Should I try increasing my pressure? Please advise

Post by cake321 » Thu Jan 26, 2012 10:01 pm

I've been on CPAP since late September. It seemed like when I first started using it, I could definitely feel the difference. It was easy to get out of bed, even when I only slept about 5-7 hours, which is what I normally get during the week. I've noticed over the last few weeks, it has once again become hard to get out of bed, and I find myself still getting tired throughout the day, especially in the afternoon (this never totally went away, but I definitely had improvement). I am not doing anything differently. I have no leaks, and my spouse confirms that I do not snore. I am wondering if I should try raising my pressure. It is currently at 11. I cannot afford to go back to the doctor as I no longer have insurance. The only thing I can possibly think of is that I may have gained 5-10 lbs over the last few months. Could this matter? I was already about 30-40 lbs overweight to begin with.

Thanks for any advice. I just can't stand feeling this way again.

cake321
Posts: 35
Joined: Mon Aug 08, 2011 1:59 pm

Re: Should I try increasing my pressure? Please advise

Post by cake321 » Thu Jan 26, 2012 10:19 pm

Some additional details that I thought of after I posted... I have no problem falling asleep or staying asleep. In fact not once have I ever woken up and had to get up and pee, which was a common problem before I started CPAP. Thanks again for any help.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Should I try increasing my pressure? Please advise

Post by BlackSpinner » Thu Jan 26, 2012 10:25 pm

I don't know if your machine has data. Someone else will help with that.
If you decide to try that up it by the smallest amount the machine will allow and leave it there for a week. Keep a journal.
There have been studies done that people are quite capable of finding their own pressure needs even without data.

The only caveat is that if you feel worse the pressure may be triggering centrals - in which case you reduce it again.

However since you use a nasal only mask you may want to try a chinstrap first. mouth breathing leaks therapy.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

cake321
Posts: 35
Joined: Mon Aug 08, 2011 1:59 pm

Re: Should I try increasing my pressure? Please advise

Post by cake321 » Thu Jan 26, 2012 10:29 pm

BlackSpinner wrote:I don't know if your machine has data. Someone else will help with that.
If you decide to try that up it by the smallest amount the machine will allow and leave it there for a week. Keep a journal.
There have been studies done that people are quite capable of finding their own pressure needs even without data.

The only caveat is that if you feel worse the pressure may be triggering centrals - in which case you reduce it again.

However since you use a nasal only mask you may want to try a chinstrap first. mouth breathing leaks therapy.
Thank you. I do use a chin strap, as well.

User avatar
porete
Posts: 230
Joined: Fri Dec 30, 2011 5:58 pm

Re: Should I try increasing my pressure? Please advise

Post by porete » Thu Jan 26, 2012 10:48 pm

I wouldn't increase the pressure without your doctor's consent. If your machine is data capable, you can take the data reports to your doctor and both of you can figure out what would be the best step for you to take.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Additional Comments: Pressure Range: 8.5-14, EPR: 3

User avatar
porete
Posts: 230
Joined: Fri Dec 30, 2011 5:58 pm

Re: Should I try increasing my pressure? Please advise

Post by porete » Thu Jan 26, 2012 10:50 pm

I looked your machine up on cpaptalk.com and it IS data capable.
"For therapy tracking, the SleepStyle 234 CPAP is compatible with Compliance Maximizer software, using F & P's Compliance Download Cable to transfer data between machine and computer."

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Additional Comments: Pressure Range: 8.5-14, EPR: 3

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Should I try increasing my pressure? Please advise

Post by Pugsy » Thu Jan 26, 2012 11:00 pm

The SleepStyle 234 has only hours of use..compliance hours. No other useful data.
http://www.fphcare.com/osa/cpap-solutio ... e-200.html

The only thing you have to go on is how you feel. It is unlikely that things would have changed so quickly to cause you to need more pressure but it would not be impossible. You have a very basic machine and there is no way to obtain any data to see what is going on. I would think something else like leaks compromising therapy. If there is no money or insurance for a doctor visit then you have to decide if you want to try a little more pressure. If you do decide to try it...go up slowly. Also before you try it be sure and check all your gear to make sure everything is fine...no tiny holes in hose...no loose fittings...mask fittings, etc. Just to make sure you aren't losing pressure via a leak that you are unaware of.

