Help. Traveling and got my machine wet
Help. Traveling and got my machine wet
I'm traveling with my family and packed my REMstar A-Flex half full of water and am trying to use it. I cleaned and dried it as best as I could. It blows out hard but that's all. It does not regulate the air pressure so I can exhale. No tools are available. Anyone with knowledge of this would be much appreciated. Thanks
- Drowsy Dancer
- Posts: 1271
- Joined: Mon Feb 28, 2011 9:35 am
- Location: here
Re: Help. Traveling and got my machine wet
Cut and pasted from cpap.com (our host):
So what should be done if water is accidentally poured into a machine? Don't be sad, you can often get it working. Here is what to do:
Assuming the wall socket area is dry, unplug the machine from all power. (Caution! never contact a wet area when power is present.)
Take the machine, hose, and mask to a sink. Empty all water from the hoses into the sink. Pour as much water out the machine as possible. Turn it every which way that allows the water to flow out of any openings.
Service the filter(s). They are often wet and need to be dried or replaced.
Completely dry the outside of the machine. Be sure the plug in area is dry. It is going to be plugged in to power, so be absolutely certain no water is on the outside of the unit.
If a hair dryer is available, blow medium heat at a maximum flow setting into the machine air inlet. Take the filters out for this! After 15 minutes, blow air into the outlet for 15 more minutes. This should dry your machine out nicely.
With a towel ready to wipe up any water blown out the main hose connection, plug the machine in and turn it on. Yes, really. Do not wear the mask as water may be blown out. Nine times out of ten, the machine will fire up and blow water out. They may not run normally at first as remaining water affects performance. Most model will run fine without the mask on a face. Let the machine run for a few hours, This will completely dry out the inside of the machine on most models. You have saved the machine from serious damage. After the three hours, breath on it a bit. If it does not feel the same, it will need to go in for service.
Some machines will not run properly when there is no person at the end of the mask. It is very difficult to get the water out of these units. Either use a hair dryer for a hour or more or send it to a dealer for repair. If less than a third of a cup of water entered the unit, it will likely be fine. If it took more, you may be forced to get it serviced.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears |
| Additional Comments: Software: SleepyHead. Pressure: APAP 9.5 min/11 max, A-Flex x2 |
How we squander our hours of pain. -- Rilke
Re: Help. Traveling and got my machine wet
Thank You!
Re: Help. Traveling and got my machine wet
Before using a hair dryer I would make sure that the air temperature is not above 140 degrees F (60 degrees C), b/c the technical specks for my S9 Autoset specify it under Storage and Transport Temp. Also, before plugging the unit to electricity I would "bake" it in an oven for an hour, or more, at the above temperature. Or cover the unit with a towel and put it in the sun for a day.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6 |
Re: Help. Traveling and got my machine wet
Most kitchen ovens don't have temperature controls that will go as low as 141 degrees F.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
jeff
Re: Help. Traveling and got my machine wet
Mine fell off the nightstand once and got wet. My DME dried it and ran it for a while and it ended up being fine. Not sure what she did to dry it out, though.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: Pressure 18/13. |
- StuUnderPressure
- Posts: 1378
- Joined: Mon Jun 18, 2012 10:34 am
- Location: USA
Re: Help. Traveling and got my machine wet
Do you have the link to that document?Drowsy Dancer wrote:Cut and pasted from cpap.com (our host):
I would like to print it out to keep in my machine bag while traveling.
EDITED to add that I found it.
It is in the Newsletter 7 at https://www.cpap.com/DisplayNewsletter/ ... sue-7.html
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap |