I change EPR to 1 and AHI is now 2.0

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
VikingGnome
Posts: 591
Joined: Fri Feb 03, 2012 6:46 pm
Location: Mississippi

I change EPR to 1 and AHI is now 2.0

Post by VikingGnome » Tue Mar 27, 2012 11:14 am

My AHI's had been creeping up to 6-7 on some nights. I decided to changed EPR from 2 to 1 to see if it helped. My AHI this morning is 2.0.

When I had my CPAP titration, the tech started at 10 and finished at 14. When my sleep doc saw me two weeks later, she said she wished the tech had gone up to 15 because at 14, my apneas who not totally gone. So she set my prescription at 15 although I was never at 15 on the study. It is possible I really need 16 but at 15 with EPR at 2, it drops to 13 on exhale. I really wish I had an AutoSet but my doctor doesn't "believe in them."

I'm on South Beach Diet now and have lost 12 pounds in 2.5 weeks. If I lose enough weight I'm going to need to have settings adjusted. I only needed 10 when I was 20 pounds lighter. If I had an APAP, I would be in fine shape. Sigh! My DME had never dispensed an AutoSet she said. That means all the sleep docs in this area are not prescribing APAP.

_________________
Mask: AirFit™ F10 For Her Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure Setting 22/18; Pregnancy pillow to keep me sleeping on side

User avatar
Pugsy
Posts: 65142
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I change EPR to 1 and AHI is now 2.0

Post by Pugsy » Tue Mar 27, 2012 11:23 am

There are ways to get the S9 Autoset as a backup machine if you are willing to go out of pocket for the costs. I am sure you are aware of this. Could you or would you be comfortable doing it on your own?

Also...Just because your particular DME has not seen to dispense an APAP doesn't necessarily mean that all doctors and all DMEs in your area are stuck on CPAPs.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
VikingGnome
Posts: 591
Joined: Fri Feb 03, 2012 6:46 pm
Location: Mississippi

Re: I change EPR to 1 and AHI is now 2.0

Post by VikingGnome » Tue Mar 27, 2012 11:36 am

No. My finances are such that purchasing an S9 Autoset on my own would not be possible. I see my sleep doc for 60 day follow-up on April 4. Maybe I can convince her that with my weight loss and continuing diet, that I really need an S9 AutoSet. If she wrote a new Rx, the DME could switch out the S9 Elite I have because I'm on Medicare.

_________________
Mask: AirFit™ F10 For Her Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure Setting 22/18; Pregnancy pillow to keep me sleeping on side

User avatar
Pugsy
Posts: 65142
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I change EPR to 1 and AHI is now 2.0

Post by Pugsy » Tue Mar 27, 2012 11:52 am

The weight loss and possible variables in pressure could present a good argument for APAP.
Of course it could always be used in CPAP mode. While APAP isn't always the "best" thing for everyone it sure is nice to have it available. My doctor didn't/doesn't offer APAPs often either. I happen to need it for variable pressures in REM. I bought my own machine but when I showed my reports documenting that APAP was beneficial for my needs then APAP was added to my records. I do have a DME for masks and supplies but have never used my insurance for a machine. My DME always tries to offer me a new APAP now. I have since bought several machines on my own and everyone of them I got such a good deal that my out of pocket was less than my copays with my insurance (which is a Medicare Advantage plan now) but for cpap stuff it pays exactly like Medicare.

Good luck and good job on the reduction in AHI with a minor change in the EPR. That works for a lot of people.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13356
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: I change EPR to 1 and AHI is now 2.0

Post by LSAT » Tue Mar 27, 2012 1:15 pm

I think too much of an issue is made about autos..Yes it would be nice to have one, but a good machine like the Elite is adequate for most people.Many with autos set their machine to CPAP anyway. Pressures do not change that much. My pressure has been in the11-12 area for as long as I have been on my CPAP 'brick'. Any time I have checked my data with an autoset,my 95% figure has been about 11.5. I rarely check my AHi, but when I do it averages about 4-6. I'm happy with that. I feel good...I sleep good. What would I gain by driving myself nuts trying to lower my AHI?...Nothing!
Last edited by LSAT on Tue Mar 27, 2012 5:19 pm, edited 1 time in total.

josh73
Posts: 61
Joined: Sat Jan 07, 2012 10:37 am

Re: I change EPR to 1 and AHI is now 2.0

Post by josh73 » Tue Mar 27, 2012 1:30 pm

LSAT, i completely agree with you on the AHI thing. There are mornings i look at my machine and my AHI is 2.0 or lower and i feel OK, but not great. And there are days that my AHI is 5 or 6 and i feel GREAT.

