ASV settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
All in Knots

ASV settings

Post by All in Knots » Mon Nov 12, 2012 7:56 am

Hi,

I've been a member for awhile, but haven't been active on the board. (I'm having trouble logging in. Whenever I hit "login," the page refreshes and doesn't take me to the "enter login/password" page. This is the only site that I have trouble with. I think it might have to do with my security setttings??)

Anyways, here's my real question. I had two recent sleep study tests (within the few months) that determined that I needed an ASV machine/adapt sv machine. I received my new S9 Resmed ASV a few days ago. I was told that the settings were the "basic" settings, and the machine was set by my DME as per my sleep doc's prescription.

The machine seems to want me to breather faster than I breathe at night/when relaxed. Before I'm finished exhaling, the machine starts to whoosh in new air. In other words, I think the rate is too fast. The new air comes in about a second too quickly, and my lungs are still busy exhaling. I'm pretty confident that the machine is not adjusting for this, because I feel this as I'm going to sleep and as I'm waking up in the morning.

Is this something I can easily tweak on my own, or do I need to call my sleep doc for a new rx? (Thx to this board, I knew how to tweak some settings on my S8 apap machine, but the adapt sv seems more complicated.) And if I call my sleep doc about this, will she recommend yet another sleep study test because of this??

Thanks,
All in Knots

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: ASV settings

Post by Sheriff Buford » Mon Nov 12, 2012 9:21 am

Try the EPR mode. It is designed to make your therapy more comfortable. EPR maintains treatment for the patient during inhalation and reduces the delivered mask pressure during exhalation. You can use EPR in either CPAP or AutoSet mode. The EPR setting on the S9 can be configured on the off, 1, 2, or 3 cm H2O. Do you have the clinical manual that shows you how to adjust the settings? If you do not, PM me with your email address and I will email you a copy.

Sheriff

All in Knots

Re: ASV settings

Post by All in Knots » Mon Nov 12, 2012 11:03 am

Hi, Sheriff!

I see that you have the S9 Autoset machine (apap). I have the ASV/adaptive servo-ventilator machine. I'm not sure if there's an EPR on that one or not (?). It's designed to treat mixed apneas, and seems extra-eager to begin shooting that air to you. I'd like it to begin shooting the air about 1 second later than it does. Does your manual explain how to do that? I have this awful feeling that I prob have to call the dr., request a settings change, wait, then have the DME RT come out and adjust...

Any ideas?

Thanks,
All in Knots

User avatar
Pugsy
Posts: 65241
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV settings

Post by Pugsy » Mon Nov 12, 2012 11:15 am

I don't believe the Adapt has EPR.
Hold on till this evening and someone with experience with your type of machine should see your post.
If you don't have the provider/clinician manual for your machine you can request it here.
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: ASV settings

Post by Madalot » Mon Nov 12, 2012 11:18 am

I had the same problem on my Trilogy Ventilator. It sounds like the Breath Rate/RR IS a little too fast for you. In order for me to be comfortable, I had to set my Breath Rate at 7. They said that they typically set it at 10 as a mid point. Anything higher for me makes me feel exactly like you describe.

I don't know anything about your specific machine, but hopefully someone can come and help you adjust that. I'm almost positive there is a setting for it.

Edited to add: John Fisher has posted below me and said there is NOT an option on the Resmed machine. I'm sure John is right since he's much more knowledgeable than I am. Sorry about that.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7
Last edited by Madalot on Mon Nov 12, 2012 11:41 am, edited 2 times in total.

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: ASV settings

Post by JohnBFisher » Mon Nov 12, 2012 11:38 am

If you provide (via PM or on a post) your email address, I can provide a copy of the clinician's manual for the S9 VPAP ASV unit.

However, the short answer is that the S9 VPAP ASV unit does not have a parameter to adjust the Respiratory Rate. The Respironics S1 ASV unit does have that as an option. The Resmed unit is based on your last three minutes as a moving average. So, if your breathing slows as your start to fall asleep, it will intervene to help sustain your rate to the average of the past three minutes.

