humidifier issue
humidifier issue
Hello, I have a new...er machine...been working fine but lately I wake often with a dry mouth..the H5i glows orangish and I have to refill it every evening ...this waking up dry is getting annoying...any suggestions? I tried to read the instructions. Either I am foggy as hell or they are just confusing ...or both. any suggestions ideas ...anything???...lost my insurance so I cannot get much help from...whoever. thankes...
Re: humidifier issue
Hello & welcome. To help us best help you, please fill in your machine & mask type by going to the User Control Panel. How old is the machine?
Sounds like you may have a mechanical problem, but lets see what other forum members may think. My hubby just got this type of humidifier and I've never seen any orange glow. I would think the machine should still be covered by warranty.
The DME who sold you the machine would be the starting point for getting it fixed, insurance should not come into play if it is under warrantly or was defective when it was sold to you.
Good luck,
Jen
Sounds like you may have a mechanical problem, but lets see what other forum members may think. My hubby just got this type of humidifier and I've never seen any orange glow. I would think the machine should still be covered by warranty.
The DME who sold you the machine would be the starting point for getting it fixed, insurance should not come into play if it is under warrantly or was defective when it was sold to you.
Good luck,
Jen
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Pressure Auto 12-20cm CPAP compliant since 2000 |
Other masks I've tried: *=liked, #= no way
Piliaro, SleepWeaver Elan*, Swift FX w&w/o Bella Loops#, OpitLife#,Simplicity*, Mirage Vista*, Go Life for Her#, IQ (original hg only)*, Quattro FX (barely)###, Wisp*, Nuance#, Swift LT for her**
Piliaro, SleepWeaver Elan*, Swift FX w&w/o Bella Loops#, OpitLife#,Simplicity*, Mirage Vista*, Go Life for Her#, IQ (original hg only)*, Quattro FX (barely)###, Wisp*, Nuance#, Swift LT for her**
-
raptor5150
- Posts: 22
- Joined: Wed Nov 21, 2012 11:56 pm
- Location: O'fallon MO.
Re: humidifier issue
yes the h5i does glow redish/orange when its preheating, mine does.
_________________
| Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: PR System One (560) Remstar Auto with A-FLEX |
Re: humidifier issue
Is the orange glow when you turn the unit on or when you wake up? Are you in a very cold area where the air is very dry? Do you have the humidifier set very high?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
-
raptor5150
- Posts: 22
- Joined: Wed Nov 21, 2012 11:56 pm
- Location: O'fallon MO.
Re: humidifier issue
Sorry, should have been more clear, its a light thats on when you preheat then it goes out.
_________________
| Mask: Mirage™ FX Nasal CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: PR System One (560) Remstar Auto with A-FLEX |
- wildflowertx
- Posts: 98
- Joined: Wed Nov 07, 2012 1:17 am
- Location: North Texas
Re: humidifier issue
Huh. I don't do the preheat and mine comes on orange sometimes; I always thought it meant low water. My humidifier light comes on -- either blue or orange -- at night when I'm putting the SD card back in just before bedtime, and then again when I actually turn the machine on. But I never actually *looked* to see what the colors mean.
_________________
| Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Heated hose; CMS-50F pulse oximeter; Zeo Bedside Sleep Manager; also Swift FX For Her mask |
Steph
Treatment started 11/27/12
Treatment started 11/27/12
Re: humidifier issue
i am not getting any help now...when I had insurance..yeah but not now....if its full its not too bad ..turned up all the way, and it is not dry here at all...damn I took it off last night ...not sure why...I hate this crap...almost rather just not wear it and feel like crap. and yeah I know the outcome possibility
Re: humidifier issue
Classic symptoms of mouth leakage. Search for topics on mouth leakage. You may have to resort to a mask that covers your mouth or tape your mouth to keep lips from blowing open.YELOOCD wrote:Hello, I have a new...er machine...been working fine but lately I wake often with a dry mouth..the H5i glows orangish and I have to refill it every evening ...this waking up dry is getting annoying...any suggestions? I tried to read the instructions. Either I am foggy as hell or they are just confusing ...or both. any suggestions ideas ...anything???...lost my insurance so I cannot get much help from...whoever. thankes...
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |
- StuUnderPressure
- Posts: 1378
- Joined: Mon Jun 18, 2012 10:34 am
- Location: USA
Re: humidifier issue
The color of the light that comes on in the ResMed S9 (whether you turn it on by pressing down the round button or it comes on automatically) means:
Orange - Humidifier is heating up, is already heated up, or is still in its cooldown mode.
