your range.
Den
.
It still in my sig....after my 30 day eval, my Sleep Doc upped it from 8 to
9.....
I know......I read that. That's what I'm talking about. Maybe you need a little more than 9 cm. (what you picked for your fixed pressure trial which is indicated in the report you posted). Maybe try 10 cm. for a week or so. I'm sure your pressure went up to or over that when you were using a range of pressures.
Den
.
Well what the heck...I upped it to 10 last nite...AHI under 2.00...Ill see how it goes over the week...nothing to lose
-
Wulfman...
Post
by Wulfman... » Sun Nov 03, 2013 10:50 am
darco wrote:Wulfman... wrote:darco wrote:Wulfman... wrote:Maybe you need a little more pressure.
Without seeing previous reports and what your pressures were doing with the range you were using, it's hard to determine what you need in CPAP mode......because when you went straight pressure, you picked the minimum pressure from your range.
Den
.
It still in my sig....after my 30 day eval, my Sleep Doc upped it from 8 to
9.....
I know......I read that. That's what I'm talking about. Maybe you need a little more than 9 cm. (what you picked for your fixed pressure trial which is indicated in the report you posted). Maybe try 10 cm. for a week or so. I'm sure your pressure went up to or over that when you were using a range of pressures.
Den
.
Well what the heck...I upped it to 10 last nite...AHI under 2.00...Ill see how it goes over the week...nothing to lose
One old saying we've used around here is......."If the numbers get better, you know you're going in the right direction."
The other thing is to give each setting 3 - 7 nights to see the (average) results. Changing the settings too often can result in frustration. That's referred to as "dial wingin' ".
Den
.
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Thu Nov 07, 2013 4:49 pm
Ok Wulfman, I took your advise and upped my pressure to 10cm.....started this on the 2nd.
Well by looking at my graphs, not only was my AHI the lowest ever, my events are dropping like a rock. Typical number of events averages prior were around 30.
The thing is, I am still very fatigued in the mornings....When the afternoons come around, there is no questions as to nap, I have to!
Any suggestions on what I can do next? Also anyone can help interpret my graphs?
thxs
d

-
LSAT
- Posts: 13393
- Joined: Sun Nov 16, 2008 10:11 am
- Location: SE Wisconsin
Post
by LSAT » Thu Nov 07, 2013 5:00 pm
If your numbers are low but you still feel fatigued, the possibility exists that there is some problem other than OSA. Have you consulted with your doctor?
-
Wulfman...
Post
by Wulfman... » Thu Nov 07, 2013 5:27 pm
darco wrote:Ok Wulfman, I took your advise and upped my pressure to 10cm.....started this on the 2nd.
Well by looking at my graphs, not only was my AHI the lowest ever, my events are dropping like a rock. Typical number of events averages prior were around 30.
The thing is, I am still very fatigued in the mornings....When the afternoons come around, there is no questions as to nap, I have to!
Any suggestions on what I can do next? Also anyone can help interpret my graphs?
thxs
d
THAT'S GREAT NEWS!
The reports look "reasonable", in my opinion. But, now that you've gotten your therapy sort of dialed in, you need to be patient and give it some time to work. You've had this setting for less than a week.......it can take weeks to months to really see the benefits. This isn't like a light switch that when it's turned on you see "immediate" results. You probably have a "sleep debt" to pay off, and that can only realistically start when the proper settings are obtained.
I'm speaking from experience when I say this. It's "baby steps" and "one day at a time".
Den
.
-
Pugsy
- Posts: 65258
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Thu Nov 07, 2013 5:35 pm
While you are giving the new settings time you might look at other possible factors.
How is your sleep quality? Looks like a good number of hours of sleep but are those hours fragmented for any reason? Do you remember many wake ups...if so..any idea why.
Your leak line is kinda all over the place. While technically it says below large leak territory is there any chance that mask issues or leaks are disturbing your sleep?
Do you take any medications even OTC stuff?
Do you have any other health issues?
Pain?
How comfortable is your bed?
Have you have a complete physical including labs (the basics like thyroids and stuff but also maybe hormones and vitamin D)?
