Help - Respironics System One - Noise?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
qwerty3656
Posts: 3
Joined: Wed Feb 27, 2013 11:17 am

Help - Respironics System One - Noise?

Post by qwerty3656 » Wed Feb 27, 2013 11:25 am

Hi New to the Forum. - Just bought the Philips Respironics System One. When the A-flex or C-flex kick in (everytime I exhale), there is a humming/whining noise - changing pitch as it reduces and then increases air pressure. It is not that loud, but in a quiet room when you are trying to get to sleep, it is very noticeable. I was wondering if this in what others have experienced, or do I need to return it? Thanks for any assistance.
Regards;


Jim

cosmo
Posts: 1371
Joined: Fri Nov 02, 2012 1:15 pm

Re: Help - Respironics System One - Noise?

Post by cosmo » Wed Feb 27, 2013 11:28 am

Whats your pressure and mask? Minus well fill in your equipment in the profile settings.

qwerty3656
Posts: 3
Joined: Wed Feb 27, 2013 11:17 am

Re: Help - Respironics System One - Noise?

Post by qwerty3656 » Wed Feb 27, 2013 11:44 am

Thanks - I found the equipment section and think I added it correctly.

My pressure is 12 and I have most recently been using a full-faced Mirage Quattro. When I am not congested I try to use a Comfortgel nasal mask. I don't use the humidifier. I have had a cpap machine for 10-years or so, but recently upgraded the machine. I've had machines die on me in the past, and I don't want to keep a "lemon". Maybe it is normal, I just was not expecting the low-level whining.

Jim

acidragon
Posts: 119
Joined: Sun Feb 03, 2013 10:08 pm

Re: Help - Respironics System One - Noise?

Post by acidragon » Wed Feb 27, 2013 2:07 pm

I just had the same problem but mine was on the inhale. I ended up getting a new machine and its a lot lower now. However, my DME said if anythign the noise should be on the exhale not the inhale. Hope this helps.

_________________
Machine

cosmo
Posts: 1371
Joined: Fri Nov 02, 2012 1:15 pm

Re: Help - Respironics System One - Noise?

Post by cosmo » Wed Feb 27, 2013 2:13 pm

I say give it another day or two to let the machine settle.

I'll see about testing my PR1 machine with the same mask. My setting is also at 12cm and I'll try to record a video clip to capture the noise.

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Help - Respironics System One - Noise?

Post by khauser » Thu Feb 28, 2013 11:38 am

acidragon wrote:I just had the same problem but mine was on the inhale. I ended up getting a new machine and its a lot lower now. However, my DME said if anythign the noise should be on the exhale not the inhale. Hope this helps.
I disagree, and that is not my experience. With AFlex or CFlex enabled, the machine is sending less pressure during exhale, and is in my opinion quieter. As soon as I begin to draw a breath it spins up. This is audible but not disturbing.

Also, when I first get in bed and mask on I am breathing somewhat heavier than when I am close to falling asleep, so it starts off which a large difference and bit of noise, and as I settle so does the machine.

Since I've been on xPAP for so long, I find the noise quite comforting. It "sings" me to sleep. (Yeah, OK, that was a bit much)

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

User avatar
Tigerlily
Posts: 341
Joined: Thu Feb 21, 2013 7:53 pm

Re: Help - Respironics System One - Noise?

Post by Tigerlily » Thu Feb 28, 2013 12:35 pm

I agree with the OP. Everyone said in reviews how quiet the 560 is.

I think mine is louder than the 150 model I had. It certainly isn't quieter. Also, when in FLEX mode it does make more noise on the exhale.

Not that it bothers me as I usually wear earplugs, but I expected it to be virtually silent from the reviews I read, and it's not.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressure: 11 , OSCAR software
~ Tigerlily ~

User avatar
Pugsy
Posts: 65374
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Help - Respironics System One - Noise?

Post by Pugsy » Thu Feb 28, 2013 12:45 pm

Noise level perception seems to also vary with people.
I hear my PR S1 machine mainly on inhale...the ResMed S9...mainly on exhale but my husband says the PR S1 he can't even tell it is on.
Some of what we "hear" is conducted noise from the machine through the hose or mattress.
I can plug my ears with my fingers...sound doesn't diminish at all...so I know it is the "noise" that is conducted through the hose that I am hearing more than the actual machine noise.

