Did I just screw everything up? Help me please! ASV unit

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Did I just screw everything up? Help me please! ASV unit

Post by sleepinow » Sat Jul 27, 2013 2:51 am

Okay so tonight I wanted to try something new. Since my previous setting 5.0 did not tackle all of my hypopneas to the point that I liked it, i wanted to raise the pressure by just 0.5. After doing so and trying to fall asleep I went into a state of panic and didnt know why. (It was probably because of the new setting)

Falling asleep was hard and now I am having trouble once again staying asleep.

Question is, does changing a setting on an ASV unit by even as little as .5 cause such a state of craziness?

Oh and what scares me more is that after looking on my machine after waking up, the darn pressure displayed a 4.5. How is that even possible!?

Will this new setting take 3 goddamn months to get used to? Or maybe around a week? At first I didnt even think I would need to really adjust. But now I am pissed off so should I stick with it or go back to my previous setting

(No one told me changing a pressure even gradually can be such a full time ultra stressful job)

Oh I was going to say sorry for the bad language. But screw it, bad language is needed at times like these. I may sound like a total jerk though. And for that I am sorry

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
Last edited by sleepinow on Sat Jul 27, 2013 6:06 am, edited 1 time in total.
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Re: Did I just screw everything up? Help me! ASV unit

Post by sleepinow » Sat Jul 27, 2013 4:28 am

Okay now I am very concerned, I do not know what to do, someone please help me

How long does it take to adapt to a pressure change on a asv unit

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Did I just screw everything up? Help me! ASV unit

Post by zoocrewphoto » Sat Jul 27, 2013 4:46 am

You probably won't get any useful replies until tomorrow sometime.

There are very few people on this forum who use an asv machine. It was mentioned on another topic that ASV machines take longer to get used to than normal machines, so those of us who have never used one can't really give you a helpful answer.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by sleepinow » Sat Jul 27, 2013 9:25 am

Bump

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

SleepyToo2
Posts: 1005
Joined: Sun Sep 11, 2011 7:55 am
Location: North of Philadelphia, PA

Re: Did I just screw everything up? Help me please! ASV unit

Post by SleepyToo2 » Sat Jul 27, 2013 10:26 am

I did not sleep with an ASV last night because I don't own one. However, I do know that there are many variables in any xpap treatment that make it impossible to tell from one night whether you messed things up or if it is any one of those other factors. The 1st rule of "fiddling" is to make the change very small, and if it doesn't work out go back to the original setting. You will then know one thing that probably doesn't work. There is absolutely no need to panic. Of course that assumes you can remember how to reverse your change! If you can't, post back here but be patient in expecting a response. Especially at the weekend.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Use SleepyHead software.
Not a medical professional - just a patient who has done a lot of reading

User avatar
2 B Sleeping Soundly
Posts: 822
Joined: Sat Nov 05, 2011 2:24 pm
Location: Southern California

Re: Did I just screw everything up? Help me! ASV unit

Post by 2 B Sleeping Soundly » Sat Jul 27, 2013 10:35 am

sleepinow wrote:Okay now I am very concerned, I do not know what to do, someone please help me

How long does it take to adapt to a pressure change on a asv unit


You might want to PM John B Fisher as he is very experienced with ASV machines and the difficult adjustments for those who are using them.
memberlist.php?mode=viewprofile&u=44411

Good luck, hang in there and don't give up!!

John

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: 11-09-11 Start PAP therapy. Current settings: APAP 13cmH2O - 17cmH2O / Ramp, off / A-Flex, off / Respironics Premium Chin Strap / Sleepyhead Software
One time a cop pulled me over for running a stop sign. He said, "Didn't you see the stop sign?" I said, "Yeah, but I don't believe everything I read." -- Steven Wright

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Did I just screw everything up? Help me please! ASV unit

Post by NateS » Sat Jul 27, 2013 10:52 am

sleepinow,

I'll try to fill in until the real experts get here.

This happened to me once with my ResMed ASV and it turned out that I had inadvertently flipped my mode from ASV to CPAP. Does your Respironics ASV unit have modes too? Check and see if that might be the problem.

Also, you say you meant to add .05 to 5.0 but now it is set to 4.5 - if so, could you have simply reduced your pressure when you meant to increase it? Did you feel like you were suffocating from inadequate air delivery?

Best wishes, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by sleepinow » Sat Jul 27, 2013 10:59 am

NateS wrote:sleepinow,

I'll try to fill in until the real experts get here.

This happened to me once with my ResMed ASV and it turned out that I had inadvertently flipped my mode from ASV to CPAP. Does your Respironics ASV unit have modes too? Check and see if that might be the problem.

Also, you say you meant to add .05 to 5.0 but now it is set to 4.5 - if so, could you have simply reduced your pressure when you meant to increase it? Did you feel like you were suffocating from inadequate air delivery?

Best wishes, Nate
I did not reduce the pressure because after turning it off and then turning it back on it went to 5.5. And yes I did feel like I was suffocating. It is also the reason why I increased pressure by .5

And I do not believe I changed it to cpap because when I did stop breathing I felt the surge of air come in which actually woke me back up this morning

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by JohnBFisher » Sat Jul 27, 2013 12:07 pm

Greetings .. If bumping it up by .5 caused so much problem .. try setting it back to 5.0 from 5.5. However, I suspect this one was more an anxiety issue than a real issue .. you worried it would cause problems so you had problems falling asleep.

Alternatively, you might want to start the ramp at 5.0 and have the ramp last for 30 minutes or so ... Hope that helps.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by sleepinow » Sat Jul 27, 2013 1:46 pm

I will give it a few days. About how lobg does it take to get used to a different pressure. I am sure that it usually is not instantly and since I am still relatively a beginner user of asv

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

User avatar
RogerSC
Posts: 1919
Joined: Sat Jul 21, 2012 1:11 pm

Re: Did I just screw everything up? Help me please! ASV unit

Post by RogerSC » Sat Jul 27, 2013 2:07 pm

I'm pretty sure that this depends on you...I've made small changes in the pressure on my machine many times (it's an apap, not an asv though), and lost no sleep over it *smile*, so it really depends on you. Just give it time, and don't stress over it, just go about your normal life and sleeping routine, and you'll get used to the change.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Additional Comments: ResMed AirFit N30 mask

User avatar
pendulumbob
Posts: 7
Joined: Tue May 28, 2013 12:26 pm

Re: Did I just screw everything up? Help me please! ASV unit

Post by pendulumbob » Sat Jul 27, 2013 2:57 pm

What setting of 5 did you change?

_________________
MaskHumidifier
Additional Comments: S9 ASV Min 7 Max 10

sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by sleepinow » Sat Jul 27, 2013 3:20 pm

EPAP why? and yeah i try to remain calm. Its hard sometimes especially last night

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by JohnBFisher » Sat Jul 27, 2013 4:15 pm

Like I said, you can use the ramp feature to make it easier to adjust to the pressure change. Set the starting pressure for the ramp at your old setting .. that way you won't notice any change. By the time the pressure changes, you should already be asleep.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

sleepinow
Posts: 470
Joined: Thu Apr 25, 2013 9:40 am

Re: Did I just screw everything up? Help me please! ASV unit

Post by sleepinow » Sat Jul 27, 2013 4:57 pm

Great thanks for the help John. I hope good things come your way

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5