Other than the leak line I don't see anything exciting on your reports and even the leak line isn't horribly exciting. Just a maybe little exciting if you wake often fiddling with the mask and leaks.
There's so much more to feeling the good numbers that we see on the report. Getting a nice low AHI is really pretty easy but sometimes it isn't all about the numbers.
I may have to RISE but I refuse to SHINE.
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Thu Nov 07, 2013 9:21 pm
Pugsy wrote:While you are giving the new settings time you might look at other possible factors.
How is your sleep quality?
***I feel I sleep thru the nite but the quality is lacking since I feel tired in AM
Looks like a good number of hours of sleep but are those hours fragmented for any reason? Do you remember many wake ups...if so..any idea why.
***I do get wake ups but are very brief. Weekends I tend to only sleep in an extra 90 mins
Your leak line is kinda all over the place. While technically it says below large leak territory is there any chance that mask issues or leaks are disturbing your sleep?
***I do notice at times when the mask leaks, I shift it back and immediately fall back asleep. I feel very comfortable w/this mask but I only had one other that I used for a week..
Do you take any medications even OTC stuff?
**No
Do you have any other health issues?
**No, 5-9 175 32w... 50yr old
Pain?
**No
How comfortable is your bed?
***its 3yr old Serta something, firm mattress..its a nice bed
Have you have a complete physical including labs (the basics like thyroids and stuff but also maybe hormones and vitamin D)?
*** I had extensive test that revealed nothing......I do take fish oil and Mens one a day , glucosamine,,,drink alcohol on weekends....
Other than the leak line I don't see anything exciting on your reports and even the leak line isn't horribly exciting. Just a maybe little exciting if you wake often fiddling with the mask and leaks.
***I missed an opportunity at the Stanford Research center where they had Mask Vendors showing off their wares..I would like to try another just to be sure...Im really think to just give it time....yes I shouldnt chase numbers....the event drop to me is huge....
There's so much more to feeling the good numbers that we see on the report. Getting a nice low AHI is really pretty easy but sometimes it isn't all about the numbers.
Last edited by
darco on Sun Nov 10, 2013 11:52 am, edited 1 time in total.
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Thu Nov 07, 2013 9:33 pm
Wulfman... wrote:darco wrote:Ok Wulfman, I took your advise and upped my pressure to 10cm.....started this on the 2nd.
Well by looking at my graphs, not only was my AHI the lowest ever, my events are dropping like a rock. Typical number of events averages prior were around 30.
The thing is, I am still very fatigued in the mornings....When the afternoons come around, there is no questions as to nap, I have to!
Any suggestions on what I can do next? Also anyone can help interpret my graphs?
thxs
d
THAT'S GREAT NEWS!
The reports look "reasonable", in my opinion. But, now that you've gotten your therapy sort of dialed in, you need to be patient and give it some time to work. You've had this setting for less than a week.......it can take weeks to months to really see the benefits. This isn't like a light switch that when it's turned on you see "immediate" results. You probably have a "sleep debt" to pay off, and that can only realistically start when the proper settings are obtained.
I'm speaking from experience when I say this. It's "baby steps" and "one day at a time".
Den
.
thxs Den.....
-
robysue
- Posts: 7520
- Joined: Sat Sep 18, 2010 2:30 pm
- Location: Buffalo, NY
-
Contact:
Post
by robysue » Thu Nov 07, 2013 9:34 pm
Also please keep in mind that it's hard to say anything after only one night at a particular setting.
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Thu Nov 07, 2013 9:37 pm
robysue wrote:Also please keep in mind that it's hard to say anything after only one night at a particular setting.
I will be a week on Saturday
-
darco
- Posts: 65
- Joined: Mon Sep 02, 2013 10:56 am
- Location: Bay Area
Post
by darco » Sun Nov 10, 2013 10:29 am
Well as Den mentioned, I am "dialed in" and I am feeling slightly more refreshed each day. I do believe in the so called "sleep debt" and feel its just a matter of time.
Also I switched over last nite to my original mask, the Zest Nasal just to compare leaks that I may be getting with my Swift. Now I remember why I switched mask, uggh the discomfort under the nose. I slept fine with it but will probably have this pain around all day...I see a post on the Nano..hmmm
Zest Nasal