Adjusting my REMstar pro2 w/C-flex myself
-
HeavyG
Adjusting my REMstar pro2 w/C-flex myself
I've had this machine for 8-9 years, recently lost over 160 lbs. I've been having to tighten my mask so tight it's giving me headaches. Had a sleep test September last year. The tech said my pressure is OK I need to try a new mask, I've tried 5 different ones full face, nose only, nasal. I'm fine with the mask i'm using I just want to have the pressure turned down a few notches. i'm currently at 14. I'd like to just try to adjust it to see if it helps. The noise and the pressure are to much and i'm tired of arguing with the tech about it. Can anyone help me with this?
-
Wulfman...
Re: Adjusting my REMstar pro2 w/C-flex myself
You have my favorite machine. (I have several of them)
Here are the instructions I wrote up years ago for getting into the setup and the options.
Den
REMstar Pro 2 Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.
a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >
c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >
d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
Select next setting with >
f. Ramp pressure setting: Select with ^ or v
Select next setting with >
g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
.
Here are the instructions I wrote up years ago for getting into the setup and the options.
Den
REMstar Pro 2 Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice.
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.
a. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >
c. CPAP or CFLE pressure setting: Select with ^ or v
Select next setting with >
d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0:00)
Select next setting with >
f. Ramp pressure setting: Select with ^ or v
Select next setting with >
g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
.
-
HeavyG
Re: Adjusting my REMstar pro2 w/C-flex myself
Thanks for your help. I'll give it a try. And yes I was lucky to get this machine, it's my third and by far the best machine I've ever used. Thanks again.