EPR setting S9 apap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Britishbulldog
Posts: 42
Joined: Sun Aug 30, 2009 1:41 pm
Location: UNITED KINGDOM

EPR setting S9 apap

Post by Britishbulldog » Sat Aug 18, 2012 12:36 pm

Hi,

Could anybody explain the EPR settings on the S9 apap

Currently set at 1

goes upto 3 then says patient ???

What is the best setting please,

Regards,

British Bulldog.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Mirage Quattro Air -Full Face Mask

User avatar
Pugsy
Posts: 65269
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR setting S9 apap

Post by Pugsy » Sat Aug 18, 2012 12:56 pm

When it says Patient that means that you can change it to whatever you want.
All Patient means is that you have access to changing the EPR setting from 1 to 2 to 3.

EPR settings are a per cm reduction in pressure during exhale...a setting of 1 is a 1 cm reduction..a setting of 3 is a 3 cm reduction. It won't ever go below 4 cm though....so if you used a 6 cm of pressure with 3 ERP..it won't drop to 3 during exhale instead it just drops to 4.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.