Measured CPAP power consumption

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Measured CPAP power consumption

Post by archangle » Sun Jul 22, 2012 2:55 pm

We often get questions about actual power consumption of CPAP machines.

For those of you with measurement capability, what do you get?

Here's one session data.

ResMed S9 AutoSet with heated humidifier. No heated hose.

Auto 18-20, EPR=3, humidifier=4. Fairly humid night. Room temp around 75F.

Kill a Watt power meter measuring AC power into the power brick.

4:32 hours.
0.15 Kilowatt (Correction Kilowatt hours)
Average power 34 watts.

Looking at the instantaneous power readings as I inhale and exhale and the heater runs, power varies all over the place. Between 15 and 90 watts read on the meter.
Last edited by archangle on Sun Jul 22, 2012 8:56 pm, edited 1 time in total.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Measured CPAP power consumption

Post by avi123 » Sun Jul 22, 2012 3:28 pm

What is a "Kill a Watt power meter" ?

And what's the meaning of this:

4:32 hours.
0.15 Kilowatt
Average power 34 watts.



You don't measure Watts but Amperes and Voltage and calculate the Watts:

Looking at the instantaneous power readings as I inhale and exhale and the heater runs, power varies all over the place. Between 15 and 90 watts read on the meter.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

123.Shawn T.W.
Posts: 748
Joined: Sun Jun 24, 2012 12:39 pm
Location: Cochise County AZ

Re: Measured CPAP power consumption

Post by 123.Shawn T.W. » Sun Jul 22, 2012 3:37 pm

"I am a man of peace, but if war comes to my door it will find me home." - Winston Churchill

User avatar
napstress
Posts: 462
Joined: Sat Apr 09, 2011 6:59 am

Re: Measured CPAP power consumption

Post by napstress » Sun Jul 22, 2012 4:20 pm

This is interesting, archangle. I unplug pretty much anything unpluggable when I'm done using them, with the exception of lamps, refrigerator, oven, in an effort to avoid phantom energy drain. I've often wondered to what degree I was canceling that out by using the CPAP all night long.
Epworth Sleepiness Scale: 14
Diagnostic study: overall AHI: 0.2 events/hour; overall RDI: 45 events/hour
Titration study: AHI: 6.1; RDI: 27; CPAP pressures: 5-8cm

Not-tired behind my eyes and with a clear, cool head!

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Measured CPAP power consumption

Post by avi123 » Sun Jul 22, 2012 6:23 pm

123.Shawn T.W. wrote:http://www.p3international.com/products ... 00-ce.html

Measures consumption
Thanks for showing the Kill a Watt.

Image
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

ozze_dollar
Posts: 611
Joined: Wed Jun 08, 2011 5:20 pm
Location: Sydney,Australia

Re: Measured CPAP power consumption

Post by ozze_dollar » Sun Jul 22, 2012 6:28 pm

I get a deduction from my power bill because I am a cpap user. It is about $19 a qtr. Dont know if it covers it or not.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Measured CPAP power consumption

Post by chunkyfrog » Sun Jul 22, 2012 6:32 pm

We used the same device to determine our power cost for distilling water.
--About 13 cents a gallon in winter; 20 cents in summer.
(seasonal electric rates are different--especially with electric heat)


0

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Measured CPAP power consumption

Post by avi123 » Sun Jul 22, 2012 7:19 pm

Here is a Typical Power Consumption of a household. What counts is Kilowatts Hours or Annual kWH:

http://www.oksolar.com/technical/consumption.html

If I take archangel's measurement numbers of:

4:32 hours.
0.15 Kilowatt = 150 Watts
Average power 34 watts

Then it means that the CPAP with the humidifier ON, is using 150 watts. If I multipy it by the 7 hours that I sleep nightly
and by 365 days in a year I get:

150 x 7 x 365 = 383,250 Watts = 383 kiloWatts Hours (KWH) per year.

So it consumes annually as a dishwasher which consume 363 KWH annually or twice as a window fan consuming 170 KWH annually.

