Bipap needs new plug

Post by Wayne02 » Wed Feb 20, 2013 1:19 pm

Hello,

A relative uses a Respironics bipap auto with bi-flex, last design in this series, 5 button. We realize the machine is old and that insurance will partially pay for a new machine every X years, but we want this one fixed to keep as a backup.

The plug on the end of the 24vdc switching power supply has worn and takes some wiggling to get it to contact and power up the machine each time the plug is removed/installed into the machine. One of these days it will fail to make contact completely. The receptacle on the machine itself looks to be in good order.

A complete new power supply is $160.

Any idea where I can source such a plug? It kind of looks like some sort of din plug but I've not seen that pin arrangement.

Thanks

Image

Image

User avatar
Pugsy
Posts: 65194
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Bipap needs new plug

Post by Pugsy » Wed Feb 20, 2013 3:15 pm

Would that be this plug you are needing? For the Legacy series (pre M series bipap)?
https://www.cpap.com/productpage/respir ... -auto.html

This place repairs machines...maybe they know of a place to get one that doesn't cost so much.
http://www.acbio.com/

You know you can often find these old Legacy series machines on craigslist pretty cheap. Might be able to buy a machine and get the power cord that comes with it for less that the online prices for just the power supply.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Wayne02
Posts: 21
Joined: Sat Jun 23, 2007 1:33 pm
Location: Western WA

Re: Bipap needs new plug

Post by Wayne02 » Wed Feb 20, 2013 5:37 pm

Pugsy wrote:Would that be this plug you are needing? For the Legacy series (pre M series bipap)?
https://www.cpap.com/productpage/respir ... -auto.html

This place repairs machines...maybe they know of a place to get one that doesn't cost so much.
http://www.acbio.com/

You know you can often find these old Legacy series machines on craigslist pretty cheap. Might be able to buy a machine and get the power cord that comes with it for less that the online prices for just the power supply.
Yep, that's the one. Good idea on craigslist, I'll take a look.

Thanks

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Bipap needs new plug

Post by JDS74 » Wed Feb 20, 2013 5:45 pm

It may be that the problem is in the cable itself and not the plug. Perhaps the wire has cracked and that is what is being moved when you wiggle the plug in the machine.

If you have a volt-ohm meter, you can check to see if the proper voltage is present when the power transformer is plugged into the wall but not the CPAP machine.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Bipap needs new plug

Post by archangle » Wed Feb 20, 2013 5:53 pm

A thorough cleaning, some judicious bending of the parts of the plug, and maybe some DeoxIT gold might help.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.