Also, I asked for an APAP when i was first diagnosed with sleep apnea and my doctor wouldn't prescribe one since my titration was successful and i didn't have any other problems besides my AHI of 15. My sleep tech told me that an APAP rental is available if changes are made, ie: weight loss. I started in mid January and have lost around 20 lbs so far but i'm not going to worry about seeing about a APAP rental until next January at the earliest.

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: I change EPR to 1 and AHI is now 2.0

Post by Papit » Tue Mar 27, 2012 4:02 pm

Pugsy wrote: . . . Of course it (your Autoset) could always be used in CPAP mode. While APAP isn't always the "best" thing for everyone it sure is nice to have it available. My doctor didn't/doesn't offer APAPs often either. I happen to need it for variable pressures in REM. I bought my own machine but when I showed my reports documenting that APAP was beneficial for my needs then APAP was added to my records. . . . Good luck and good job on the reduction in AHI with a minor change in the EPR. That works for a lot of people.
I'm using an Autoset in CPAP mode since I began three months ago. Please help me understand better when it might make sense, if ever, for someone to switch their Autoset machine from CPAP over to Auto mode. What are the most common reasons to do that?

Also, you might explain how the Autoset is useful, and why you need it, for "variable pressures" in REM. What was it in the printouts you brought in that showed that the Autoset set for Auto was beneficial for your needs? --thanks, Mike

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: I change EPR to 1 and AHI is now 2.0

Post by Uncle_Bob » Tue Mar 27, 2012 4:39 pm

Papit wrote:
I'm using an Autoset in CPAP mode since I began three months ago. Please help me understand better when it might make sense, if ever, for someone to switch their Autoset machine from CPAP over to Auto mode. What are the most common reasons to do that?
someone using an AUTO in CPAP mode may switch to APAP mode in an attempt to self titrate themselves to see if a changed CPAP pressure may be needed.

User avatar
Pugsy
Posts: 65142
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I change EPR to 1 and AHI is now 2.0

Post by Pugsy » Tue Mar 27, 2012 4:45 pm

Papit wrote:Also, you might explain how the Autoset is useful, and why you need it, for "variable pressures" in REM. What was it in the printouts you brought in that showed that the Autoset set for Auto was beneficial for your needs?
Rather than hijack this thread with my reports and stuff I answered your questions here.
Look on last page if you don't want to read my whole boring story of how I went to bilevel.
viewtopic/t71442/viewtopic.php?f=1&t=71 ... te#p656179

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Tue Mar 27, 2012 7:16 pm, edited 2 times in total.
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13356
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: I change EPR to 1 and AHI is now 2.0

Post by LSAT » Tue Mar 27, 2012 5:16 pm

Papit wrote:

I'm using an Autoset in CPAP mode since I began three months ago. Please help me understand better when it might make sense, if ever, for someone to switch their Autoset machine from CPAP over to Auto mode. What are the most common reasons to do that?

Also, you might explain how the Autoset is useful, and why you need it, for "variable pressures" in REM. What was it in the printouts you brought in that showed that the Autoset set for Auto was beneficial for your needs? --thanks, Mike
If you are using CPAP mode and your AHI is low, but you feel crappy, you might benefit from a titration with an auto...or...
If your AHI is high and you don't know why.

User avatar
teachcsg
Posts: 189
Joined: Sun Mar 11, 2012 2:39 pm

Re: I change EPR to 1 and AHI is now 2.0

Post by teachcsg » Tue Mar 27, 2012 10:47 pm

Just from experience as an RT. Auto cpap's are great but dont think they will solve all your problems. I have seen many patients on auto cpap's where their AHI was still >5. So dont feel like if you dont get an auto cpap your therapy is not any good. There really is not one size fits all in sleep apnea land.
Cesar Garza, RRT
Registred Respiratory Therapist
http://www.lastminuteceus.com