A few extra pounds shouldn't make a huge difference (heck if we lose that much we don't often get to reduce the pressure) but again, it wouldn't be impossible especially if you were perhaps borderline needing a bit more.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

cake321
Posts: 35
Joined: Mon Aug 08, 2011 1:59 pm

Re: Should I try increasing my pressure? Please advise

Post by cake321 » Fri Jan 27, 2012 7:27 am

Pugsy wrote:The SleepStyle 234 has only hours of use..compliance hours. No other useful data.
http://www.fphcare.com/osa/cpap-solutio ... e-200.html

The only thing you have to go on is how you feel. It is unlikely that things would have changed so quickly to cause you to need more pressure but it would not be impossible. You have a very basic machine and there is no way to obtain any data to see what is going on. I would think something else like leaks compromising therapy. If there is no money or insurance for a doctor visit then you have to decide if you want to try a little more pressure. If you do decide to try it...go up slowly. Also before you try it be sure and check all your gear to make sure everything is fine...no tiny holes in hose...no loose fittings...mask fittings, etc. Just to make sure you aren't losing pressure via a leak that you are unaware of.

A few extra pounds shouldn't make a huge difference (heck if we lose that much we don't often get to reduce the pressure) but again, it wouldn't be impossible especially if you were perhaps borderline needing a bit more.

The pressure settings go up in increments of 0.5, so the next highest setting would be 11.5. Do you all suggest going up in .5 increments or 1.0? Also, how many days should I wait before upping it if I feel no improvement?

User avatar
Kairosgrammy
Posts: 529
Joined: Mon Jan 02, 2012 8:13 am

Re: Should I try increasing my pressure? Please advise

Post by Kairosgrammy » Fri Jan 27, 2012 7:54 am

Is your machine data capable? If so, have you checked your ahi, rera's etc. What are those numbers.

Gaining weight can make a change in OSA sometimes but at other times, it is inconsequential. There are people of normal weight who have severe OSA and in fact there is the arguement of "Which came first" OSA and then weight gain or weight gain and then OSA. There is a lot of research out there that would favor OSA & then weight gain and with me, I would say there is a lot of truth, Even if I were to lose weight and get down to a healthy weight, I will still probably need cpap for the rest of my life and that's fine. I feel better.

If I had to change something, I would first work on increasing # of hours sleeping. As the old saying goes, we can't be teenagers forever living on 4 or 5 hours of sleep.
cake321 wrote:Some additional details that I thought of after I posted... I have no problem falling asleep or staying asleep. In fact not once have I ever woken up and had to get up and pee, which was a common problem before I started CPAP. Thanks again for any help.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: I'm starting to use sleepyhead.

User avatar
freshair
Posts: 129
Joined: Tue Oct 26, 2010 10:05 am
Location: Kentucky

Re: Should I try increasing my pressure? Please advise

Post by freshair » Fri Jan 27, 2012 8:21 am

I have had my pressure increased twice since triation...I requested another...they said no...the reason is if my pressure is too high it could shut off my breathing.....so in may when i return hm i'll setup another study....she did tell me to change my c-flex to setting one and also turn off the system one humification selection...its working better now....me and my sleep tech are 1000 miles apart.....tks--FA

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Should I try increasing my pressure? Please advise

Post by archangle » Fri Jan 27, 2012 9:43 am

freshair wrote:I have had my pressure increased twice since triation...I requested another...they said no...the reason is if my pressure is too high it could shut off my breathing.....so in may when i return hm i'll setup another study....she did tell me to change my c-flex to setting one and also turn off the system one humification selection...its working better now....me and my sleep tech are 1000 miles apart.....tks--FA
Geez, what they're saying is "We ripped you off by giving you a non-data capable CPAP machine. Now we're going to refuse to adjust your pressure upwards since it might give you central apneas, and the machine we screwed you with can't tell the difference. How would you like another uncomfortable, expensive, sleep test that may not tell us much because you may not sleep the same way you do at home?"

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

Guest

Re: Should I try increasing my pressure? Please advise

Post by Guest » Fri Jan 27, 2012 10:47 am

my insurance pays 100%

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Should I try increasing my pressure? Please advise

Post by Pugsy » Fri Jan 27, 2012 12:55 pm

With no data to rely on....I would only increase the pressure in 0.5 cm increments. I would leave it for 7 days. I would start a journal and document how I felt and how many hours of sleep I got. Did I wake often during the night? If so, for what reason.

Keep a journal so have something to correlate things with. Hours of sleep are critical....so is waking often for any reason...it messes with how we feel.

You will need to be very thorough with your journal... I know that it is real easy to forget how bad or good I felt 2 days ago... Looks for patterns. Don't expect a sudden change in how you feel with a small pressure change. We don't sleep the same each night even if we don't change anything.

If you start feeling horribly worse....discontinue the experiments and go back to RX pressure.

Finally, if your finances would permit it at all or if you had a friend with a full data machine you could borrow....you need the data available to really see if you even need a pressure increase...it may not be the pressure causing your problems...it might be something else. Could you scrape together a couple hundred dollars to get a machine that offered data? Craigslist sometimes has some good deals. Just have to shop smart. Even if it was limited to on screen data (like the ResMed S8) it would be better than flying blind like you are now.