If I assume that the average price of residential electricity was 12¢/kWh in the U.S. in December 2011, and ranged from 7.5¢ in Idaho to 36¢ in Hawaii, then the CPAP electrical consumption costs me:

383 KWH x 12 cents/KWH = 4,596 cents = app $46 annually.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: Measured CPAP power consumption

Post by Lizistired » Sun Jul 22, 2012 8:32 pm

Breathing all night.... Priceless!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Measured CPAP power consumption

Post by archangle » Sun Jul 22, 2012 9:02 pm

avi123 wrote: 4:32 hours.
0.15 Kilowatt = 150 Watts
Average power 34 watts
Sorry about that. It was 0.15 Kilowatt hours. Blame it on dain bramage.

For 4:32 hours, that's 34 watts average power usage.

Roughly 0.3 kWh for one night's use. 3-5 cents worth of electricity a night. $10 to 15 a year.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Measured CPAP power consumption

Post by archangle » Sun Jul 22, 2012 10:43 pm

Further information:

My S9 AutoSet power supply draws about 2 watts when the machine is turned off.

My PRS1 Auto power supply appears to draw less than 1 watt when turned off.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Measured CPAP power consumption

Post by avi123 » Sun Jul 22, 2012 11:16 pm

Further information

The power supply (Brick) for the S9 comes in 90 watts and 30 watts capacities. I assume that the 90 watts is for use with the humidifier. On the S9 Autoset body it's written 24 volts DC and 3.75 amps (for system). If I multiply the two numbers I get 90 voltampers, which jibes with the 90 watts power supply.

So I am back to my first calculation and I don't need the Kill-a-Watts gadget.

The annual cost of electricity to run the S9 Autoset with the humidifier:

90 watts x 7 hours x 365 days = 229,950 watts hours= 230 KWH
230 KWH x 12 cents/KWH= 2,760 cents = $27.6 annually.
If I don't use the humidifier then its one third as much, about $9.5 annually.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Measured CPAP power consumption

Post by archangle » Mon Jul 23, 2012 12:10 am

avi123 wrote:Further information

The power supply (Brick) for the S9 comes in 90 watts and 30 watts capacities. I assume that the 90 watts is for use with the humidifier. On the S9 Autoset body it's written 24 volts DC and 3.75 amps (for system). If I multiply the two numbers I get 90 voltampers, which jibes with the 90 watts power supply.

So I am back to my first calculation and I don't need the Kill-a-Watts gadget.
The label rating on almost any electrical device will be a maximum number. Actual usage is almost always less.

This thread is about actual measured power consumption, not the ratings. In my case, it's about 1/3 of the rated maximum. If I were deciding what size battery I need for emergency back, I'd need a battery 3 times a large if I wanted to use the rated maximum.

If I want to figure out actual yearly cost, I need to use actual consumption, not rated maximum. You can use the rated maximum, but realize you're probably overestimating actual cost.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Measured CPAP power consumption

Post by Goofproof » Mon Jul 23, 2012 12:32 am

ozze_dollar wrote:I get a deduction from my power bill because I am a cpap user. It is about $19 a qtr. Dont know if it covers it or not.
That means your niebhors are paying $15 extra a quarter, because your electric company allows you a discount that is out of line with your use, Socialism at it's best. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
snoozysue
Posts: 287
Joined: Sun Oct 17, 2010 2:32 am
Location: Australia

Re: Measured CPAP power consumption

Post by snoozysue » Mon Jul 23, 2012 2:15 am

Once again things seem to be much more expensive Down Under (Aust). Our electricty went up on 1st July due to the Carbon Tax. I received a letter from my supplier today with the new rates.
I have a "use of time" meter and now pay:
Peak Consumption (2pm - 10pm) - 53.59 cents
Shoulder Consumption (8am - 2pm) - 21.82 cents
Off Peak Consumption (10pm - 8am) - 13.42cents

Luckily when I sleep its Off Peak - not that would stop me sleeping now that I actually sleep!