So there's vent noise from the mask...there's machine noise that may vary if exhale relief is used...there's conductive noise and finally just simple machine noise because machines just have a hard time being totally silent and some people are more sensitive to sounds than others.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

spitfire

Re: Help - Respironics System One - Noise?

Post by spitfire » Thu Jul 11, 2013 3:45 pm

Hello-

I recently upgraded my PRS1 from model 550 to model 560. Was really happy with the 550 for the past two years and wanted to get something a little newer with my insurance money. However, I also noticed the 'whirring' noise on the 560. My pressure is 14 and I use the Opus 360 mask as well as a newer Wisp.

At first I thought the humming was something rattling on my night table or in the mask. However, I put the 550 and 560 side by side, in a quiet room, and with the exact same settings (A-Flex, etc) it is very apparent that the new 560 is 'louder' on inhale/exhale than the older 550.

I am talking with my DME supplier about swapping the 560 for something else, considering Resmed S9. I'm curious about the poster who said it takes time for the machine to settle. It'd be nice if this is the case. My 550 is 2.5 years old but I never remember the 'noise.' The 560 has only been used for 15 nights. I'm going to continue to use it for another week while I work out an equipment swap.

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Help - Respironics System One - Noise?

Post by greatunclebill » Thu Jul 11, 2013 4:16 pm

check the total hours on your new machine. maybe your new machine isn't really new. maybe its really old and used alot.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Help - Respironics System One - Noise?

Post by jdm2857 » Thu Jul 11, 2013 4:46 pm

qwerty --

Even with the exhale noise you describe, the 560 should be quieter than just
about any older generation machine.

Do you find that the 560 is louder than the machine it replaced? And, if
so, what older machine did you have?
jeff

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Help - Respironics System One - Noise?

Post by Janknitz » Thu Jul 11, 2013 4:48 pm

I found that my machine is much quieter with flex at its lowest setting, and the setting doesn't bother me at all. After 3 years I have to think about it to hear my machine. I suppose it's because I'm used to it, not that the machine changed at all.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

bondoj

Re: Help - Respironics System One - Noise?

Post by bondoj » Thu Aug 27, 2015 6:59 am

Late response but I have had the same noise problem.(high pitched noise)Machine was silent at first but after awhile it started to make noise. Have replaced 3 machines .I now know noise is caused by myself overfilling water tank. When tank is overfilled water I believe will enter machine where motor is located and over time will deteriorate and or wash away any lubricate and or cause motor to labour and cause noise and become over time too load to sleep. Take water tank and see how high you can fill with water before it will overfill and be sucked into fan side of machine.

marklafan
Posts: 3
Joined: Fri Sep 25, 2015 8:49 pm

Re: Help - Respironics System One - Noise?

Post by marklafan » Fri Sep 25, 2015 9:02 pm

I have returned 2 "Phillips Remstar Auto A-Flex" machines in the past 6 months for this problem. they always start out fine for the first 2 moths or so. on returning the second machine they claimed that the whiny high pitch noise is due to the fact that the pressure of my exhale was stronger then the pressure of the machine thereby causing the motor spin the opposite direction. this was burning out the motor?!? I'm not sure of how valid this is but thought maybe others have heard the same thing or know how to prevent this noise?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Help - Respironics System One - Noise?

Post by Wulfman... » Fri Sep 25, 2015 9:11 pm

marklafan wrote:I have returned 2 "Phillips Remstar Auto A-Flex" machines in the past 6 months for this problem. they always start out fine for the first 2 moths or so. on returning the second machine they claimed that the whiny high pitch noise is due to the fact that the pressure of my exhale was stronger then the pressure of the machine thereby causing the motor spin the opposite direction. this was burning out the motor?!? I'm not sure of how valid this is but thought maybe others have heard the same thing or know how to prevent this noise?
If you're using a "Flex" setting, try turning it off.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow