Res Med Mirrage Quattro Full Face Mask Help!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
reffus1881
Posts: 19
Joined: Sat Jan 07, 2012 9:04 pm

Res Med Mirrage Quattro Full Face Mask Help!

Post by reffus1881 » Thu Feb 02, 2012 8:06 pm

First night with machine S9 Res Med Elite/Heated humidifier.Full face mask is Mirage Quattro Medium. Problem: The valve where the air enters the face mask is closing when i put the mask on Hence no pressure into mask. i take the mask off and the valve opens. The DME set up the machine everything worked fine in his office. So i guess i need to know what is the trick for keeping the valve open.
Thanks

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: Res Med Mirrage Quattro Full Face Mask Help!

Post by RandyJ » Thu Feb 02, 2012 9:07 pm

The little flap on the elbow between the hose and the mask closes when the air comes through (when machine is on); it's the anti-asphyxiation valve. It should be open when machine is off. That is how it is supposed to be so that if your machine loses power and stops running during the night, you would get fresh air through the valve instead of re-breathing your own exhaled carbon dioxide.

That said, if you are not getting air into the mask from your hose, something else may be wrong.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
LSAT
Posts: 13401
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Res Med Mirrage Quattro Full Face Mask Help!

Post by LSAT » Thu Feb 02, 2012 9:40 pm

Randy is absolutely right...as long as you feel exhaust coming out of the vents, you are getting plenty of air. When you turn on the CPAP the flap shuts to keep the pressure at your prescribed setting. If the power goes off the vent opens so you can breathe. It sounds like everything is working properly.