This is one of those things that can make it difficult to become acclimated to an ASV unit.

Two suggestions. First, if this really causes you a problem, then ask if a Respironics unit would be an option. That unit allows the Respiratory Rate (RR) to be adjusted. Second, learn to start breathing slowly when you put on the mask. You will eventually adjust to the unit (in about 2 or 3 months).

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ASV settings

Post by SleepingUgly » Mon Nov 12, 2012 11:59 am

This is a long shot, but... When you get into the clinical menu (pushing the button with the check mark and the big round button to it's right simultaneously), is the AB Filter set to "on"? It should not be set to "on" unless you're actually using an AB filter. The AB filter is not the filter you have on the back of the machine, but rather is something like this:

https://www.cpap.com/productpage/respir ... -pack.html
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

Taringa542
Posts: 207
Joined: Mon Jul 16, 2012 7:26 pm
Location: Bundaberg QLD Australia

Re: ASV settings

Post by Taringa542 » Mon Nov 12, 2012 1:57 pm

All in Knots wrote:Hi,

I've been a member for awhile, but haven't been active on the board. (I'm having trouble logging in. Whenever I hit "login," the page refreshes and doesn't take me to the "enter login/password" page. This is the only site that I have trouble with. I think it might have to do with my security setttings??)

Anyways, here's my real question. I had two recent sleep study tests (within the few months) that determined that I needed an ASV machine/adapt sv machine. I received my new S9 Resmed ASV a few days ago. I was told that the settings were the "basic" settings, and the machine was set by my DME as per my sleep doc's prescription.

The machine seems to want me to breather faster than I breathe at night/when relaxed. Before I'm finished exhaling, the machine starts to whoosh in new air. In other words, I think the rate is too fast. The new air comes in about a second too quickly, and my lungs are still busy exhaling. I'm pretty confident that the machine is not adjusting for this, because I feel this as I'm going to sleep and as I'm waking up in the morning.

Is this something I can easily tweak on my own, or do I need to call my sleep doc for a new rx? (Thx to this board, I knew how to tweak some settings on my S8 apap machine, but the adapt sv seems more complicated.) And if I call my sleep doc about this, will she recommend yet another sleep study test because of this??

Thanks,
All in Knots
The Adapt has Pressure Support, shows as PS min & PS Max, If memory serves me correctly the default PS are 3 min & 15 max, try upping the PS min to 5

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 VPAP Adapt 36037. Rescan 4.3 & Sleepyhead 9.2 (Mac)

All in Knots

Re: ASV settings

Post by All in Knots » Mon Nov 12, 2012 2:33 pm

You guys & gals are the best!

I checked my machine, and the AB filter was set to "NO," so that wasn't the prob, but it was worth a try!

Checked the pressure supports. Sure enough, it was set to min. 3, max 15. I reset it to min. 5 (kept max @ 15). I briefly put on the mask, and voila! The new whoosh of air began *after* exhalation, instead of near the end of the inhalation. That feels much better!!

I have no idea why it was set to 3 to begin with? Perhaps that would be what I'd need if I were jogging with the unit on <tee hee>. Heck, it's almost a pant!

I hope this works for me. Will try this new setting tonight.

Thx again everyone,
All in Knots

User avatar
Xney
Posts: 842
Joined: Fri Jul 20, 2012 6:53 pm

Re: ASV settings

Post by Xney » Tue Nov 13, 2012 1:17 am

The S9 vpap adapt doesn't let you set an EPR or BPM. What I do so it doesn't freak out is just breathe evenly yet not too quickly or slowly when it's starting to get its timing down. If you breathe really slowly, like a calm awake breathing, it seems to make the machine trigger faster breaths. But if you're just a bit slower than it likes, it seems to accept that.

Don't ask me why, just my experience.