Blue - Humidifier has already cooled down or was never heated up.
That light (no matter what color) will go off by itself automatically after a short period of tiem.
Orange - Humidifier is heating up, is already heated up, or is still in its cooldown mode.
Blue - Humidifier has already cooled down or was never heated up.
That light (no matter what color) will go off by itself automatically after a short period of tiem.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap |
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
Re: humidifier issue
The display turns off after a while when you're using the machine if you don't touch the knobs. It comes back on when you turn the machine off.
The humidifier light turns on and off with the display light.
The humidifier light is orange if it's heating, or if it's cooling down after you turn the machine off. It's blue if the humidifier heat is off. It will normally be orange for a while after you turn the machine off. It doesn't matter if the tank is full or not.
The OP has a full face mask, so mouth leaks shouldn't be a problem.
You could have leaks around the mask increasing your water usage. Get the SleepyHead software and check to see if you're having air leaks during the night. Also, disconnect the hose from your mask, plug up the end of the hose, turn the machine on, and listen for leaks somewhere in the system.
Other than that, setting the humidity high or having dry room air will increase usage. Even if it doesn't feel humid, humidity is lower when the air is cold. You'll also use more water the higher your pressure is because the mask vents more air at higher pressure.
Unless you have excessive air leaks or a water leak, the only place the water can go is into your mask and then out through the vent like it's supposed to. If you need the amount of humidity you're getting there's no way for the humidifier to "lose" the water.
The humidifier light turns on and off with the display light.
The humidifier light is orange if it's heating, or if it's cooling down after you turn the machine off. It's blue if the humidifier heat is off. It will normally be orange for a while after you turn the machine off. It doesn't matter if the tank is full or not.
The OP has a full face mask, so mouth leaks shouldn't be a problem.
You could have leaks around the mask increasing your water usage. Get the SleepyHead software and check to see if you're having air leaks during the night. Also, disconnect the hose from your mask, plug up the end of the hose, turn the machine on, and listen for leaks somewhere in the system.
Other than that, setting the humidity high or having dry room air will increase usage. Even if it doesn't feel humid, humidity is lower when the air is cold. You'll also use more water the higher your pressure is because the mask vents more air at higher pressure.
Unless you have excessive air leaks or a water leak, the only place the water can go is into your mask and then out through the vent like it's supposed to. If you need the amount of humidity you're getting there's no way for the humidifier to "lose" the water.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
- Denial Dave
- Posts: 1334
- Joined: Wed Sep 19, 2012 6:45 am
- Location: Connecticut
Re: humidifier issue
I alternate between the 2 different tanks used on the H5i... 5 days a week, I use the dishwasher safe tank. But when it is in the dishwasher, I use the disposable/ non opening tank.
my experience is that I use more water when I use the tank that can't be opened... the bottom of it is lighter and tends to heat up more that the heavier dishwasher safe humidity tank.
my experience is that I use more water when I use the tank that can't be opened... the bottom of it is lighter and tends to heat up more that the heavier dishwasher safe humidity tank.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Started at VPAP pressure setting of 20/14.4. I've survived Stage 1 cancer and lost 80+ lbs. Pressure is now 14.5 / 11 |
if you don't know where you are going... any road will take you there.... George Harrison
-
helpmesleep66
- Posts: 3
- Joined: Sun Jan 13, 2013 2:34 pm
Re: humidifier issue
I occasionally have the same problem. After about 8 years I went to an ENT doctor and now can easily breathe through my nose. If I have a dry mouth, is is usually because my full face mask is leaking air or my nose is stopped up. It has been 10 years now and Iim finally getting better rest. Because of age and weight, I still wake up 3-5 times a night unless I use a sleep aid.
-
bjcyorkiemom
- Posts: 33
- Joined: Sat Dec 29, 2012 12:50 pm
- Location: Kansas
Re: humidifier issue
I have the H5i and was having the same problem. First, the lights turn on when you mess with the machine. Blue means your decision has been locked in, if you push the "turning knob" it will turn the light orange, and lets you adjust the heat level. I was going through almost a full tank a night on a setting of 5.5, with no mouth leaks, but I was still dry, so I got the Climateline Max heated tube and everything is working much better for me. Have had the tube for almost 2 weeks.
_________________
| Mask | Humidifier | |||
![]() | ![]() | |||
| Additional Comments: sized (not adjustable) Ruby style chin strap, Climateline Max tubing, NeilMed saline nasal rinse, RoEzIt nasal mositurizer | ||||
Integrity - is what / how you do when no one is looking...