So basically as I'm falling asleep, I breathe a little faster than I would normally, but I don't start speeding it up more than that. That seems okay until i fall asleep.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP

All in Knots

Re: ASV settings

Post by All in Knots » Tue Nov 13, 2012 8:06 am

OK, so I was a little prematurely excited. Setting the PS to 5 made the air come in a little too forcefully for my taste, so I set it to 4.2. The extra pressure kind of forced me to inhale more efficiently, thereby faster, so I prob. began exhaling a little sooner than normal--but still, the new pressure was coming in. John, you mentioned that ResMed does not have a setting to adjust the breathing rate on this thing, but that the Respironics does. Why do you suppose that ResMed would not include that as a setting? Seems like a breath rate setting is pretty important! I get it, that the machine adjusts a bit to your breathing rate, but it kind of bugs me that it interferes with your exhalation (if you're a slower breather). Actually, so many disciplines try to get you to breathe slowly when relaxed (yoga, etc.), why would a machine want you to do differently?? I'm really thinking about requesting the Respironics.

All in Knots

User avatar
Bons
Posts: 810
Joined: Fri Sep 03, 2010 9:27 am

Re: ASV settings

Post by Bons » Tue Nov 13, 2012 8:30 am

John will correct me if I'm wrong, but I don't think min and max pressure support has anything to do with the speed of response from exhalation to inhalation pressures. Pressure support has nothing to do with timing of breaths.

I think he has a good suggestion is asking if you can be switched to a Phillips Respironics ASV. That machine has a rise rate, which sets the parameter for the timing of pressure changes from EP to IP, and a control for breaths per minute.

That being said, I still had a terrible time adjusting to ASV. I have a very slow awake bpm (8 bpm) and the machine setting cannot be set that low. It is now set for auto (doc originally set it for 12 ). The rise rate is set for 4, which is the slowest response rate available, and that is still too fast for my awake breathing (tech originally had that set at 1, which is the fastest switch time).

So, I did what John suggests. The machine tries to adapt to you, and at the same time your body learns to adjust to it. It is like a carefully executed dance. When I go to bed I take long, deep breaths, and breathe OVER the machine. If it switches to inhalation pressure while I am exhaling, I continue exhaling at the pace I feel comfortable with and the machine will drop back to exhalation. Same with inhaling. The machine will adjust as best it can according to the parameters that are set for it. Even though you don't FEEL in control, for the most part you are. The machine responds to your breathing, and then tries to keep you at the same pace and volume that you set, as long as the machine interprets that rate as healthy. When it determines that it is not healthy (too long a pause and it senses a central apnea) it will then try to get you back in the rhythm of the "dance".

This is not always a pleasant experience (the ASV can really shudder when you scream into a fullface mask ). Once I'm asleep, I breathe a little faster as the ASV and I compromise - I'm the only patient my doctor has who actually breathes faster when asleep than I do when I'm awake.

But in the long run, it works. Two years ago it took me an hour or more to fall asleep with the ASV. Now, it's a matter of minutes. You can still tell from the data when I'm awake during the night because my bpm drops.

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: ASV settings

Post by Madalot » Tue Nov 13, 2012 9:05 am

Bons wrote:I'm the only patient my doctor has who actually breathes faster when asleep than I do when I'm awake.
I'm very similar. We had to set the Trilogy to 7 BPM. Any faster, I felt rushed to inhale before I fell asleep. Once asleep, my AVERAGE BPM is between 12 & 13.

During my last sleep study, when I woke up for the day, I asked the tech if he had raised my BPM because when I woke up, I felt I was being rushed to inhale. Sure enough, he had raised it to 9.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Xney
Posts: 842
Joined: Fri Jul 20, 2012 6:53 pm

Re: ASV settings

Post by Xney » Tue Nov 13, 2012 12:35 pm

The good thing about the resmed ASVs is they're almost automatic, no settings.

The bad thing about the resmed ASVs is they're almost automatic, no settings.


_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Actually a S9 VPAP Adapt, and Respironics M Series Auto BiPAP

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ASV settings

Post by SleepingUgly » Tue Nov 13, 2012 12:42 pm

Think of the ASV as a lover that has to be trained to give you what you need.

...That said, there's a limit to how trainable some lovers are.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly